kladit:
Rob, thanks for you quick answer.

I did a SOLT calibration and tryed to put in the parameters for Load
in USER1, but it is still grayed out so no input possible.

I think it should go the other way round, but may be I am wrong.

First I buy a calibration kit.
With it I get a sheet of paper with its correct data,
describing its differences from an ideal calibration kit,
measured by highest precision analyzers.

These differences I have  to put in in USER1(or2) to tell the analyzers math,
which always calculates with the (not existing) ideal calibration kit the
differences to it. And then do the calibration.

The efforts to manufacture a calibration kit near the unreacheable perfection
one and to determine the small differences to this is resposible for the
enormous price differences in calibration kits.

I started the exploration of my network analyzer with the attempt to
measure the quality of my Miny-Ciruits HAT BNC attenuators.
These are specified for DC to 2GHz, so I assumed to get a return loss
of better 20dB at 2 GHz.

I used a male BNC calibration kit from sdr-kits, set the anlyzer to measure S11 and did 1 port SOL calibration.

Then I measured a 30 dB attenuator and got a return loss of 35 db (100kHz) and only 10 db (2 GHz).
Next I used a Spinner BN533863 calibration kit at the same cable but necessarily with and additional
BNC double-female and an BNC to N connector.

After calibration I used the port extensions to move the calibration plane back to the end of the BNC cable (-274ps)
and then measured the same 30 dB attenuator.

This time I got a return loss of 35 db (100kHz) and 24 dB (2 GHz), not changing
with and without 50 Ohm termination at the end oft the attenuator (as expected).

May be other owners of an SVA will try to reproduce a similiar procdure and post
their results, to enlighten us all about the right way to get resonable results
when using these analyzers.

Many users have reported to be happy to have unlocked their SSA3021X+ to an SVA
but I have not read much of their experiences in using these as SVA up to now.

The next procdure will be to build a simple fixuture for smd elements, measure
the fixtures data and save it in the analyzer for correction of the measurement plane.
This is called de-embedding.
After that the real usage of the network analyzer can start.

If anybody has done this already and is willing to share its results this will
helpfull for all users here, the manual is not helpful in this respect, to express
it friendly.

Rob, please ask Siglent for a detailed description of how to add the data for a
calibration kit to USER1 USER2 and why it is not possible to enter data for Load.

tautech:

--- Quote from: kladit on May 20, 2024, 05:00:11 pm ---.........
The next procedure will be to build a simple fixture for smd elements, measure the fixtures data and save it in the analyzer for correction of the measurement plane.
This is called de-embedding.
After that the real usage of the network analyzer can start.

If anybody has done this already and is willing to share its results this will helpful for all users here, the manual is not helpful in this respect, to express it friendly.

--- End quote ---
A simple fixture for TH components I cobbled together here:
https://www.eevblog.com/forum/testgear/test-equipment-anonymous-(tea)-group-therapy-thread/msg4345558/#msg4345558

Inspired by Chris Grossman's work:

tautech:
For completeness, a pic of the 2 SMA TH fixtures I've cobbled together.

They are across and through incarnations with the through with a wire shorting link fitted for Through calibration.
ST42 SMD tweezers case for scale.

kladit:
Here again, this times I hope I will get the pictures up to eevblog.

First S11 calibraition with sdr-kits cal-kits.
The 50 Ohm load is 49.85 Ohm.
Then measure of mini_circuits HAT30+ attenuator Smith and return-loss.

Second calibraition with Spinner BN533863R000.
The 50 Ohm load is 50.05 Ohm.
Then measure of (the same) mini_circuits HAT30+ attenuator Smith and return-loss.


The results of these simple S11 measures with different calibration kits
are very different.
So to make a statement about return loss is impossible with this instrument
(SSA3032X-R). Or is there a failure in my approach to measure?

The question is will it be possible to improve the accurracy of Siglent-VNAs
by input the correct load resistance of the cal-kit in use into the definition
of a USER cal-kit?

With current firmware is it not possible, the input-fields are grayed out.

I have read the high accuracy of an DG8SAQ-VNA(sdr-kits) or the Funkamateur VN-VA5
depends only on the acurracy of the correction data, delivered with an calibration kit.

To measure and get this correction data a very high precision network analyzer is needed,
unpayable for hobbyist/amateurs. But by simply transfer the correction data of an cal-kit
ascertained by a high precision analyzer the accurray of this is transferred to a
relative cheap analyzer with very sophisticated firmware capable to use it.

Can Siglent for their network analyzers provide a firmware which will give them
the same possiblilities and so accuracy as the much cheaper ones metioned above?

ozkarah:
Could you please first calibrate with the Spinner kit, then measure the SDR Kit components.
This should give a brief info about how they differ. 