Hello there, I stumbled across the SDG6022X as an signal generator for a reasonable price. Looking at the data sheet I am missing information about rise/fall times for different output voltages. Does anybody know what times to expect for 5, 10, 20V outputs?
Thanks, Jan
Performa01:
I only have SDG6052X here, but this shouldn’t make any difference:
Firstly, I won’t be able to show you the rise time at 20 Vpp, because this instrument, like any other AWG in this class, can only output 10 Vpp into a 50 ohm load. There’s certainly no way to get a fast pulse with single digit nanosecond risetime into a high-Z input without significant degradation. So we have to stick to 50 ohms.
The fastest specified rise-time of the SDG6000X is 2 ns. At least my specimen is a little faster and manages to stay below 1.7 ns at 5 Vpp:
SDG6052X_5Vpp_T2ns
For 10 Vpp it was a bit of a hassle already, because that level is too high for a 50 ohm DSO input. I had to fit a 6 dB inline attenuator and consequently you’ll still see only a 5 Vpp pulse in the screenshot. But it is 10 Vpp originally – you can also see that the rise time got quite a bit slower now (~2.3 ns).
SDG6052X_10Vpp_T2ns_A6dB
JanL:
Thank you very much. It helps a lot with my decision.
Performa01:
Attached is an example for the SDG6052X after a power loss.
It was set to a 100 Hz, 1 Vpp sine wave.
Utility – System – PowerOn was set to “Last”.
Utility – Output Setup – Power On State was set to “On”
Utility – Output Setup – Amplitude was set to 20.000 Vpp, i.e. no limit set.
That’s how a well behaved AWG handles mains power outage and -return.
SDG6052X Sine_100Hz_1Vpp_Power Loss
Martin72:
What I ask myself all the time is whether the "gap" between SDG2000 and SDG6000 will ever be filled....