| Products > Test Equipment |
| Siglent .ads firmware file format |
| << < (46/85) > >> |
| rf-loop:
--- Quote from: firstcolle on October 20, 2018, 11:27:19 am ---hi, I searched in the forum for a couple of hours but I didn't find anything.. where can I find a patched version of the latest firmware for a sds1104x-e with known telnet password? thanks --- End quote --- Read first all Instructions download links looks like broken. I do not understand at all why peoples use these total crap file sharing sides with tons of ads and popups etc junk and very limited time. better quality place for files: Here SDS1004X-E_OSV1_EN_eevblog.zip |
| cguareschi:
Very interesting thread and great work on decrypting ADS files. Read the entire thread several times and I would like to get my hands dirty with this as well. I tried janekivi tools v 0.1.4 on SDS1004X_E_6.1.26 but it didn't work. It says: This is not SDS firmware file. Do I need to reverse-xor the file before it can be parsed by janekivi tools? It would be great to be able to find out where the model is stored (EEPROM?), then change it to the 1004X model then use regular upgrades. Cheers |
| tv84:
--- Quote from: cguareschi on October 21, 2018, 12:03:42 pm ---I tried janekivi tools v 0.1.4 on SDS1004X_E_6.1.26 but it didn't work. It says: This is not SDS firmware file. Do I need to reverse-xor the file before it can be parsed by janekivi tools? It would be great to be able to find out where the model is stored (EEPROM?), then change it to the 1004X model then use regular upgrades. Cheers --- End quote --- I didn't test with janekivi's tool but it should work. It's a normal SDS X-E FW file. Parsing: --- Code: ---File Header Size: 00000070 00000000 - File Checksum: C60CADAC [00000004-00765FB8] (with only the File Header decrypted) CKSM OK 00000004 - File Size: 00765F49 (without 0x70 bytes of the File Header) 0000000C - Product_ID: 13501 00000026 - Vendor/Brand: SIGLENT 0000003A - USB Host Controller: ISP1763 **************************************************** Decrypting the 0x2800 and 0x1400 blocks... Reversing file... XORing with 0xFF (incrementing pattern)... XORing with 0xFF from 0x003B2FA5 until 0x00765F48 **************************************************** 00000000 --- Section Checksum: C302218F 00000004 --- Section Size: 00765F15 [00000034-00765F48] CKSM OK 00000008 --- Section # 00000007 00000034 --- 00765F48 ***** ZIP file ***** Offset Ver Flag Comp Size Packed Modified CRC32 Name Permissions Extra Details 00000034 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [00000056-00000055] lib/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 00000056 2.0 0000 0008 00005F13 00001B04 13-07-2018 10:23:21 9F6ED423 [00000084-00001B87] lib/brusvxi11.so MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 00001B88 2.0 0000 0008 0000916E 00003F99 13-07-2018 10:23:21 5B27B9AB [00001BBE-00005B56] lib/ext_device_driver.so MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 00005B57 2.0 0000 0008 00065C2C 00022A1A 13-07-2018 10:23:21 577A82D1 [00005B86-0002859F] lib/libEasyLib.so MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000285A0 2.0 0000 0008 0001E649 00008EBF 13-07-2018 10:23:21 AB77BB49 [000285D1-0003148F] lib/libvncclient.so MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00031490 2.0 0000 0008 0001E649 00008EBF 13-07-2018 10:23:21 AB77BB49 [000314C3-0003A381] lib/libvncclient.so.1 MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0003A382 2.0 0000 0008 00047CE2 000160A2 13-07-2018 10:23:21 3E6ADA03 [0003A3B3-00050454] lib/libvncserver.so MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00050455 2.0 0000 0008 00047CE2 000160A2 13-07-2018 10:23:21 3E6ADA03 [00050488-00066529] lib/libvncserver.so.1 MTime: 13-07-2018 03:23:43 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0006652A 1.0 0000 0000 00000000 00000000 09-08-2018 08:47:07 00000000 [00066550-0006654F] ui_data/ MTime: 09-08-2018 01:47:15 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 00066550 2.0 0000 0008 0001984A 00003053 16-07-2018 14:38:06 A321AADF [0006658A-000695DC] ui_data/arabic_help_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000695DD 2.0 0000 0008 00022BB7 00002B21 14-08-2018 15:23:13 3FBAF4FD [00069617-0006C137] ui_data/arabic_menu_info.xml MTime: 14-08-2018 08:23:26 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0006C138 2.0 0000 0008 00008529 00001991 16-07-2018 14:38:06 722FD019 [0006C172-0006DB02] ui_data/arabic_text_info.xml MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0006DB03 2.0 0000 0008 000003F8 00000058 16-07-2018 14:38:06 9DBCA89C [0006DB30-0006DB87] ui_data/dev.bmp MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0006DB88 2.0 0000 0008 00019F89 000030A7 27-07-2018 08:53:04 DB7E908B [0006DBC3-00070C69] ui_data/english_help_info.xml MTime: 27-07-2018 01:53:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00070C6A 2.0 0000 0008 00022F6D 00002B62 14-08-2018 14:51:22 6F53507E [00070CA5-00073806] ui_data/english_menu_info.xml MTime: 14-08-2018 07:51:45 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00073807 2.0 0000 0008 00008529 00001991 16-07-2018 14:38:06 722FD019 [00073842-000751D2] ui_data/english_text_info.xml MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000751D3 2.0 0000 0008 000198F7 00003088 27-07-2018 08:53:27 7AFB4C5C [0007520D-00078294] ui_data/french_help_info.xml MTime: 27-07-2018 01:53:55 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00078295 2.0 0000 0008 00023243 00002CAB 14-08-2018 14:52:01 F4FCDF5A [000782CF-0007AF79] ui_data/french_menu_info.xml MTime: 14-08-2018 07:52:03 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0007AF7A 2.0 0000 0008 0000850D 000019B5 16-07-2018 14:38:06 C784D9DF [0007AFB4-0007C968] ui_data/french_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0007C969 2.0 0000 0008 000198A1 00003080 27-07-2018 08:54:08 19D632E0 [0007C9A3-0007FA22] ui_data/german_help_info.xml MTime: 27-07-2018 01:54:17 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0007FA23 2.0 0000 0008 00023700 00002D88 14-08-2018 14:52:14 0E9BACA8 [0007FA5D-000827E4] ui_data/german_menu_info.xml MTime: 14-08-2018 07:52:29 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000827E5 2.0 0000 0008 00008C82 00001BDC 16-07-2018 14:38:06 096321F9 [0008281F-000843FA] ui_data/german_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000843FB 2.0 0000 0008 000198C4 00003080 27-07-2018 08:54:19 5B695C02 [00084436-000874B5] ui_data/italian_help_info.xml MTime: 27-07-2018 01:54:38 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000874B6 2.0 0000 0008 0002350C 00002CC1 14-08-2018 14:52:19 AED3E8E6 [000874F1-0008A1B1] ui_data/italian_menu_info.xml MTime: 14-08-2018 07:52:39 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0008A1B2 2.0 0000 0008 00008522 000019AA 16-07-2018 14:38:06 5488AB5E [0008A1ED-0008BB96] ui_data/italian_text_info.xml MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0008BB97 2.0 0000 0008 000198C4 00003081 27-07-2018 08:54:27 CEF5D080 [0008BBD3-0008EC53] ui_data/japanese_help_info.xml MTime: 27-07-2018 01:54:55 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0008EC54 2.0 0000 0008 00023F0F 00002F04 14-08-2018 14:52:23 585805B7 [0008EC90-00091B93] ui_data/japanese_menu_info.xml MTime: 14-08-2018 07:52:47 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00091B94 2.0 0000 0008 00008529 00001991 16-07-2018 14:38:06 722FD019 [00091BD0-00093560] ui_data/japanese_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00093561 2.0 0000 0008 000198C5 00003080 27-07-2018 08:55:08 355082EB [0009359B-0009661A] ui_data/korean_help_info.xml MTime: 27-07-2018 01:55:16 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0009661B 2.0 0000 0008 00023974 00002E32 14-08-2018 14:52:29 E6282D8E [00096655-00099486] ui_data/korean_menu_info.xml MTime: 14-08-2018 07:52:59 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00099487 2.0 0000 0008 00008529 00001991 16-07-2018 14:38:06 722FD019 [000994C1-0009AE51] ui_data/korean_text_info.xml MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0009AE52 2.0 0000 0008 000198C4 00003080 27-07-2018 08:55:17 AA3AC251 [0009AE90-0009DF0F] ui_data/portuguese_help_info.xml MTime: 27-07-2018 01:55:34 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0009DF10 2.0 0000 0008 00023191 00002C4B 14-08-2018 14:53:04 874BE64C [0009DF4E-000A0B98] ui_data/portuguese_menu_info.xml MTime: 14-08-2018 07:53:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000A0B99 2.0 0000 0008 0000851C 000019B1 16-07-2018 14:38:06 ED13903B [000A0BD7-000A2587] ui_data/portuguese_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000A2588 2.0 0000 0008 000198C4 00003080 27-07-2018 08:58:03 5B695C02 [000A25C3-000A5642] ui_data/russian_help_info.xml MTime: 27-07-2018 01:58:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000A5643 2.0 0000 0008 00024A30 00003006 14-08-2018 14:56:05 10CA9361 [000A567E-000A8683] ui_data/russian_menu_info.xml MTime: 14-08-2018 07:56:11 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000A8684 2.0 0000 0008 00008584 00001A4E 16-07-2018 14:38:06 40A19302 [000A86BF-000AA10C] ui_data/russian_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000AA10D 2.0 0000 0008 00018C1B 000032D0 27-07-2018 09:05:23 43A609DF [000AA145-000AD414] ui_data/simp_help_info.xml MTime: 27-07-2018 02:05:46 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000AD415 2.0 0000 0008 00023864 00002E41 14-08-2018 14:56:11 81A5D501 [000AD44D-000B028D] ui_data/simp_menu_info.xml MTime: 14-08-2018 07:56:22 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000B028E 2.0 0000 0008 00008473 00001AB2 16-07-2018 14:38:06 E027E215 [000B02C6-000B1D77] ui_data/simp_text_info.xml MTime: 16-07-2018 07:38:12 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000B1D78 2.0 0000 0008 000198C8 00003082 27-07-2018 09:01:23 2072E7B2 [000B1DB3-000B4E34] ui_data/spanish_help_info.xml MTime: 27-07-2018 02:01:47 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000B4E35 2.0 0000 0008 00023426 00002C4B 14-08-2018 14:56:15 A89A8946 [000B4E70-000B7ABA] ui_data/spanish_menu_info.xml MTime: 14-08-2018 07:56:30 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000B7ABB 2.0 0000 0008 00008540 000019B7 16-07-2018 14:38:05 3D98F0D4 [000B7AF6-000B94AC] ui_data/spanish_text_info.xml MTime: 16-07-2018 07:38:11 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000B94AD 2.0 0000 0008 0001849E 0000328A 27-07-2018 09:13:11 AFFC2F20 [000B94E5-000BC76E] ui_data/trad_help_info.xml MTime: 27-07-2018 02:13:23 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000BC76F 2.0 0000 0008 00023858 00002EE8 14-08-2018 14:56:19 3E117598 [000BC7A7-000BF68E] ui_data/trad_menu_info.xml MTime: 14-08-2018 07:56:38 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000BF68F 2.0 0000 0008 0000846C 00001AF7 16-07-2018 14:38:06 5395CA64 [000BF6C7-000C11BD] ui_data/trad_text_info.xml MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C11BE 2.0 0000 0008 00000308 00000066 16-07-2018 14:38:06 539A7E1A [000C11EB-000C1250] ui_data/usb.bmp MTime: 16-07-2018 07:38:13 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C1251 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [000C1273-000C1272] www/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:12 000C1273 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [000C129F-000C129E] www/attchment/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 000C129F 2.0 0000 0008 00005AD3 00005AC5 16-07-2018 16:54:16 E23B57EF [000C12D2-000C6D96] www/BinToCSV_tool.zip MTime: 16-07-2018 09:54:33 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:12 000C6D97 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [000C6DBD-000C6DBC] www/css/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 000C6DBD 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [000C6DEA-000C6DE9] www/css/device/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 000C6DEA 2.0 0000 0008 00001A0A 0000065D 13-07-2018 10:21:01 DA06DD28 [000C6E29-000C7485] www/css/device/device_control.css MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C7486 2.0 0000 0008 00000E6F 00000492 13-07-2018 10:21:01 731108C6 [000C74BB-000C794C] www/css/device/main.css MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C794D 2.0 0000 0008 000007DD 0000029B 13-07-2018 10:21:01 C8FCE2D7 [000C797C-000C7C16] www/css/login.css MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C7C17 2.0 0000 0008 000011D5 00000577 13-07-2018 10:21:01 26D75603 [000C7C48-000C81BE] www/css/welcome.css MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000C81BF 2.0 0000 0008 000007C4 00000239 13-07-2018 10:21:02 D3E45378 [000C81F1-000C8429] www/deviceupdate.php MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:12 000C842A 2.0 0000 0008 000005F2 0000017D 13-07-2018 10:21:03 7BB9B512 [000C8461-000C85DD] www/device_read_write.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:12 000C85DE 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [000C8606-000C8605] www/fonts/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 000C8606 2.0 0000 0008 00011497 0001146B 13-07-2018 10:21:02 A3758D75 [000C8645-000D9AAF] www/fonts/fontawesome-webfont.eot MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000D9AB0 2.0 0000 0008 00059430 0001A95E 13-07-2018 10:21:02 CFF95578 [000D9AEF-000F444C] www/fonts/fontawesome-webfont.svg MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 000F444D 2.0 0000 0008 00022AF8 0001453A 13-07-2018 10:21:02 AC39751F [000F448C-001089C5] www/fonts/fontawesome-webfont.ttf MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001089C6 2.0 0000 0008 00014684 0001464C 13-07-2018 10:21:02 C17DCBF4 [00108A06-0011D051] www/fonts/fontawesome-webfont.woff MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0011D052 2.0 0000 0008 00010440 00010431 13-07-2018 10:21:02 B5F6695A [0011D093-0012D4C3] www/fonts/fontawesome-webfont.woff2 MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0012D4C4 2.0 0000 0008 0001AC78 00015EFD 13-07-2018 10:21:02 98F5FCC9 [0012D4FB-001433F7] www/fonts/FontAwesome.otf MTime: 13-07-2018 03:21:04 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001433F8 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [0014341E-0014341D] www/img/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 0014341E 2.0 0000 0008 00004723 0000112A 13-07-2018 10:21:04 0A7D1CDE [0014344D-00144576] www/img/about.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00144577 2.0 0000 0008 00006682 00002200 13-07-2018 10:21:04 A3952F7A [001445A6-001467A5] www/img/Home1.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001467A6 2.0 0000 0008 000056E1 00001973 13-07-2018 10:21:04 3F022755 [001467D5-00148147] www/img/Home2.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00148148 2.0 0000 0008 000001A9 000001AE 13-07-2018 10:21:04 2B0955C8 [0014817A-00148327] www/img/ic_login.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00148328 2.0 0000 0008 0000015B 00000160 13-07-2018 10:21:04 26CF6E8F [0014835D-001484BC] www/img/ic_password.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001484BD 2.0 0000 0008 0000011E 00000121 13-07-2018 10:21:04 2D89CBF8 [001484EE-0014860E] www/img/ic_user.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0014860F 2.0 0000 0008 00006E08 00002A4F 13-07-2018 10:21:04 6A01E632 [00148643-0014B091] www/img/Intrument1.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0014B092 2.0 0000 0008 00006444 00002088 13-07-2018 10:21:04 54331AA2 [0014B0C6-0014D14D] www/img/Intrument2.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0014D14E 2.0 0000 0008 0000704D 00002BC9 13-07-2018 10:21:04 7A6E3F62 [0014D17C-0014FD44] www/img/LAN1.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0014FD45 2.0 0000 0008 00006179 00002345 13-07-2018 10:21:04 F227C9F5 [0014FD73-001520B7] www/img/LAN2.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001520B8 2.0 0000 0008 000156C4 00012101 13-07-2018 10:21:04 4766C70E [001520E6-001641E6] www/img/logo.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001641E7 2.0 0000 0008 0000740F 00002F8F 13-07-2018 10:21:04 A4E5054B [00164216-001671A4] www/img/SCPI1.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001671A5 2.0 0000 0008 00005DD6 000019E5 13-07-2018 10:21:04 10F8D9C4 [001671D4-00168BB8] www/img/SCPI2.png MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00168BB9 2.0 0000 0008 00000261 00000101 13-07-2018 10:21:03 59765DD4 [00168BE4-00168CE4] www/index.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:12 00168CE5 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [00168D12-00168D11] www/Instrument/ MTime: 09-08-2018 01:44:13 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:13 00168D12 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [00168D45-00168D44] www/Instrument/novnc/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:13 00168D45 2.0 0000 0008 00000012 00000012 13-07-2018 10:21:04 09EA74AA [00168D83-00168D94] www/Instrument/novnc/favicon.ico MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00168D95 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [00168DCF-00168DCE] www/Instrument/novnc/images/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:14 00168DCF 2.0 0000 0008 00000153 00000151 13-07-2018 10:21:04 E8B725E1 [00168E10-00168F60] www/Instrument/novnc/images/alt.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00168F61 2.0 0000 0008 000001F5 000001FA 13-07-2018 10:21:04 D063D8CA [00168FA8-001691A1] www/Instrument/novnc/images/clipboard.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001691A2 2.0 0000 0008 00000194 00000199 13-07-2018 10:21:04 310798F7 [001691E7-0016937F] www/Instrument/novnc/images/connect.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00169380 2.0 0000 0008 00000162 00000161 13-07-2018 10:21:04 6610B793 [001693C2-00169522] www/Instrument/novnc/images/ctrl.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00169523 2.0 0000 0008 0000013D 00000142 13-07-2018 10:21:04 5FA26818 [0016956B-001696AC] www/Instrument/novnc/images/ctrlaltdel.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001696AD 2.0 0000 0008 00000562 00000567 13-07-2018 10:21:04 2EABA092 [001696F5-00169C5B] www/Instrument/novnc/images/disconnect.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00169C5C 2.0 0000 0008 000003C3 000003C8 13-07-2018 10:21:04 06D319EE [00169C9E-0016A065] www/Instrument/novnc/images/drag.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016A066 2.0 0000 0008 00000181 00000182 13-07-2018 10:21:04 F86CB8F8 [0016A0A7-0016A228] www/Instrument/novnc/images/esc.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016A229 2.0 0000 0008 0000019E 00000189 13-07-2018 10:21:04 FD031A25 [0016A26D-0016A3F5] www/Instrument/novnc/images/expand.gif MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016A3F6 2.0 0000 0008 0000047E 000000FF 13-07-2018 10:21:04 B17B6C52 [0016A43B-0016A539] www/Instrument/novnc/images/favicon.ico MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016A53A 2.0 0000 0008 000001C5 000001C5 13-07-2018 10:21:04 86BA4433 [0016A57F-0016A743] www/Instrument/novnc/images/favicon.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016A744 2.0 0000 0008 000002ED 000002F2 13-07-2018 10:21:04 3956D9F0 [0016A78D-0016AA7E] www/Instrument/novnc/images/full_screen.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016AA7F 2.0 0000 0008 000001A7 00000199 13-07-2018 10:21:04 FFF8D302 [0016AAC1-0016AC59] www/Instrument/novnc/images/hide.gif MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016AC5A 2.0 0000 0008 00000503 00000508 13-07-2018 10:21:04 0E44D2DA [0016ACA0-0016B1A7] www/Instrument/novnc/images/keyboard.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016B1A8 2.0 0000 0008 0000014A 00000141 13-07-2018 10:21:04 297C782F [0016B1EA-0016B32A] www/Instrument/novnc/images/left.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016B32B 2.0 0000 0008 000001FF 00000204 13-07-2018 10:21:04 33C0F7B9 [0016B373-0016B576] www/Instrument/novnc/images/mouse_left.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016B577 2.0 0000 0008 00000205 0000020A 13-07-2018 10:21:04 DE48BB49 [0016B5C1-0016B7CA] www/Instrument/novnc/images/mouse_middle.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016B7CB 2.0 0000 0008 000001F1 000001F6 13-07-2018 10:21:04 B8155471 [0016B813-0016BA08] www/Instrument/novnc/images/mouse_none.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016BA09 2.0 0000 0008 00000201 00000206 13-07-2018 10:21:04 4BDAF377 [0016BA52-0016BC57] www/Instrument/novnc/images/mouse_right.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016BC58 2.0 0000 0008 000000DF 000000D1 13-07-2018 10:21:04 5BC997F5 [0016BCA1-0016BD71] www/Instrument/novnc/images/new-down-24.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016BD72 2.0 0000 0008 000000EE 000000E0 13-07-2018 10:21:04 059AC92F [0016BDBB-0016BE9A] www/Instrument/novnc/images/new-left-24.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016BE9B 2.0 0000 0008 000000EA 000000DD 13-07-2018 10:21:04 D3CBF039 [0016BEE5-0016BFC1] www/Instrument/novnc/images/new-right-24.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016BFC2 2.0 0000 0008 000000B7 000000AB 13-07-2018 10:21:04 3407F6E1 [0016C009-0016C0B3] www/Instrument/novnc/images/new-up-24.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C0B4 2.0 0000 0008 00000186 00000188 13-07-2018 10:21:04 4EA5EAAC [0016C0F7-0016C27E] www/Instrument/novnc/images/power.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C27F 2.0 0000 0008 00000257 0000025C 13-07-2018 10:21:04 4DA3EF2C [0016C2C6-0016C521] www/Instrument/novnc/images/press_new.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C522 2.0 0000 0008 00000122 0000011C 13-07-2018 10:21:04 11609785 [0016C565-0016C680] www/Instrument/novnc/images/right.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C681 2.0 0000 0008 00000141 0000013E 13-07-2018 10:21:04 0AA95190 [0016C6C8-0016C805] www/Instrument/novnc/images/r_left-16.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C806 2.0 0000 0008 00000142 0000013D 13-07-2018 10:21:04 D26DA198 [0016C84E-0016C98A] www/Instrument/novnc/images/r_right-16.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016C98B 2.0 0000 0008 000031EA 00002ED0 13-07-2018 10:21:04 BB29E23A [0016C9D7-0016F8A6] www/Instrument/novnc/images/screen_320x460.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 0016F8A7 2.0 0000 0008 0000070F 00000714 13-07-2018 10:21:04 6D5EA4CD [0016F8F1-00170004] www/Instrument/novnc/images/screen_57x57.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00170005 2.0 0000 0008 0000460A 00003DE3 13-07-2018 10:21:04 82300E91 [00170051-00173E33] www/Instrument/novnc/images/screen_700x700.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00173E34 2.0 0000 0008 000009BF 00000977 13-07-2018 10:21:04 4A582C85 [00173E7A-001747F0] www/Instrument/novnc/images/settings.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 001747F1 2.0 0000 0008 000002DF 000002E4 13-07-2018 10:21:04 AD86DFB3 [0017483C-00174B1F] www/Instrument/novnc/images/showextrakeys.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00174B20 2.0 0000 0008 000000AD 000000A0 13-07-2018 10:21:04 76A67EB7 [00174B62-00174C01] www/Instrument/novnc/images/stop.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00174C02 2.0 0000 0008 00000183 00000186 13-07-2018 10:21:04 5026226C [00174C43-00174DC8] www/Instrument/novnc/images/tab.png MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00174DC9 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [00174E04-00174E03] www/Instrument/novnc/include/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:13 00174E04 2.0 0000 0008 00002A17 000009EF 14-07-2018 15:16:18 BDD7A20C [00174E47-00175835] www/Instrument/novnc/include/base.css MTime: 14-07-2018 08:16:36 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00175836 2.0 0000 0008 0000153E 0000088A 13-07-2018 10:21:03 7A8AC16F [0017587A-00176103] www/Instrument/novnc/include/base64.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00176104 2.0 0000 0008 00002AAF 00000A00 14-07-2018 15:16:18 0177A851 [0017614E-00176B4D] www/Instrument/novnc/include/base_mobile.css MTime: 14-07-2018 08:16:36 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00176B4E 2.0 0000 0008 00001205 000002F4 13-07-2018 10:21:03 32FDD938 [00176B92-00176E85] www/Instrument/novnc/include/black.css MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00176E86 2.0 0000 0008 00000534 000001A5 13-07-2018 10:21:03 F31B00F0 [00176EC9-0017706D] www/Instrument/novnc/include/blue.css MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017706E 2.0 0000 0008 000007EE 000002B4 13-07-2018 10:21:03 58434E20 [001770B4-00177367] www/Instrument/novnc/include/browsers.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00177368 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [001773AE-001773AD] www/Instrument/novnc/include/chrome-app/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 001773AE 2.0 0000 0008 000024F9 00000A27 13-07-2018 10:21:03 17252ED4 [00177401-00177E27] www/Instrument/novnc/include/chrome-app/tcp-client.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:14 00177E28 2.0 0000 0008 000029BA 00000F2C 13-07-2018 10:21:03 C52426D5 [00177E69-00178D94] www/Instrument/novnc/include/des.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00178D95 2.0 0000 0008 00005575 0000175C 13-07-2018 10:21:03 CD04C438 [00178DDA-0017A535] www/Instrument/novnc/include/display.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017A536 2.0 0000 0008 00002C4A 00000A8C 13-07-2018 10:21:03 5D13357B [0017A57F-0017B00A] www/Instrument/novnc/include/domkeytable.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017B00B 2.0 0000 0008 000004E0 000001EF 13-07-2018 10:21:03 66DA53C8 [0017B052-0017B240] www/Instrument/novnc/include/encodings.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017B241 2.0 0000 0008 000010A2 000005A8 13-07-2018 10:21:03 7934485D [0017B285-0017B82C] www/Instrument/novnc/include/events.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017B82D 2.0 0000 0008 00000409 00000196 13-07-2018 10:21:03 C7D62201 [0017B876-0017BA0B] www/Instrument/novnc/include/eventtarget.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017BA0C 2.0 0000 0008 00000EC4 00000467 13-07-2018 10:21:03 6AAB8870 [0017BA53-0017BEB9] www/Instrument/novnc/include/fixedkeys.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017BEBA 2.0 0000 0008 0000049A 000001CE 13-07-2018 10:21:03 6C773BC2 [0017BF00-0017C0CD] www/Instrument/novnc/include/inflator.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017C0CE 2.0 0000 0008 0000C6A5 00002DAD 13-07-2018 10:21:03 CB3AA190 [0017C111-0017EEBD] www/Instrument/novnc/include/input.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0017EEBE 2.0 0000 0008 00004AC1 00001319 13-07-2018 10:21:03 564AF4D0 [0017EF03-0018021B] www/Instrument/novnc/include/jsunzip.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0018021C 2.0 0000 0008 00005168 000015AA 13-07-2018 10:21:03 C9CCAD61 [00180262-0018180B] www/Instrument/novnc/include/keyboard.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 0018180C 2.0 0000 0008 000057FE 000011D5 13-07-2018 10:21:03 BFC59E51 [00181850-00182A24] www/Instrument/novnc/include/keysym.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00182A25 2.0 0000 0008 00005422 00001F89 13-07-2018 10:21:03 BA5BA9A2 [00182A6C-001849F4] www/Instrument/novnc/include/keysymdef.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 001849F5 2.0 0000 0008 00000536 00000209 13-07-2018 10:21:03 156BE805 [00184A3A-00184C42] www/Instrument/novnc/include/logging.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:44 CTime: 09-08-2018 01:44:13 00184C43 2.0 0000 0008 00006BDA 00004E97 13-07-2018 10:21:03 83395258 [00184C85-00189B1B] www/Instrument/novnc/include/logo.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00189B1C 2.0 0000 0008 00002740 00000A7D 13-07-2018 10:21:03 4F35F4AF [00189B5F-0018A5DB] www/Instrument/novnc/include/mouse.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 0018A5DC 2.0 0000 0008 000096B4 00004435 13-07-2018 10:21:03 75810239 [0018A626-0018EA5A] www/Instrument/novnc/include/Orbitron700.ttf MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 0018EA5B 2.0 0000 0008 00004440 0000441B 13-07-2018 10:21:03 9B7440BD [0018EAA6-00192EC0] www/Instrument/novnc/include/Orbitron700.woff MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00192EC1 2.0 0000 0008 000009BF 00000396 13-07-2018 10:21:03 EB0D8CA4 [00192F07-0019329C] www/Instrument/novnc/include/playback.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 0019329D 2.0 0000 0008 0000074F 000002F4 13-07-2018 10:21:03 CA2202E0 [001932E3-001935D6] www/Instrument/novnc/include/polyfill.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001935D7 2.0 0000 0008 0000F593 00003A03 13-07-2018 10:21:03 965E43CA [00193618-0019701A] www/Instrument/novnc/include/rfb.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 0019701B 2.0 0000 0008 0000013B 000000E7 13-07-2018 10:21:03 312DF302 [00197060-00197146] www/Instrument/novnc/include/strings.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00197147 2.0 0000 0008 0000901A 00001C34 13-07-2018 10:21:03 8C2E1648 [00197187-00198DBA] www/Instrument/novnc/include/ui.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00198DBB 2.0 0000 0008 00002425 00000B93 13-07-2018 10:21:03 D823E18B [00198DFD-0019998F] www/Instrument/novnc/include/util.js MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00199990 2.0 0000 0008 000009D7 000003B6 13-07-2018 10:21:03 BFC66E5F [001999D3-00199D88] www/Instrument/novnc/include/vkeys.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00199D89 2.0 0000 0008 0000043F 000001F6 13-07-2018 10:21:03 66464436 [00199DCA-00199FBF] www/Instrument/novnc/include/vnc.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 00199FC0 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [0019A009-0019A008] www/Instrument/novnc/include/web-socket-js/ MTime: 09-08-2018 01:44:14 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:14 0019A009 2.0 0000 0008 00001753 00001758 13-07-2018 10:21:03 F2D8E66E [0019A05C-0019B7B3] www/Instrument/novnc/include/web-socket-js/README.txt MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:14 0019B7B4 2.0 0000 0008 000027EC 00000F3C 13-07-2018 10:21:03 C1200AA1 [0019B809-0019C744] www/Instrument/novnc/include/web-socket-js/swfobject.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:14 0019C745 2.0 0000 0008 0002B3F3 0002981D 13-07-2018 10:21:03 044DC2E9 [0019C79F-001C5FBB] www/Instrument/novnc/include/web-socket-js/WebSocketMain.swf MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:14 001C5FBC 2.0 0000 0008 000032F0 0000104B 13-07-2018 10:21:03 8FA21DD0 [001C6012-001C705C] www/Instrument/novnc/include/web-socket-js/web_socket.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:14 001C705D 2.0 0000 0008 000023E2 00000B42 13-07-2018 10:21:03 FF6488B2 [001C70A2-001C7BE3] www/Instrument/novnc/include/websock.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001C7BE4 2.0 0000 0008 00001041 00000652 13-07-2018 10:21:03 F887C1C8 [001C7C29-001C827A] www/Instrument/novnc/include/webutil.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001C827B 2.0 0000 0008 000037B0 00000F15 13-07-2018 10:21:03 2076F13A [001C82C4-001C91D8] www/Instrument/novnc/include/xtscancodes.js MTime: 13-07-2018 03:21:07 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:14 001C91D9 2.0 0000 0008 00000AE2 00000AA5 13-07-2018 10:21:03 5DF86B18 [001C9217-001C9CBB] www/Instrument/novnc/LICENSE.txt MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001C9CBC 2.0 0000 0008 00001512 000009A4 13-07-2018 10:21:04 ECFC1F4A [001C9CF8-001CA69B] www/Instrument/novnc/README.md MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001CA69C 2.0 0000 0008 0000C466 0000253F 13-07-2018 10:21:04 1D39BAF3 [001CA6DB-001CCC19] www/Instrument/novnc/vnc_auto.php MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001CCC1A 2.0 0000 0008 0000D2F2 000027A4 13-07-2018 10:21:04 4418B719 [001CCC60-001CF403] www/Instrument/novnc/vnc_auto_mobile.php MTime: 13-07-2018 03:21:08 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001CF404 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001CF429-001CF428] www/js/ MTime: 09-08-2018 01:44:13 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:12 001CF429 2.0 0000 0008 000017EF 00000836 13-07-2018 10:21:04 29BD1571 [001CF456-001CFC8B] www/js/cycle.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001CFC8C 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001CFCB8-001CFCB7] www/js/device/ MTime: 09-08-2018 01:44:13 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:13 001CFCB8 2.0 0000 0008 00012AEE 000031D3 13-07-2018 10:21:04 C39FF42A [001CFCEE-001D2EC0] www/js/device/t_table.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001D2EC1 2.0 0000 0008 0000069E 000002D8 13-07-2018 10:21:04 D03B120A [001D2EF2-001D31C9] www/js/home_info.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001D31CA 2.0 0000 0008 00010C6C 00002D85 13-07-2018 10:21:04 5E8976A5 [001D31F9-001D5F7D] www/js/iscroll.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001D5F7E 2.0 0000 0008 00000C31 00000539 13-07-2018 10:21:04 DB5D4D50 [001D5FB3-001D64EB] www/js/jquery.cookie.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001D64EC 2.0 0000 0008 0000FD23 00003377 13-07-2018 10:21:04 F4479356 [001D6525-001D989B] www/js/jquery.fileupload.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001D989C 2.0 0000 0008 00002C0C 00000B06 13-07-2018 10:21:04 A612B1BD [001D98DB-001DA3E0] www/js/jquery.iframe-transport.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001DA3E1 2.0 0000 0008 000176F8 00008125 13-07-2018 10:21:04 6AE93E35 [001DA413-001E2537] www/js/jquery.min.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E2538 2.0 0000 0008 000013E4 000006A3 13-07-2018 10:21:04 40D7DC92 [001E2571-001E2C13] www/js/jquery.pagination.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E2C14 2.0 0000 0008 00001E96 00000AA1 13-07-2018 10:21:04 8C7EA0B6 [001E2C4E-001E36EE] www/js/jquery.transit.min.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E36EF 2.0 0000 0008 00000FC3 00000544 13-07-2018 10:21:04 5737EAD4 [001E3728-001E3C6B] www/js/jquery.wordexport.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E3C6C 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001E3CA0-001E3C9F] www/js/JSON-js-master/ MTime: 09-08-2018 01:44:13 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:12 001E3CA0 2.0 0000 0008 000017EF 00000836 13-07-2018 10:21:04 29BD1571 [001E3CDC-001E4511] www/js/JSON-js-master/cycle.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E4512 2.0 0000 0008 00004800 0000157C 13-07-2018 10:21:04 685DAC78 [001E454E-001E5AC9] www/js/JSON-js-master/json2.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001E5ACA 2.0 0000 0008 000023FF 00000A6B 13-07-2018 10:21:04 FE620627 [001E5B0B-001E6575] www/js/JSON-js-master/json_parse.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001E6576 2.0 0000 0008 00003341 00000CFC 13-07-2018 10:21:04 C908B106 [001E65BD-001E72B8] www/js/JSON-js-master/json_parse_state.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001E72B9 2.0 0000 0008 00000689 0000035D 13-07-2018 10:21:04 B4BE88FD [001E72F3-001E764F] www/js/JSON-js-master/README MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:13 001E7650 2.0 0000 0008 00004800 0000157C 13-07-2018 10:21:04 685DAC78 [001E767D-001E8BF8] www/js/json2.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E8BF9 2.0 0000 0008 000023FF 00000A6B 13-07-2018 10:21:04 FE620627 [001E8C2B-001E9695] www/js/json_parse.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001E9696 2.0 0000 0008 00003341 00000CFC 13-07-2018 10:21:04 C908B106 [001E96CE-001EA3C9] www/js/json_parse_state.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EA3CA 2.0 0000 0008 000046A4 00000C68 13-07-2018 10:21:04 106FAAB6 [001EA3FC-001EB063] www/js/lan_config.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EB064 2.0 0000 0008 00000085 00000062 13-07-2018 10:21:04 872B6CB6 [001EB092-001EB0F3] www/js/log_in.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EB0F4 2.0 0000 0008 0000215A 00000BE2 13-07-2018 10:21:04 A11D7515 [001EB120-001EBD01] www/js/main.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EBD02 2.0 0000 0008 00001DD7 00000A40 13-07-2018 10:21:04 02D7D4BB [001EBD30-001EC76F] www/js/mcplib.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EC770 2.0 0000 0008 000001E4 000000B9 13-07-2018 10:21:04 CC50E34D [001EC79B-001EC853] www/js/npm.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001EC854 2.0 0000 0008 00019503 000084B5 13-07-2018 10:21:04 FC554B04 [001EC888-001F4D3C] www/js/TweenMax.min.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F4D3D 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001F4D69-001F4D68] www/js/vendor/ MTime: 09-08-2018 01:44:12 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:12 001F4D69 2.0 0000 0008 00003ED9 00001325 13-07-2018 10:21:04 B8246F6C [001F4DA8-001F60CC] www/js/vendor/jquery.ui.widget.js MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F60CD 2.0 0000 0008 00000B2D 0000045B 13-07-2018 10:21:04 F719B09D [001F60F7-001F6551] www/js/xhtml MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F6552 2.0 0000 0008 00007D6F 00001F54 13-07-2018 10:21:04 0AE8B9F2 [001F658E-001F84E1] www/js/xhtml1-transitional.dtd MTime: 13-07-2018 03:21:09 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F84E2 2.0 0000 0008 00001E9B 00000686 13-07-2018 10:21:03 7F4DEC6B [001F850B-001F8B90] www/lan.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F8B91 2.0 0000 0008 00002BDE 000008FB 13-07-2018 10:21:03 CB8B927E [001F8BC2-001F94BC] www/lan_setting.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F94BD 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001F94E3-001F94E2] www/log/ MTime: 09-08-2018 01:44:12 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:12 001F94E3 1.0 0000 0000 00000000 00000000 13-07-2018 10:21:01 00000000 [001F951B-001F951A] www/log/lighttpd_error.log MTime: 13-07-2018 03:21:02 ATime: 09-08-2018 01:44:12 CTime: 09-08-2018 01:44:12 001F951B 2.0 0000 0008 00000FE9 000005CB 13-07-2018 10:21:03 DC1C907D [001F9546-001F9B10] www/login.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F9B11 2.0 0000 0008 0000011B 000000B7 13-07-2018 10:21:03 562D1E79 [001F9B3D-001F9BF3] www/logout.php MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001F9BF4 2.0 0000 0008 000010AE 00000546 13-07-2018 10:21:01 F7CB535A [001F9C26-001FA16B] www/SCPI_control.php MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001FA16C 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:06 00000000 [001FA192-001FA191] www/src/ MTime: 09-08-2018 01:44:12 ATime: 15-08-2018 08:26:43 CTime: 09-08-2018 01:44:12 001FA192 2.0 0000 0008 000000B9 0000007F 13-07-2018 10:21:01 E60190A8 [001FA1BF-001FA23D] www/src/IDN.php MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001FA23E 2.0 0000 0008 000000C8 00000082 13-07-2018 10:21:01 9E8519B7 [001FA26E-001FA2EF] www/src/index.html MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001FA2F0 2.0 0000 0008 000000CE 00000085 13-07-2018 10:21:03 4734FAA2 [001FA31F-001FA3A3] www/vncserver.txt MTime: 13-07-2018 03:21:06 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001FA3A4 2.0 0000 0008 00001366 00000502 13-07-2018 10:21:01 0C3FC166 [001FA3D1-001FA8D2] www/welcome.php MTime: 13-07-2018 03:21:02 ATime: 15-08-2018 08:26:45 CTime: 09-08-2018 01:44:12 001FA8D3 2.0 0000 0008 0000CDB2 00000FF2 14-08-2018 14:56:25 EAE3BFFB [001FA904-001FB8F5] factory_setting.xml MTime: 14-08-2018 07:56:50 ATime: 15-08-2018 08:26:07 CTime: 09-08-2018 01:44:11 001FB8F6 2.0 0000 0008 003DBB6D 00158549 26-07-2018 11:46:04 F3302A7F [001FB92E-00353E76] top_sds1000b_fpga_256M.bit MTime: 26-07-2018 04:46:09 ATime: 15-08-2018 08:26:07 CTime: 09-08-2018 01:44:12 00353E77 2.0 0000 0008 00000E34 00000335 30-05-2018 11:47:04 19E22C47 [00353E9E-003541D2] update.sh MTime: 30-05-2018 04:47:08 ATime: 15-08-2018 08:26:07 CTime: 09-08-2018 01:44:12 003541D3 2.0 0000 0008 00006831 000024BC 13-07-2018 10:23:22 66D144AF [003541FA-003566B5] vncserver MTime: 13-07-2018 03:23:44 ATime: 15-08-2018 08:26:07 CTime: 09-08-2018 01:44:12 003566B6 2.0 0000 0008 00CA9B1C 003F194E 15-08-2018 15:51:14 26EE22F2 [003566E0-0074802D] sds1000b.app MTime: 15-08-2018 08:51:28 ATime: 15-08-2018 08:44:33 CTime: 09-08-2018 01:44:11 0074802E 1.0 0000 0000 00000000 00000000 09-08-2018 08:44:07 00000000 [00748054-00748053] drivers/ MTime: 09-08-2018 01:44:15 ATime: 15-08-2018 08:26:39 CTime: 09-08-2018 01:44:15 00748054 2.0 0000 0008 000068AD 000027D3 08-07-2017 00:10:13 FAAA09A7 [00748085-0074A857] drivers/g_usbtmc.ko MTime: 07-07-2017 17:10:26 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 0074A858 2.0 0000 0008 0000ECB5 00005D2C 08-07-2017 00:10:13 6E3E714E [0074A88D-007505B8] drivers/libcomposite.ko MTime: 07-07-2017 17:10:26 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 007505B9 2.0 0000 0008 000118E6 0000778E 17-11-2017 10:16:20 4C1729CD [007505E9-00757D76] drivers/mt7601u.ko MTime: 17-11-2017 02:16:40 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 00757D77 2.0 0000 0008 000028C5 00000DF7 08-07-2017 00:10:12 EADEFABE [00757DA9-00758B9F] drivers/siglentkb.ko MTime: 07-07-2017 17:10:25 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 00758BA0 2.0 0000 0008 00002E6E 000010D2 08-07-2017 00:10:12 2DAC2FC4 [00758BD4-00759CA5] drivers/siglent_dma.ko MTime: 07-07-2017 17:10:24 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 00759CA6 2.0 0000 0008 00002396 00000D35 08-07-2017 00:10:12 888DAB3C [00759CDB-0075AA0F] drivers/siglent_vdma.ko MTime: 07-07-2017 17:10:25 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 0075AA10 2.0 0000 0008 00002288 00000B80 20-03-2018 11:04:02 E0989036 [0075AA44-0075B5C3] drivers/siglent_vnc.ko MTime: 20-03-2018 03:04:05 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 0075B5C4 2.0 0000 0008 00003729 00001639 08-07-2017 00:10:12 AF8F4F6D [0075B5FB-0075CC33] drivers/xilinx_axicdma.ko MTime: 07-07-2017 17:10:25 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 0075CC34 2.0 0000 0008 00003B9B 00001756 08-07-2017 00:10:12 A70C189A [0075CC6A-0075E3BF] drivers/xilinx_axidma.ko MTime: 07-07-2017 17:10:25 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 0075E3C0 2.0 0000 0008 0000412D 00001999 08-07-2017 00:10:12 6F40AEAD [0075E3F4-0075FD8C] drivers/xilinx_vdma.ko MTime: 07-07-2017 17:10:25 ATime: 15-08-2018 08:26:46 CTime: 09-08-2018 01:44:15 Disk Entries: 226 Total Entries: 226 Directory Size: 24998 bytes [0075FD8D-00765F32] **************************************************** File Processed OK --- End code --- |
| cguareschi:
Oh well, it worked on SDS200x ads firmware file straight out, but not on 1xx4X-E ads, tried .25R2 and .26. @tv84, thanks for your reply, by any chance did you post your parsing script anywhere? couldn't find it. If you didn't, would you mind sharing? Thanks |
| tv84:
--- Quote from: cguareschi on October 21, 2018, 12:27:48 pm ---@tv84, thanks for your reply, by any chance did you post your parsing script anywhere? couldn't find it. If you didn't, would you mind sharing? --- End quote --- My "script" is part of a full-blown work in progress. Sorry, no share. Janekivi's tool and all other spinoffs/infos that have been made available (in this thread) should be enough to accomplish any need for the most persistents. |
| Navigation |
| Message Index |
| Next page |
| Previous page |