Author Topic: Siglent 1202X-E Fan Noise  (Read 2188 times)
Offline Roger Need
« on: December 24, 2018, 09:25:03 am »
I am a new subscriber to this forum.  I read many of the posts concerning the Siglent SDS1202X-E oscilloscope before ordering one 10 days ago from Amazon Canada (no local dealers in Western Canada).  There were a few comments about Siglent fan noise but mostly they were about other models like the 1104. .  So I hope those of you that own or sell these scopes can help me out.
The unit that I received is a BB model but it must have been in stock for awhile because the calibration letter was dated last February.  It was new because the power on screen showed that I was the first to power it up.  The firmware was 5.1.3.17 which is not the latest September 2018 release so I will have to update if I don't return the unit.  The scope works okay and seems quite accurate in terms of the frequency counter, voltage measurements and FFT operation.  Offsets are pretty good on the 500 uV scale.
The problem is the noise.  When I power it up in the morning it has some fan noise which is fairly noticeable and at a higher level than I expected.  I have a quiet lab without any other fans running most of the time.  What I thought I noticed was that the noise was getting louder after a few minutes and then about 10-15  minutes after start there is a whine which may be coming from the fan or the power supply.  In order to verify my observation I made a recording by placing a microphone near the fan exhaust grill on the right hand side.   I recorded the ambient noise with the 1202X-E turned off.   Then a 15 minute recording was made with the scope powered up and nothing connected. I then analyzed the recording with spectral analysis software and confirmed what I had been observing.   The plot below shows the difference from ambient noise after at 2, 9 and 13 minutes.  One can clearly see the noise slowly getting worse and several spectral peaks which are the "whine" I was referring to earlier.
As it stands it is not enjoyable to use this instrument.  I will either have to build a sound-proof enclosure around it (with quiet fans to provide airflow) or return it to Amazon for exchange or refund.  If this is the way this product normally operates there is no point in going through the hassle of return shipping just to get the same issue again.
I hope I can get some feedback on these questions.
1. Is this typical for the 1202X-E or do I have a bad fan?
2. Why does it get worse as the unit warms up?  The fan is still pushing the same CFM of air - it is just warmer.
3. Could the spectral peaks be something to do with audio noise from the switching power supply?
4. It shipped without a CD in the Quick Start guide.  Anything of value on the CD that can't be downloaded?
Thanks for reading and any comments....
« Last Edit: December 24, 2018, 01:52:08 pm by Roger Need »
Offline Performa01
  • Country: at
Re: Siglent 1202X-E Noise Issues
« Reply #1 on: December 24, 2018, 10:23:19 am »
First I have to say that most modern test instruments are pretty annoying for someone used to a quiet lab. Tiny fans can be found in just about any (semi-)modern test gear...
The Siglent DSOs are not very quiet, but also not particularly loud in my book and I've never experienced a noticable change in volume or tone over time. The fan is just blowing air, that produces some (pink?) noise, but there are no particularly annoying tones like hum or whine.
I'm currently not set up for sound recordings, but if you provide a short sample for me to listen, then I can tell you whether it's normal or not.
« Reply #2 on: December 24, 2018, 10:42:36 am »
The Siglent DSOs are not very quiet, ......
Latest SDS1202X-E’s are quite quiet, somewhat more than early units.
They don’t have as much internal HW as 4ch units so can use a lower capacity quieter fan.
Avid Rabid Hobbyist
Offline TheSteve
  • GHz
« Reply #3 on: December 24, 2018, 11:14:20 am »
Welcome to the forum(I am also in BC).
I don't know how loud it should be, but if you find it irritating I would return it as it will forever annoy you. The fact that the noise changes is also a concern unless the fan is just breaking in.
VE7FM
The following users thanked this post: Roger Need
Offline DaJMasta
    • medpants.com
« Reply #4 on: December 24, 2018, 11:14:50 am »
Are you sure the fan isn't speed controlled?  They certainly can be, and it would explain the warm up sound difference.  Also would be useful to look at the raw audio plots rather than difference plots, and it's worth making sure that your recording device doesn't have any auto-gain enabled or any noise gating, compression, or other features enabled.  Unlikely for a standalone mic and sound equipment, but quite likely with a device with an integrated mic and software.
When it comes down to it, though, if the sound bothers you, you really only have two options:  return it and either try another unit or try another kind of scope, or go in there and replace the fan with a quieter one.  As mentioned, it's atypical for fan noise to be a primary concern for test equipment, especially when you start moving towards higher end stuff, so while a small amount of noise on a scope may be noticeable in the quiet office a hobbyist works in, it's going to be well below ambient in most electronics lab settings.
Online Rerouter
« Reply #5 on: December 24, 2018, 11:24:43 am »
At least for my 1104X-E, I can best describe the fan noise as noticeable, but not loud from a 1m distance,
Offline Roger Need
Re: Siglent 1202X-E Fan Noise
« Reply #6 on: December 24, 2018, 11:27:24 am »
I edited the recorded file to show what happens when turned on, after 2, 9 and 13 minutes and when the fan is blocked.  I annotated the waveform plot so you can see the events and the increasing noise.  Hoefully the recording attaches to this post.  If not it is on Google Drive here >> https://drive.google.com/open?id=14QU10jiJB_haCWSX3yFGMHBqU_aRLkhO
« Last Edit: December 24, 2018, 01:56:36 pm by Roger Need »
« Reply #7 on: December 24, 2018, 11:49:21 am »
For factory ‘new’ 2 and 4ch units there is a noticeable difference in fan noise levels....for October ‘18 produced units.
Avid Rabid Hobbyist
Online Rerouter
« Reply #8 on: December 24, 2018, 11:52:23 am »
So was a different model fan used? or something different, as mine was pre august.
« Reply #9 on: December 24, 2018, 12:01:21 pm »
Couldn’t say as I haven’t investigated it, I just know the 2ch models are now quieter.....I run every unit before dispatch.
Avid Rabid Hobbyist
Offline Roger Need
« Reply #10 on: December 24, 2018, 12:06:24 pm »
Thank you for this information.  How can one tell if the unit is October 2018 or later?  By certification letter or serial number?  Are the latest one still using BB in the serial number?
« Last Edit: December 24, 2018, 01:56:54 pm by Roger Need »
« Reply #11 on: December 24, 2018, 12:16:42 pm »
The Cal cert is probably just a few days older than production date.
Recent SN# sold: SDS1ECDQ2R**** so CD is latest HW version in my stock.
Avid Rabid Hobbyist
Offline Performa01
  • Country: at
« Reply #12 on: December 24, 2018, 12:23:07 pm »
Well, this doesn't sound abnormal to me, especially considering that the mic is close to the fan.
I think it might just be your first more modern unit with a fan - I felt the same annoyance when I got my first equipment with fan, introducing a lot of noise to the previously very quiet lab. And the smaller units are even worse - have you ever heard e.g. a Keithley 2001 DMM? That really has some whining tone to it...
Then there is the change of the tone on your unit. I remember that there have been complaints about fan noise in the past and this has subsequently been discussed in Siglent R&D. Back then I mentioned that a temperature controlled fan doesn't add significant cost and maybe this is what has been implemented by now (since Tautech mentions that recent SDS1202X-E are less noisy than the previous units).
The problem is, you will not find anything less noisy out of the box, except a faceless USB-scope, and the same problem will arise with pretty much any modern gear you're going to add to your lab, be it a signal/function generator, spectrum analyzer or even a DMM. Ironically, there are quite a few power supplies without fan, either with a big heatsink for the linear ones or SMPS designs with low power dissipation in the first place.
Anyway, fan noise is an ongoing topic here, for pretty much all devices. Fanless designs are really rare nowadays - and certainly not found in the class of an SDS1202X-E DSO. Users are sometimes "tuning" their gear, fitting less noisy fans, adding temp controllers and/or mechanic decoupling from the chassis.
Offline Roger Need
« Reply #13 on: December 24, 2018, 12:31:19 pm »
Thank you to everyone that has taken the time to respond to my questions. I appreciate all the information and suggestions.  The features and performance in the 1202X-E are what I am looking for but at this point I think it best to send this unit back to Amazon for a refund as the noise level is too high to enjoy using this particular scope.   As Performa01 pointed out fan noise is a fact of life but hopefully the latest hardware version will be better.  The noise when it starts and runs for a few minutes is OK but not so great after 15 minutes
Tautech posted that the units from October 2018 are quieter and he should know because he tests each one leaving his shop.  I will try and find one of the CD or later versions from a supplier in Eastern Canada.  One wonders what else they might have changed besides the fans.
Wishing you all a Merry Xmas and a Happy New Year....
« Last Edit: December 24, 2018, 01:57:10 pm by Roger Need »
« Reply #14 on: December 24, 2018, 12:41:32 pm »
As mentioned to Roger via PM’s, I’m reasonably sure another member has added a series resistor to the fan to slow it some and reduce fan noise.
This topic would benefit from other owners experience if they would care to share what they’ve done.
There’s also adding rubber mounts to reduce resonant coupled noise........also mentioned by Performa01.
Avid Rabid Hobbyist
Offline Roger Need
« Reply #15 on: December 24, 2018, 12:57:05 pm »
That is a good question.  From the screenshot of Dave's teardown they are only using two wires to the fan so it is not feedback speed controlled but maybe they just change the voltage to the fan as the temperature rises.  Maybe someone that has taken their unit apart has some info.
« Last Edit: December 24, 2018, 01:57:25 pm by Roger Need »
Offline CDaniel
« Reply #16 on: December 24, 2018, 09:41:09 pm »
I looked some time ago ( maybe 1 year ) in my unit and has 2 wires . I didn't notice to be adjustable , it is just spinning all the time .
I think I mounted something rubbery underneath it , it's the metalic chassis that is vibrating from the fan . And this could be variable in time ...
« Last Edit: December 24, 2018, 09:46:40 pm by CDaniel »
Offline tek2232
« Reply #17 on: December 24, 2018, 10:42:01 pm »
Mine 1104Xe fan is loud and always blowing at full speed can be heard in every  corner of
my "lab"   calibrated date = Jun-20-2018 hardware version 01-03 , maybe your 1104XE
is from a later date. ?
Online Rerouter
« Reply #18 on: December 24, 2018, 10:48:34 pm »
Away from my lab until boxing day, but It was a pre-august model, so If there was a change around then, then you missed out by less than a month
Offline tek2232
« Reply #19 on: December 24, 2018, 11:15:34 pm »
Yes, that would be the case.
Offline Roger Need
« Reply #20 on: December 25, 2018, 04:52:48 am »
The fan may be speed controlled but I do not have a device to measure CFM at the fan output port.  Here is the graph you requested comparing the ambient sound level with the sound after 9 minutes of operation.   Recorded in Windows 8.1 with an external microphone and no auto gain.  You can see that the difference varies from 10 to 15 db at the low to mid frequencies.  To the human ear a 10 dB increase in sound level will be perceived as twice as loud which is what I observe with my unit.  The spike in levels from 2800 to 3000 Hz. may be the "whine" that I hear.
« Last Edit: December 25, 2018, 04:59:19 am by Roger Need »
Offline DaJMasta
    • medpants.com
« Reply #21 on: December 25, 2018, 05:20:02 am »
I sort of meant more of a sanity check - while zeroing out backround noise is useful, in a variable environment to corroborate a subjective (human) observation is a very easy way to convince yourself there's a big issue when the measurement doesn't actually show it.
In any case, you should be able to use the frequency content as your RPM indicator - if it goes up over time, then the fan is probably speeding up with increased load.  If you suspect vibration plays a part, you may be able to touch parts of the chassis near the fan or put a plastic piece through the grill (being careful not to obstruct the blades) to try and dampen any vibration of the fan itself.  While the design may not have emphasized silence, it also could be that there was a little wobble to the fan installed on the unit.
In any case, though, while it appears very fixable from the teardown photos by replacing the fan, if you are not comfortable doing so or don't think it's worth the effort, return is probably the best option if it bothers you.
Offline Roger Need
« Reply #22 on: December 26, 2018, 07:39:36 am »
There has been a lot of useful feedback on this thread.  Based on the comments above I made another series of measurements to make sure that the microphone was properly set-up (no AGC etc.) and to determine if a variable speed fan was being used.
In order to confirm that the fan was speeding up with temperature I ran the tests in the unheated section of my shop.  The temperature there is 13C (55F) compared to the lab which is 21C (70F).  The first thing I noticed is that the unit started up with hardly any fan noise at all.  Even after 2 minutes in the unheated shop it was still quiet compared to what I find in the lab.  I recorded measurements for about 20 minutes and then plotted the spectrum vs. time.  The results can be seen below. The first graph shows noise at ambient and after 2,4, 8 and 16 minutes of operation. The change from ambient to 16 minutes and 2 to 16 minutes are shown on separate plots.  One can observe that over time the noise is increasing due to the fan speeding up.  Spectral peaks from 2300 to 2800 Hz are evident and may be due to fan noise or the air passing through the grill at a higher CFM.
There are only 2 wires on this fan so they must be using voltage control, without feedback, to increase the fan speed and airflow. This is not a good design choice if different fan suppliers are used.  This may account for Tautech's observation that new units are less noisy. A different fan would not have the same CFM vs. voltage characteristics and might be less noisy for the same voltage.  The other thing that occurred to me is that the fan control algorithm may have have changed in the latest firmware release (1.3.23 - Sept. 20, 2018).  My firmware (5.1.3.17) is one release behind the current one.  Since I am returning this unit I am reluctant to update the firmware.  Perhaps someone else may want to comment or try an experiment with the updated firmware.
I hope this investigation is useful to others owning one of these scopes or for those considering a purchase.
Offline tinhead
« Reply #23 on: December 26, 2018, 08:28:55 am »
i have Oct 30th calibrated 4ch version, loud as shit and always running at full speed as well :\
Online Rerouter
« Reply #24 on: December 26, 2018, 08:49:10 am »
Then perhaps they got louder after the change date? Or perhaps just a differemt noise distribution from different fan blade noise
Have any cold rooms you could run it in for a few minutes. May be able to figure out how things change with different temperatures.
