I recently purchased a Siglent SDS 1102CML (100Mz) scope and was using it to look at some noise on a power supply. I noticed two stable peaks in the FFT at 50Mz and 100Mz which remained even when I removed the probes.. I am relatively new to scopes and am wondering if it is typical to have FFT peaks on harmonics of the scope frequency. I have attached a screenshot. Any suggestions are welcome.
Ron
-
I recently purchased a Siglent SDS 1102CML (100Mz) scope and was using it to look at some noise on a power supply. I noticed two stable peaks in the FFT at 50Mz and 100Mz which remained even when I removed the probes.. I am relatively new to scopes and am wondering if it is typical to have FFT peaks on harmonics of the scope frequency. I have attached a screenshot. Any suggestions are welcome.
Ron
Welcome to the forum.
It is normal to see all sorts of garbage if the probes are not terminated to a DUT.
Internally the SDS1000 series have 3 clocks, 25, 30 and 100MHz. (check the Service manual)
Even the lowest BW model in the SDS1**2C*L has the same clocks.
BTW to check PSU noise/ripple one would normally use AC input coupling and mV/div amplitude setting, then adjust timebase and triggering to analyse the displayed waveform.