Hi
I was comparing my new to me sds1104x-e with my other scopes and i noticed that in some setting the scope won't trigger in a repeatable way on the rising edge of the probe compensation signal and shows a flat line but if i press more or less any key or knob it will trigger correctly but if i disconnect and reconnect the probe i have a 50/50 chance of not triggering
It does that on the auto trigger but not on the normal trigger
What am i doing wrong?
 

Online TheDefpom

  • Frequent Contributor
  • **
  • Posts: 719
  • Country: nz
  • YouTuber Nerd - I Fix Stuff
    • The Defpom's Channel
Re: Siglent 1104x-e not triggering on probe compensation signal
« Reply #1 on: May 22, 2024, 07:05:07 pm »
Make sure you have hold off turned off or down to it’s minimum setting.
Cheers Scott

Check out my Electronics Repair, Mailbag, or Review Videos at https://www.youtube.com/TheDefpom
 

Offline LorisermejoTopic starter

  • Newbie
  • Posts: 3
  • Country: it
Re: Siglent 1104x-e not triggering on probe compensation signal
« Reply #2 on: May 22, 2024, 07:50:36 pm »
Well
I remember I modified it some days ago but then i disabled it
I checked now and reports "holdoff close" (probably meaning minimum)
I pressed the button and shows 50ms
I pressed again bringing it back to " holdoff close" and now it works
Thanks
I thought it was a bug
 