Siglent SDG1032 Amplitude not constant when sweeping
« on: May 18, 2024, 10:42:48 am »
Is there a way to get a constant amplitude, or just much less?
Sweep goes from 1KHz to 30Mhz.

"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1536
  • Country: be
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #1 on: May 18, 2024, 10:58:34 am »
Some remarks:
- I see the scope in sampling at 50MS/s, this is too slow for capturing 30MHz (minimum 60MS/s required)
- I presume you are using the 1M input on the scope? Can you give it a try on 50 Ohm input (with a 50 ohm adapter is ok)

 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #2 on: May 18, 2024, 11:21:57 am »
- I see the scope in sampling at 50MS/s, this is too slow for capturing 30MHz (minimum 60MS/s required)
- I presume you are using the 1M input on the scope? Can you give it a try on 50 Ohm input (with a 50 ohm adapter is ok)

Yes, with 50MS/s, the waveform looks rather bad, but its not the form that matters. I have checked with 2Gsa/s, which confirm this issue.

I do not have 50 Ohm, but i set the load option of the gen to HighZ. Should that help?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6915
  • Country: hr
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #3 on: May 18, 2024, 11:28:55 am »
How many topics will you open?
Will you open individual topic for EVERY SINGLE thing you don't understand and then blame equipment for your lack of knowledge.
Are you going to open a topic per buton per piece equipment you own.
3 topics on same generator so far?
 
The following users thanked this post: Mortymore

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4126
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #4 on: May 18, 2024, 11:43:48 am »
- I see the scope in sampling at 50MS/s, this is too slow for capturing 30MHz (minimum 60MS/s required)
- I presume you are using the 1M input on the scope? Can you give it a try on 50 Ohm input (with a 50 ohm adapter is ok)

Yes, with 50MS/s, the waveform looks rather bad, but its not the form that matters. I have checked with 2Gsa/s, which confirm this issue.

I do not have 50 Ohm, but i set the load option of the gen to HighZ. Should that help?
|O |O |O

Could you really stop fooling around and learn even the basics.
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: newbrain

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4126
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #5 on: May 18, 2024, 12:39:33 pm »
Here SDG1000X  sweep from 1Hz to 30MHz
Signal set in SDG, Sine, 0dBm and sweep time 10ms
In scope (SDS3104X HD,  image zoom window is top of signal. (10x vertical magnification)
Cursors in zoom window..  if I drop signal level around 0.5dB then signal top is just as cursor Y2 level.

And naturally this include also oscilloscope level imperfection (nonsense in this case).


Signal from SDG to oscilloscope 50 ohm input directly using 1m Suhner RG223/U

Where is this problem? 
I don't see any other problem here than your fooling around with total lack of enough knowledge and experience.


« Last Edit: May 18, 2024, 12:42:35 pm by rf-loop »
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: Performa01, TurboTom, newbrain, Mortymore, KungFuJosh, Martin72

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #6 on: May 18, 2024, 01:47:44 pm »
Where is this problem? 
I don't see any other problem here than your fooling around with total lack of enough knowledge and experience.

Damn, i thought i could get through with my things that i want to do, with that level of knowledge i have. Seems i cant do so. Thank you investing your time to show that to me.

I thought that this problem would only be problematic with fast rising signals.  :o

Should it work with sticking a 50R resitor in a y-adapter?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27232
  • Country: nl
    • NCT Developments
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #7 on: May 18, 2024, 01:53:22 pm »
Where is this problem? 
I don't see any other problem here than your fooling around with total lack of enough knowledge and experience.

Damn, i thought i could get through with my things that i want to do, with that level of knowledge i have. Seems i cant do so. Thank you investing your time to show that to me.

I thought that this problem would only be problematic with fast rising signals.  :o

Should it work with sticking a 50R resitor in a y-adapter?
Yes, a 50 Ohm resistor would do just fine for 30MHz. But I think the problem you are seeing might be due to overdrive recovery as the sweep you are showing shouldn't be so attenuated (with peak-detect enabled as you seem the have done already) even when using an 1MOhm input especially at the lower frequencies (assuming your sweep starts at the low frequency and ends at 30MHz). When you do the measurement again, make sure the entire amplitude of the signal fits vertically on the screen of the oscilloscope to be sure no effect from the analog front-end of the oscilloscope distorts the signal. If you still get a similar attenuation, zoom in at the start of the sweep further to get an idea what the waveform looks like.
« Last Edit: May 18, 2024, 01:56:08 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1536
  • Country: be
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #8 on: May 18, 2024, 04:36:53 pm »
I do not have 50 Ohm, but i set the load option of the gen to HighZ. Should that help?

No, this will not help.

HighZ-mode and 50 ohm-mode on the siggen just change the signal output level. Your signal generator has a 50 ohm resistor in its output path. When outputting a signal to a 50 ohm input receiver, the receiver would only "see" half the signal (due to the voltage divider created by the two 50-ohm resistors in "series"). By configuring the siggen in 50ohm-mode, the siggen doubles its output voltage to ensure the receiver sees the signal level you entered on the siggen. This is the ONLY difference between these two modes, so it will not help...
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #9 on: May 18, 2024, 05:34:20 pm »
Yes, a 50 Ohm resistor would do just fine for 30MHz. But I think the problem you are seeing might be due to overdrive recovery as the sweep you are showing shouldn't be so attenuated (with peak-detect enabled as you seem the have done already) even when using an 1MOhm input especially at the lower frequencies (assuming your sweep starts at the low frequency and ends at 30MHz). When you do the measurement again, make sure the entire amplitude of the signal fits vertically on the screen of the oscilloscope to be sure no effect from the analog front-end of the oscilloscope distorts the signal. If you still get a similar attenuation, zoom in at the start of the sweep further to get an idea what the waveform looks like.

Here are the results. They dont look much better. If i turn to 1X, it lookes like without termination. The wave fits into the vertical space. (see the cursors)
I guess i should go and learn basics then... that time is probably better spent there.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #10 on: May 18, 2024, 06:14:27 pm »
I have found a really helpful video:


I now put the probe directly into the AWG. But the result is only somewhat better. (see picture)
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1536
  • Country: be
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #11 on: May 18, 2024, 06:29:35 pm »
I now put the probe directly into the AWG. But the result is only somewhat better. (see picture)

Are you using a scope probe instead of a BNC cable? A scope probe uses resistive coax and cannot be used for this purpose. Just do "siggen => bnc cable => 50-ohm adaptor => Scope"
 
The following users thanked this post: KungFuJosh

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #12 on: May 18, 2024, 06:45:31 pm »
I now put the probe directly into the AWG. But the result is only somewhat better. (see picture)

Are you using a scope probe instead of a BNC cable? A scope probe uses resistive coax and cannot be used for this purpose. Just do "siggen => bnc cable => 50-ohm adaptor => Scope"
But what do i do, if i want to measure that signal? It confuses me, that i cannot use a probe for measuring a signal, as you say.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27232
  • Country: nl
    • NCT Developments
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #13 on: May 18, 2024, 07:20:41 pm »
I now put the probe directly into the AWG. But the result is only somewhat better. (see picture)

Are you using a scope probe instead of a BNC cable? A scope probe uses resistive coax and cannot be used for this purpose. Just do "siggen => bnc cable => 50-ohm adaptor => Scope"
But what do i do, if i want to measure that signal? It confuses me, that i cannot use a probe for measuring a signal, as you say.
Follow this rule: Either use the probe as-is (no 50 Ohm termination at the oscilloscope side) OR use a BNC cable with 50 Ohm terminator at the oscilloscope side.

Did you adjust the probe correctly? See probe compensation procedure.

But this still doesn't explain why the amplitude is near zero at low frequencies. First use a BNC cable between the generator and the oscilloscope and do the measurement again. Leave the probe out.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #14 on: May 18, 2024, 09:05:49 pm »
It made "click"

The solution i did: i put a 50 Ohm termination on the point, where i measured with the probe. It doesnt look perfect, but good enough for me. (Its  51Ohm but never mind)

I tried to put up a little breadboard with a trimmer, to get to exactly 50Ohm, but then it looked like a cross section of a mountain range.  ;D
So i guess, everything, that is behind the 50 Ohm termination, can still reflect waves. Or is it because of the rather bad connections of the breadboard?

"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27232
  • Country: nl
    • NCT Developments
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #15 on: May 18, 2024, 10:01:45 pm »
This sweep is shorter compared to the previous one. For repeatability, don't change the signal settings you are using. Unless you are using an extremely long coax cable, reflections shouldn't be an issue at all. Either way, you should zoom in on the areas which show low amplitude. A potential problem could be the decimation (sample skipping) in the DSO to create the screen causes the signal envelope shown to be lower than it actually is.
« Last Edit: May 18, 2024, 10:04:06 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #16 on: May 19, 2024, 05:37:48 am »
This sweep is shorter compared to the previous one. For repeatability, don't change the signal settings you are using. Unless you are using an extremely long coax cable, reflections shouldn't be an issue at all. Either way, you should zoom in on the areas which show low amplitude. A potential problem could be the decimation (sample skipping) in the DSO to create the screen causes the signal envelope shown to be lower than it actually is.

The sweep has the same length. It did not start at 0ms. I had no trigger there.

From the video of dave, i think reflection DO play a role? Your statement conflicts with my understanding of what i learned from the video. Did i got something wrong?

I could not reproduce the waveform on that other picture, but a different one. I have testet it from 10ksa/s to 500Msa/s: The amplitudes stayed the same on the zoomed out view.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #17 on: May 19, 2024, 09:37:14 am »
This is now the best result i could achieve on the sweep.

Could it be improved by some more measures?

"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27232
  • Country: nl
    • NCT Developments
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #18 on: May 19, 2024, 09:40:35 am »
Can you please show 3 screendumps where you can see the actual waveforms at the start, middle and end of the sweep? And make sure that the signal fits on the screen of your oscilloscope vertically? Only then it is possible to understand what is the problem with your measurement.
« Last Edit: May 19, 2024, 09:44:08 am by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline ebastler

  • Super Contributor
  • ***
  • Posts: 6668
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #19 on: May 19, 2024, 09:51:30 am »
Could it be improved by some more measures?

You should really measure with the probe directly at the generator's BNC output. I am not quite sure from your photo, but it seem like you still have a length of two loose wires as part of your connection, which obviously don't have 50 Ohm impedance and may cause reflections.

There are dedicated adapters for that, e.g. here, or they come included with some scope probes. (I got mine with the Testec probes I bought a few years ago.) But in a pinch you can improvise by plugging the bare probe tip into the BNC jack and making the ground connection with a piece of wire or foil.
 

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #20 on: May 19, 2024, 10:06:13 am »
... but it seem like you still have a length of two loose wires as part of your connection, which obviously don't have 50 Ohm impedance and may cause reflections.

There are dedicated adapters for that, e.g. here, or they come included with some scope probes. (I got mine with the Testec probes I bought a few years ago.) But in a pinch you can improvise by plugging the bare probe tip into the BNC jack and making the ground connection with a piece of wire or foil.

Look at the picture again. Its in plain sight  ;-)

I had that adapter, and i tried it, it looks better then, but i would like to use that signal of the gen, and not just measure it. I gues the right thing would then to build my own ouput stage, that would limit problematic voltages, and cope with the termination. But i want to finish my project and dont do new ones.  :o

Maybe there is something to buy like this?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline ebastler

  • Super Contributor
  • ***
  • Posts: 6668
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #21 on: May 19, 2024, 10:38:38 am »
Look at the picture again. Its in plain sight  ;-)

Sorry, we may be misunderstanding each other. What is in "plain sight" for me in the photo is

- a long red/black wire which I assume comes from the generator,
- a resistor at the end of that wire,
- a pice of cardboard (?) in a crocodile clip,
- a probe lying next to it, apparently unconnected to anything.

So I am not sure at all how you had your probe connected to the generator.  ::)

Edit: Oh, and if you need to deliver that 1..30 MHz sweep to your device under test, you need to provide a proper input for it on the DUT. (BNC, 50 Ohm terminated.) Then measure with the high-impedance probe tip right where the signal enters your DUT.  I was under the impression that the sweep was just for performance/stability testing of the generator, not related to a current project.

Edit²: Or did your "in plain sight" refer to the scope probe, which does look like a Testec probe? So you have one of those BNC adapters? If so, I would love to see a frequency sweep measured directly at the generator's BNC output, with nothing else connected. (And the same for the spurious voltages you see when powering off the generator, discussed in the other thread.) It's the cleanest connection you can make with the equipment at hand.
« Last Edit: May 19, 2024, 11:42:56 am by ebastler »
 
The following users thanked this post: Aldo22

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #22 on: May 19, 2024, 11:15:24 am »
Look at the picture again. Its in plain sight  ;-)

Sorry, we may be misunderstanding each other. What is in "plain sight" for me in the photo is

- a long red/black wire which I assume comes from the generator,
- a resistor at the end of that wire,
- a pice of cardboard (?) in a crocodile clip,
- a probe lying next to it, apparently unconnected to anything.

So I am not sure at all how you had your probe connected to the generator.  ::)

Edit: Oh, and if you need to deliver that 1..30 MHz sweep to your device under test, you need to provide a proper input for it on the DUT. (BNC, 50 Ohm terminated.) Then measure with the high-impedance probe tip right where the signal enters your DUT.  I was under the impression tjhat the sweep was just for performance/stability tesintg of the generator, not related to a current project.
That resistor is the termination, i thought i mentioned it. Should have mentioned it again.

Didnt notice, that the cardboard did actually hide the pin header (where the probe would go) on that angle. It would not stick in there on its own.


"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online Aldo22

  • Frequent Contributor
  • **
  • Posts: 790
  • Country: ch
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #23 on: May 19, 2024, 11:24:08 am »
Edit: Oh, and if you need to deliver that 1..30 MHz sweep to your device under test, you need to provide a proper input for it on the DUT. (BNC, 50 Ohm terminated.)
Or SMA with adapter, which is handier.
I always try to keep the distances as short as possible so that there is less noise etc.  ;)
« Last Edit: May 19, 2024, 11:28:34 am by Aldo22 »
 
The following users thanked this post: ebastler

Offline eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 797
  • Country: de
Re: Siglent SDG1032 Amplitude not constant when sweeping
« Reply #24 on: May 19, 2024, 11:28:23 am »
Edit: Oh, and if you need to deliver that 1..30 MHz sweep to your device under test, you need to provide a proper input for it on the DUT. (BNC, 50 Ohm terminated.)
Or SMA with adapter, which is handier.
I always try to keep the distances as short as possible so that there is less noise etc.
Are you supporting the cable? If not, you put a lot of stress on the right side SMA adapter.

What is this board??
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 