Author Topic: Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X  (Read 522023 times)
Online Martin72
Re: Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X
« Reply #2000 on: November 30, 2024, 10:52:43 pm »
Pics I´ve made from the reference.
The following users thanked this post: Performa01, Bad_Driver, Detlev
Online KungFuJosh
  • TEAS is real.
« Reply #2001 on: December 01, 2024, 01:36:27 am »
Don't waste time trying to reverse engineer this upgrade, it can't be done. The standard SDM3065X doesn't have the interface components installed which is what makes it a factory only option selected when you purchase it.
It probably can be done. Not having a few minor parts and a mount isn't an issue if you have an SC model to reference what parts are needed. I think that would be the easy part of making the upgrade. The card itself, the limitations of the firmware/interface, and the switching scheme might be slightly trickier than my skills, but I'm sure somebody could figure that out.
I won't likely spend any time on it now that I have a Keithley on the way with the specs I need.
Thanks,
Josh
"Be nice to your children. After all, they are going to choose your nursing home." - Steven Wright
Online KungFuJosh
  • TEAS is real.
« Reply #2002 on: December 03, 2024, 03:42:25 am »
Here's a 3-hour test comparing the DAQ6510 (same analog board as DMM6500) and my SDM3065X. Both had hours of warmup time. Both were set to 10PLC. Everything I could set the same, I did.
The Siglent appeared more stable*, but also more susceptible to environmental change. By the end of the test, my office wasn't just colder, the humidity dropped to 26%. Gross.
The DAQ was less susceptible to environmental change, but was slightly further off from the expected ref value.
Last cal for the DAQ6510 was in April 2024. SDM3065X was September 2023.
Overall, neither of them suck. 😉
Thanks,
Josh
*I found the setting, the DAQ was sampling too fast. I'll eventually do another test with the slower sample rate.
ETA: The SDM3065X is my loudest DMM. The DAQ is much quieter.
« Last Edit: December 03, 2024, 03:50:50 am by KungFuJosh »
The following users thanked this post: Performa01, Korchahin
Online KungFuJosh
  • TEAS is real.
« Reply #2003 on: January 02, 2025, 08:23:48 pm »
Here's a weird bug on the SDM3065X:
Condition: Continuity mode with screensaver active for extended period.
Action: Press key to wake screen (not sure if specific key matters).
Result: Beeper stops working, but still display correct value on screen.
Fix: Adjust Threshold value (up or down and back to same value will work), or power cycle. Nothing else works to remedy it (including deactivating/reactivating beeper or changing volume level).
Firmware V3.01.01.12R1.
Thanks,
Josh
Offline msemhs
« Reply #2004 on: February 14, 2025, 05:11:02 pm »
I just got the SDM3065X-SC, and the FW loaded on it is 3.02.01.13. I can't find any info on their site.  Has anyone seen any release notes for this?  The latest one Siglent has on their site is V3.01.01.12R1
Offline t1d
« Reply #2005 on: February 14, 2025, 11:10:07 pm »
Re: Siglent SDM3065X - Firmware is current @ V3.01.01.12R1
Original Post
https://www.eevblog.com/forum/testgear/siglent-new-bench-dmm-sdm3055/msg4144669/#msg4144669
Beginning at Page 46 @ Post #1134
Ending at Page 47 @ Post #1152
I would like to (finally) address my units low buzzer volume. I have found a convenient mounting location, on the faceplate. I want to discuss 1) any consequences of using that location and, 2) if there are no problems with using the location, how best to wire the buzzer.
Location
I have discovered that there is a sixth meter probe hole, in the front, metal, face bracket. It is below the black, LoSense Probe Hole and to the left of the Red, 10A RMS Probe Hole. See picture. This hole is perfectly sized to accommodate the buzzer. I have not removed the face bracket, yet, but it appears that there may be space to tuck the buzzer in between the front of the metal, face bracket and the back of the plastic front panel.
Note: The hole in the metal, face bracket is covered by a very thin shim of black plastic. I confirmed the hole's location, by shining a flashlight through the plastic faceplate.
Accommodations
A hole would need to be opened in the plastic faceplate. For appearance sake, I recommend only making the opening the size of the small hole that is on the top of the buzzer. I don't think a larger hole in the faceplate will make the buzzer louder, because the hole in the top of the buzzer is the limiting factor.
Wiring
I am concerned with this hack introducing noise into the case. So, I am considering the following method of wiring...
Common copper solder wire wick is woven in the shape of a tube. I would use it to create a shielded wire loom in which to run a twisted pair of wire, for the buzzer. The wire loom would be grounded to the case/Earth GND. Care would be taken to not ground it to Circuit GND. I will cover the loom with heat shrink, as well.
There appears to be room to attach the loom to the side of the case. However, it will have to be routed under the fan housing. There is space to do this, as the PCB is cut out, under the fan.
I would simply glue the buzzer to the metal, face bracket. I prefer a product called Shoe Goo.
I apologize for the lack of case pictures. Health problems prevent that effort, presently.
I look forward to your thoughts and suggestions. Thank you for your help.
EDIT:
I am moving this topic out of the Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X thread, in the hopes that it will make the information easier to be found and to be followed.
Please join us, at the new thread!
https://www.eevblog.com/forum/testgear/siglent-sdm3065x-low-buzzer-volume-repairhack-attempt/new/#new
« Last Edit: February 17, 2025, 01:21:33 pm by t1d »
Offline t1d
« Reply #2006 on: February 14, 2025, 11:15:16 pm »
tautech shared these thoughts:
"Use a audio twin coax for the buzzer cable.
Maybe just gluing the buzzer to the front panel and nothing else will increase its volume."
@ tautech
I found a nice space, between the plastic faceplate and the metal, face bracket. I put these details in the post, above, so you will need to read that post, again.
Great ideas! Thank you!
t1d
The coax shield would be tied to Earth Ground, not Circuit Ground; correct?
« Last Edit: February 14, 2025, 11:18:18 pm by t1d »
Online tautechTopic starter
« Reply #2007 on: February 15, 2025, 07:22:20 am »
You might also consider a replacement/new buzzer, maybe one with a louder output.
Dunno anything about these things but other members will have some ideas.
Online KungFuJosh
  • TEAS is real.
« Reply #2008 on: February 18, 2025, 08:08:17 pm »
At 10PLC measuring VDC the SDM3065X does 6 readings per second here in the US. Is there also some sort of filtering/averaging besides that? Or is it just a delay between readings?
Thanks,
Josh
Online 2N3055
  • Country: hr
« Reply #2009 on: February 18, 2025, 09:33:33 pm »
Thanks,
Josh
ADC is averaging for 10NPLC, pushes result out, rinse and repeat.
It is not quick measure, pause, measure, pause...
Dr. Richard W. Hamming
Online KungFuJosh
  • TEAS is real.
« Reply #2010 on: February 18, 2025, 11:04:47 pm »
Thanks,
Josh
Interesting. Keithley and Keysight both offer additional averaging on top of the PLC, so I was curious if there's something else going on there.
I have the extra averaging off for the attached test. It was interesting watching the Keysight/Siglent compete for biggest PP. 😉
Online mawyatt
« Reply #2011 on: February 18, 2025, 11:50:47 pm »
With KS34465A you can get an additional digit with Math Function MX + B in the display.
Best
« Last Edit: February 19, 2025, 12:15:07 am by mawyatt »
~Wyatt Labs by Mike~
Offline skander36
« Reply #2012 on: February 19, 2025, 03:49:07 pm »
Best
Keithley also do this.
Also Keysight can be "determined" to display 7,5 digit.
But as a professional said here on the forum, we can trust only the numbers that manufacturer warrant for them in the specifications. The rest is "uncertainty fog".
The following users thanked this post: KungFuJosh, mawyatt
Online mawyatt
« Reply #2013 on: February 19, 2025, 04:38:57 pm »
Best
Yes, the DMM6500 has the MX + B Math feature.
Did not know that the KS34465A Digit Mask had 7 1/2 digit selection. Ours don't show this. Is this a select option or something with the newer firmware?
Agree, these "Extra Digits" don't convey much is absolute terms, and are often just a "Random Number" but can help with direct comparisons or relative measurements.
Best
Offline skander36
« Reply #2014 on: February 19, 2025, 05:23:52 pm »
I didn't done myself this mod, someone did this for me. The firmware must be unpacked enable the option (the fw. is common for 6.5 and 7.5 models), and repacked.
Online KungFuJosh
  • TEAS is real.
« Reply #2015 on: February 19, 2025, 05:52:06 pm »
While fun, and I would certainly do it for that purpose, there isn't a lot of value in those extra digits...except maybe the Keithley. It's interesting to me how similar the Siglent and Keysight behave. I'll post a few more screenshots in the other thread.
Online Kleinstein
« Reply #2016 on: February 19, 2025, 06:03:26 pm »
If comparing different meters, one should have the same scale. It is confusing with different scales.
This thread is already long and about the Sigilent meters - so better keep the compare part out of this thread. Much is about different meters and the reference used.
Offline skander36
« Reply #2017 on: February 19, 2025, 06:25:11 pm »
... yeah, is somewhat confusing.
I think they want to signal that the result is math processed, I don't have an explanation about why they choose to display with X.
Online hfleming
  • Country: no
« Reply #2018 on: February 21, 2025, 11:53:56 am »
Can confirm. Just received my new SDM3065X with firmware version 3.02.01.13. Weird that it is not on Siglent’s website.
Offline valcher
  • Country: ru
« Reply #2019 on: February 21, 2025, 06:22:01 pm »
I bought mine in April last year and the firmware was already 3.02.01.13. It will soon be a year.
Online Martin72
« Reply #2020 on: February 21, 2025, 10:19:07 pm »
It may well be that a change in the hardware, due to obsolescence or similar, triggered a change in the firmware.
However, this is only necessary for new production and was therefore not launched as an official update.
So my thoughts on this, I would leave it at that for now.
The following users thanked this post: hfleming, KungFuJosh
Online Mahagam
« Reply #2021 on: March 26, 2025, 10:03:51 pm »
SDM3055. hw/sw on the first picture
Img 2 and 3 - temperature measurement with the only difference: REL button was pressed, but no reaction!
Img 3 and 4 - same type K thermocouple connected, and REL button works perfect
Also, I`ve got some strange readings (text and excel chart below) which looks a little bit strange, but I cannot reproduce this behavior now.
*idn?
Siglent Technologies,SDM3055,SDM35HBD7R1727,1.02.01.28
fetch?
+2.25929622E+01,+2.26003335E+01,+2.25975692E+01,+2.25929622E+01,+2.25948050E+01,+2.25938836E+01,+2.30929622E+01,+2.30929622E+01,+2.30948050E+01,+2.25957264E+01,+2.25920408E+01,+2.26012549E+01,+2.26067833E+01,+2.25994120E+01,+2.25994120E+01,+2.30975692E+01,+2.25938836E+01,+2.25948050E+01,+2.25975692E+01,+2.25994120E+01,+2.26030977E+01,+2.26058619E+01,+2.31040191E+01,+2.26030977E+01,+2.26030977E+01,+2.26067833E+01,+2.26021763E+01,+2.25957264E+01,+2.26003335E+01,+2.26067833E+01,+2.26049405E+01,+2.26113903E+01,+2.31123117E+01,+2.31169188E+01,+2.26178402E+01,+2.26104689E+01,+2.26049405E+01,+2.31067833E+01,+2.26067833E+01,+2.26049405E+01,+2.31067833E+01,+2.26104689E+01,+2.26169188E+01,+2.31187616E+01,+2.26169188E+01,+2.26132331E+01,+2.31104689E+01,+2.31113903E+01,+2.31104689E+01,+2.26095475E+01,+2.26104689E+01,+2.26132331E+01,+2.26113903E+01,+2.26077047E+01,+2.26030977E+01,+2.31021763E+01,+2.26030977E+01,+2.26067833E+01,+2.26067833E+01,+2.26058619E+01,+2.26058619E+01,+2.26049405E+01,+2.26113903E+01,+2.26086261E+01,+2.26067833E+01,+2.26095475E+01,+2.26077047E+01,+2.31086261E+01,+2.26141546E+01,+2.26159974E+01,+2.26123117E+01,+2.26077047E+01,+2.26095475E+01,+2.26132331E+01,+2.31150760E+01,+2.31141546E+01,+2.26058619E+01,+2.31123117E+01,+2.31169188E+01,+2.26159974E+01,+2.26150760E+01,+2.26058619E+01,+2.26003335E+01,+2.26086261E+01,+2.26077047E+01,+2.26040191E+01,+2.26058619E+01,+2.26086261E+01,+2.26086261E+01,+2.26086261E+01,+2.26040191E+01,+2.26058619E+01,+2.26021763E+01,+2.25994120E+01,+2.26040191E+01,+2.26021763E+01,+2.26030977E+01,+2.26030977E+01,+2.31049405E+01,+2.26012549E+01,+2.26049405E+01,+2.26049405E+01,+2.30984906E+01,+2.30984906E+01,+2.25984906E+01,+2.26012549E+01,+2.25984906E+01,+2.30938836E+01,+2.31003335E+01,+2.26040191E+01,+2.31021763E+01,+2.31012549E+01,+2.26049405E+01,+2.26021763E+01,+2.25975692E+01,+2.25948050E+01,+2.30966478E+01,+2.25984906E+01,+2.25994120E+01,+2.31049405E+01,+2.31077047E+01,+2.31058619E+01,+2.31040191E+01,+2.26040191E+01,+2.26030977E+01,+2.26021763E+01,+2.26021763E+01,+2.26058619E+01,+2.26058619E+01,+2.26058619E+01,+2.31104689E+01,+2.26123117E+01,+2.26095475E+01,+2.31095475E+01,+2.31077047E+01,+2.26086261E+01,+2.26104689E+01,+2.31058619E+01,+2.26049405E+01,+2.26049405E+01,+2.26021763E+01,+2.31049405E+01,+2.31077047E+01,+2.26049405E+01,+2.26095475E+01,+2.26077047E+01,+2.26003335E+01,+2.30994120E+01,+2.30994120E+01,+2.31003335E+01,+2.26086261E+01,+2.26095475E+01,+2.26030977E+01,+2.25948050E+01,+2.25957264E+01,+2.30938836E+01,+2.30938836E+01,+2.30948050E+01,+2.30948050E+01,+2.30975692E+01,+2.30966478E+01,+2.31003335E+01,+2.25975692E+01,+2.26003335E+01,+2.26003335E+01,+2.30957264E+01,+2.30892766E+01,+2.25920408E+01,+2.30957264E+01,+2.30938836E+01,+2.25938836E+01,+2.25892766E+01,+2.25938836E+01,+2.30883552E+01,+2.25865124E+01,+2.25837481E+01,+2.30911194E+01,+2.30901980E+01,+2.30883552E+01,+2.30920408E+01,+2.30948050E+01,+2.30994120E+01,+2.30929622E+01,+2.25920408E+01,+2.25883552E+01,+2.30920408E+01,+2.25966478E+01,+2.25957264E+01,+2.26012549E+01,+2.31030977E+01,+2.31003335E+01,+2.31040191E+01,+2.31021763E+01,+2.31049405E+01,+2.31003335E+01,+2.26021763E+01,+2.26012549E+01,+2.25984906E+01,+2.25984906E+01,+2.26030977E+01,+2.25984906E+01,+2.30975692E+01,+2.31030977E+01,+2.25938836E+01,+2.30883552E+01,+2.30948050E+01,+2.31003335E+01,+2.26003335E+01,+2.25975692E+01,+2.25994120E+01,+2.26003335E+01,+2.25994120E+01,+2.25966478E+01,+2.25920408E+01,+2.25948050E+01,+2.31012549E+01,+2.30984906E+01,+2.30938836E+01,+2.26012549E+01,+2.26021763E+01,+2.26012549E+01,+2.25957264E+01,+2.30948050E+01,+2.31040191E+01,+2.26095475E+01,+2.31067833E+01,+2.30994120E+01,+2.26040191E+01,+2.31077047E+01,+2.31067833E+01,+2.26104689E+01,+2.26132331E+01,+2.26113903E+01,+2.26095475E+01,+2.26086261E+01,+2.31150760E+01,+2.31113903E+01,+2.31030977E+01,+2.26040191E+01,+2.31030977E+01
« Last Edit: March 26, 2025, 10:19:22 pm by Mahagam »
Online Mahagam
« Reply #2022 on: March 27, 2025, 05:49:27 pm »
SDM powered on at evening, and bug with temperature readings reappeared.
Again, I`ve got data via SCPI:
*idn?
fetch?
+2.13961050E+01,+2.13970264E+01,+2.13951836E+01,+2.13924194E+01,+2.13933408E+01,+2.14016334E+01,+2.14034762E+01,+2.14025548E+01,+2.18961050E+01,+2.18868909E+01,+2.13896551E+01,+2.13905765E+01,+2.13896551E+01,+2.13951836E+01,+2.13961050E+01,+2.13868909E+01,+2.13841267E+01,+2.13887337E+01,+2.13896551E+01,+2.13850481E+01,+2.13887337E+01,+2.13914980E+01,+2.13914980E+01,+2.13924194E+01,+2.13905765E+01,+2.13887337E+01,+2.13859695E+01,+2.13878123E+01,+2.13896551E+01,+2.18878123E+01,+2.18832053E+01,+2.13868909E+01,+2.13868909E+01,+2.13868909E+01,+2.13942622E+01,+2.13914980E+01,+2.13924194E+01,+2.18859695E+01,+2.13804411E+01,+2.13767554E+01,+2.13868909E+01,+2.13878123E+01,+2.13859695E+01,+2.13896551E+01,+2.13841267E+01,+2.13813625E+01,+2.13758340E+01,+2.13795197E+01,+2.13822839E+01,+2.13914980E+01,+2.13832053E+01,+2.13693842E+01,+2.13730698E+01,+2.18832053E+01,+2.18832053E+01,+2.13776769E+01,+2.13749126E+01,+2.13804411E+01,+2.13813625E+01,+2.13859695E+01,+2.13822839E+01,+2.13749126E+01,+2.18739912E+01,+2.18758340E+01,+2.13776769E+01,+2.13767554E+01,+2.13785983E+01,+2.13776769E+01,+2.13776769E+01,+2.13785983E+01,+2.13785983E+01,+2.13776769E+01,+2.13749126E+01,+2.13749126E+01,+2.18758340E+01,+2.18813625E+01,+2.13795197E+01,+2.13758340E+01,+2.13703056E+01,+2.13721484E+01,+2.13712270E+01,+2.13758340E+01,+2.13795197E+01,+2.13693842E+01,+2.13693842E+01,+2.13666200E+01,+2.13656986E+01,+2.13749126E+01,+2.18776769E+01,+2.18703056E+01,+2.18656986E+01,+2.13638558E+01,+2.13675414E+01,+2.13675414E+01,+2.13666200E+01,+2.18721484E+01,+2.18758340E+01,+2.13684628E+01,+2.13693842E+01,+2.13712270E+01,+2.18656986E+01,+2.13721484E+01,+2.13675414E+01,+2.13656986E+01,+2.13721484E+01,+2.13739912E+01,+2.18703056E+01,+2.18656986E+01,+2.18620129E+01,+2.13656986E+01,+2.13647772E+01,+2.13656986E+01,+2.13610915E+01,+2.13583273E+01,+2.13610915E+01,+2.18629343E+01,+2.13647772E+01,+2.13656986E+01,+2.13656986E+01,+2.13601701E+01,+2.13620129E+01,+2.13647772E+01,+2.18620129E+01,+2.18675414E+01,+2.13601701E+01,+2.13583273E+01,+2.13574059E+01,+2.13574059E+01,+2.18546417E+01,+2.18527989E+01,+2.18500347E+01,+2.13509561E+01,+2.13574059E+01,+2.13564845E+01,+2.13537203E+01,+2.13564845E+01,+2.18537203E+01,+2.13564845E+01,+2.13527989E+01,+2.13555631E+01,+2.13546417E+01,+2.13527989E+01,+2.18601701E+01,+2.13583273E+01,+2.13500347E+01,+2.13509561E+01,+2.13481918E+01,+2.13491132E+01,+2.18546417E+01,+2.13509561E+01,+2.13527989E+01,+2.18555631E+01,+2.18537203E+01,+2.18481918E+01,+2.18546417E+01,+2.13574059E+01,+2.13527989E+01,+2.18426634E+01,+2.13463490E+01,+2.13592487E+01,+2.18555631E+01,+2.18481918E+01,+2.18417420E+01,+2.13371350E+01,+2.13362136E+01,+2.13518775E+01,+2.18546417E+01,+2.13509561E+01,+2.13417420E+01,+2.13491132E+01,+2.13509561E+01,+2.13472704E+01,+2.18463490E+01,+2.13472704E+01,+2.13472704E+01,+2.13463490E+01,+2.13454276E+01,+2.13491132E+01,+2.18472704E+01,+2.13463490E+01,+2.13481918E+01,+2.13445062E+01,+2.13426634E+01,+2.13426634E+01,+2.18380564E+01,+2.18472704E+01,+2.18445062E+01,+2.13417420E+01,+2.18435848E+01,+2.18398992E+01,+2.18389778E+01,+2.13417420E+01,+2.13398992E+01,+2.18380564E+01,+2.18426634E+01,+2.18389778E+01,+2.18362136E+01,+2.13352921E+01,+2.13398992E+01,+2.13426634E+01,+2.18389778E+01,+2.18343707E+01,+2.13371350E+01,+2.13408206E+01,+2.13417420E+01,+2.13352921E+01,+2.13362136E+01,+2.13389778E+01,+2.13408206E+01,+2.18352921E+01,+2.18352921E+01,+2.18316065E+01,+2.18325279E+01,+2.18362136E+01,+2.18398992E+01,+2.18334493E+01,+2.18352921E+01,+2.18352921E+01,+2.18306851E+01,+2.18297637E+01,+2.18288423E+01,+2.18288423E+01,+2.18288423E+01,+2.18269995E+01,+2.13269995E+01,+2.13260781E+01,+2.13251567E+01,+2.18269995E+01,+2.18279209E+01,+2.18325279E+01,+2.18306851E+01,+2.18251567E+01,+2.18279209E+01,+2.18306851E+01,+2.18316065E+01,+2.18325279E+01,+2.18297637E+01,+2.18251567E+01,+2.18288423E+01,+2.18297637E+01,+2.18196282E+01,+2.13187068E+01,+2.18205496E+01,+2.18233139E+01,+2.18288423E+01,+2.18325279E+01,+2.18297637E+01,+2.18242353E+01,+2.18214711E+01,+2.18260781E+01,+2.18269995E+01,+2.18260781E+01,+2.18251567E+01,+2.18187068E+01,+2.18260781E+01,+2.18260781E+01,+2.18168640E+01,+2.18196282E+01,+2.18168640E+01,+2.13196282E+01,+2.18214711E+01,+2.18223925E+01,+2.18177854E+01,+2.18251567E+01,+2.18279209E+01,+2.18214711E+01,+2.18242353E+01,+2.18168640E+01,+2.18168640E+01,+2.13196282E+01,+2.13205496E+01,+2.18168640E+01,+2.18177854E+01,+2.18177854E+01,+2.18223925E+01,+2.18223925E+01,+2.18214711E+01,+2.18205496E+01,+2.18242353E+01,+2.18214711E+01,+2.13177854E+01,+2.18205496E+01,+2.18205496E+01,+2.18196282E+01,+2.18233139E+01,+2.18140998E+01,+2.13187068E+01,+2.18150212E+01,+2.18122570E+01,+2.18150212E+01,+2.18113356E+01,+2.18122570E+01,+2.18140998E+01,+2.18168640E+01,+2.18196282E+01,+2.18104142E+01,+2.18085714E+01,+2.18113356E+01,+2.18150212E+01,+2.18104142E+01,+2.18159426E+01,+2.18122570E+01,+2.18140998E+01,+2.18177854E+01,+2.13140998E+01,+2.18104142E+01,+2.18067285E+01,+2.18048857E+01,+2.18067285E+01,+2.18094928E+01,+2.18104142E+01,+2.18104142E+01,+2.18104142E+01,+2.18076500E+01,+2.18058071E+01,+2.18094928E+01,+2.18113356E+01,+2.18067285E+01,+2.18021215E+01,+2.18048857E+01,+2.18067285E+01,+2.18085714E+01,+2.18076500E+01,+2.18039643E+01,+2.18048857E+01,+2.18094928E+01,+2.18039643E+01,+2.18021215E+01,+2.18085714E+01,+2.18076500E+01,+2.18002787E+01,+2.18076500E+01,+2.18085714E+01,+2.18067285E+01,+2.18058071E+01,+2.17984359E+01,+2.18012001E+01,+2.18058071E+01,+2.18002787E+01,+2.17965931E+01,+2.18012001E+01,+2.18058071E+01,+2.18048857E+01
Online Kleinstein
« Reply #2023 on: March 27, 2025, 08:41:07 pm »
Looks like 2 values about 0.5 K separated. With a type K thermocouple this would be some 20 µV, or quite a bit. The steps are sharp and too fast to be a thermal contact problem.
It is possible that the problem is with the probe and not with the meter. The meter can not tell the difference between a probe at the cold junction temperature and a normal wire short. So one could use a normal short instead of the TC probe for a test.
Another way to test the probe would be to look at the restance of the probe in 2 wire mode. One possible failure mode for the probe is a bad contact. A good TC should show a reasonable stable resistance with no jumps if moved.
Online Mahagam
« Reply #2024 on: March 27, 2025, 08:59:32 pm »
Exactly!
I did such test. Short wire (~0.14Ω) shows me a ~25.0°C. But it can stay stable for around 1 hour, and then can suddenly start to jump to 0.5° up and back. It`s not so obvious when your are watching at numbers. But its clear when you looking at the chart.