Author Topic: Siglent SDS800X HD Review & Demonstration Thread  (Read 39582 times)

ksjh, gf, Martin72, Mortymore, Electro Fan, Bad_Driver, egonotto, newbrain and 18 Guests are viewing this topic.

Online gf

  • Super Contributor
  • ***
  • Posts: 1213
  • Country: de
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #300 on: Today at 11:51:23 am »
Why do you use a lower signal level with the DHO1074?
It looks the same to me:
0 dBm ~= -13.5 dBV (for a 50  \$\Omega\$ Z0}

Sorry, I missed dBm vs. dBV scale.

 
The following users thanked this post: egonotto, newbrain

Offline markone

  • Frequent Contributor
  • **
  • Posts: 707
  • Country: it
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #301 on: Today at 12:05:29 pm »
Why do you use a lower signal level with the DHO1074?
It looks the same to me:
0 dBm ~= -13.5 dBV (for a 50  \$\Omega\$ Z0}

That's correct, dear friend  :-+
 
The following users thanked this post: egonotto

Offline markone

  • Frequent Contributor
  • **
  • Posts: 707
  • Country: it
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #302 on: Today at 12:25:07 pm »

-snip

Let's reverse the picture: if you create dual tone, let say 1MHz 0dBm combined with 1,1 MHz -80 dBm and feed that to Rigol what you will see?
Will you see -80dBm peak?
Following your logic, from your FFT plots this -80dBm should stick out nicely from noise floor.

Here we go with test signal below the  theoretical resolution limit 12x6 = 72 dB, I adjusted the vertical scale to fullfill the ADC range with 0dBm signal with 81mV/div :







Then I applied a 40dB attenuator at the SDG2042x output :



and I started to step down the signal level obtaining the following result :















as expected the DSO is going well below the theoretical limit of 72dB and I would say that I like what I see ;-)

For some reason @ -56dBm  (-96dBm after attenuator) the peak is there with correct level but it does not appear in peak list, my guess is that the Peack list function is not working on the averaged signal but instead on the "clear" trace, I would call it a bug.

Of course I expect that with a 2 components combined signal we will get some spurious here and there, but of course this will NOT cancel signal below 72dB full scale.

HDO's measure coming.
 
The following users thanked this post: Electro Fan, egonotto, 2N3055

Offline markone

  • Frequent Contributor
  • **
  • Posts: 707
  • Country: it
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #303 on: Today at 01:53:35 pm »
Here below DHO measures, same signal approach with DSO input set at 81mV, first spectrum image with 0dBm signal, then the addition of 40dB attenuator and subsequent  signal level stepping down.

Being the DHO math system an utter shit compared to the SDS800x, we here are missing the average function on FFT (AFAIK), dBm scaling, so I stopped at -50dBm (last image, -90dBm at DSO input), where the signal peak is still clearly visible at 90dB under the full scale theoretical level, so the famous -80dB FS signal "stick out" with some margin (last but one image) :













Anyway, shame on Rigol for the missing FFT average feature, this would change a lot the whole experience.
 
The following users thanked this post: egonotto

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6702
  • Country: hr
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #304 on: Today at 03:23:03 pm »

For some reason @ -56dBm  (-96dBm after attenuator) the peak is there with correct level but it does not appear in peak list, my guess is that the Peack list function is not working on the averaged signal but instead on the "clear" trace, I would call it a bug.


For peak to be detected, both detection level and excursion needs to be set.
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11770
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #305 on: Today at 03:41:48 pm »
Vintage 8-bit DSO input -110dBm 1MHz applied.

Offline mawyatt

  • Super Contributor
  • ***
  • Posts: 3307
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #306 on: Today at 04:04:46 pm »
Interesting we see some signal "artifacts" on the Rigol #0 but not on the Siglent #52. Signal is from AWG (SDG2042X) at 2mvpp @ 50 ohms (-50dBm). Both have Flattop Windows, Signal has 4 average (not available on Rigol). BTW these artifacts originate within the Rigol and not from the AWG.

Used a cheap 50 ohm termination and connected to both DSOs, and no the artifact is not from the Siglent contaminating the Rigol or vice versa as both were also checked by themselves!!

Both read the correct level, altho the Rigol is in dBV as it has no dBm reference ability and both produce a nice FFT however we prefer the Siglent display and FFT setup.

Edit: Added another plot with Rigol at a lower sample rate #1, and with Blackman Window #2 and Siglent with Blackman Window #53.

As always YMMV.

Best,
« Last Edit: Today at 06:02:44 pm by mawyatt »
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11770
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #307 on: Today at 04:05:42 pm »
-120dBm 1MHz applied.  Wrong tool for the job but gives some comparison of vintage tech. 

***
Should add that the display update rate, with the scope set to maximize computations is about 7 seconds!  It's too bad that used CPU upgrade did not work out as I am sure it would help in this case. 
« Last Edit: Today at 04:09:16 pm by joeqsmith »
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6702
  • Country: hr
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #308 on: Today at 04:56:06 pm »
SDS2000xHD 1MHz -125dBm applied.
 
The following users thanked this post: Performa01, Electro Fan

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6702
  • Country: hr
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #309 on: Today at 05:17:41 pm »
SDS3104xHD 1MHz -125dBm 1kHz 100% AM modulation applied.
 
The following users thanked this post: Performa01, joeqsmith

Offline mawyatt

  • Super Contributor
  • ***
  • Posts: 3307
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #310 on: Today at 06:04:42 pm »
Impressive results from the LeCroy and Siglent :-+

@ 2N3055 the two sidebands are down 6dB as they should be  :)

Best,
« Last Edit: Today at 06:12:44 pm by mawyatt »
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 
The following users thanked this post: 2N3055

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11770
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #311 on: Today at 06:06:29 pm »
Very impressive to see that AM modulation.  How long does it take to display? 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6702
  • Country: hr
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #312 on: Today at 06:19:21 pm »
Very impressive to see that AM modulation.  How long does it take to display?

Long.
Each acquisition run is 1 sec (100ms/div) , some time for FFT and then 16x averages. Didn't actually measure time. Maybe later.
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6702
  • Country: hr
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #313 on: Today at 06:21:18 pm »
Impressive results from the LeCroy and Siglent :-+

@ 2N3055 the two sidebands are down 6dB as they should be  :)

Best,

Mike,
that is what I wanted to show.  ;D

Best,
Siniša
 

Online Electro Fan

  • Super Contributor
  • ***
  • Posts: 3204
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #314 on: Today at 06:58:40 pm »
SDS2000xHD 1MHz -125dBm applied.

2N3055, thanks for doing and posting the measurements.  A couple questions and maybe a request, please, thanks.

1. Why is the amplitude measurement different with the 2kHD vs 3kHD?
2. What happens when you run the modulation test on the 2k?  Similar results but just takes longer to complete?
 

Online Performa01Topic starter

  • Super Contributor
  • ***
  • Posts: 1654
  • Country: at
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #315 on: Today at 07:05:04 pm »
Since this thread is about the SDS800X HD, you shall not get the impression that it cannot do something similar. It can do it, despite its FFT producing spurious responses at lower RBW (error report has already been filed) – it is just a little less accurate and I had to look for a sweet spot in the spectrum where there were no spurious signals getting into the way…


SDS824X HD_FFT_1.77MHz_-125dBm_AM100_10kHz

The signal was -125 dBm at 1.77 MHz, 100% amplitude modulated with 10 kHz.

The level measurement is 1.175 dB high, and the sidebands are -5.668 and -5.276 dBc.

Well, there should be a little gap between a low end SDS800 and the 2000/3000 series.

Regarding the speed, it takes about 1.28 seconds to produce a single 2 Mpts FFT result (including acquisition). Even tough I’ve used 64x averaging in this example, usable (but less nice looking) results can be expected with just 4 averages already.
 
The following users thanked this post: Martin72

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11770
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #316 on: Today at 07:08:40 pm »
1MHz -125dBm 1kHz 99.9% AM modulation applied. 

You can see the center spur and side lobes but they are lower than the scope's internal noise.   Scope also will not readout this low. Turning off the 3D view does speed up the display but still a few seconds.  Guessing most modern scopes would blow the doors off it.   

My SA can easily measure down this low.  Note the -135dBm peak readout.   I normally keep a 10dB attenuator attached.   Using the right tool for the job can improve the measurements. 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11770
  • Country: us
Re: Siglent SDS800X HD Review & Demonstration Thread
« Reply #317 on: Today at 07:11:54 pm »
Very impressive to see that AM modulation.  How long does it take to display?

Long.
Each acquisition run is 1 sec (100ms/div) , some time for FFT and then 16x averages. Didn't actually measure time. Maybe later.

I assume it uses a round robin buffer and once loaded up, you're about a second update.  Much faster than mine. 


Share me