@tonyalbus Since my own SDM3065X has always stabilised within half an hour of being powered up after an overnight shut down, your observation of an extremely protracted 4 hour warm up time with yours rather suggested that it must have a fault (slow running or failed cooling fan?).
However, after watching your warm up timings test video, I think the problem is quite likely to be blocked fan intake and exhaust ports due to it being tightly hemmed in all round by that rather tightly packed stack of test equipment.
May I kindly suggest you pull it out of the stack so that it has at least 50mm clearance either side of any obstructive surfaces, preferably clear of other localised heat source, and repeat another warm up test? This may well answer your question over its unusually long , out of specification
(30 minutes according to the manual for my meter) warm up time. Sorry for that erroneous information. I think I had that figure in mind from the DSO user guides. The SDM3065X user guide is rather remiss in this regard by failing to mention any warm up time requirement.
The only mention of any warm up time at all is in the data sheet in regard of calibration checks/adjustments where, understandably, they state that rather protracted 90 minute figure to ensure it'll be as stable as possible for such critical checks. Very likely based on the typical 30 minutes warm up time multiplied up by a factor of three to be triply sure its temperature is stable enough for such tests.