Siglent SDL1000X/SDL1000X-E Electronic Load
Martin72:
I actually have everything together for the tests.
I also bought the small black power supply unit to have more than 3A available (10A).
Then an LED driver from meanwell with 1050mA constant current.
With the other power supplies I can test the maximum power, e.g. the 2x60V to 120V and 3 A are 360W.
Plus my Tektronix current clamp and a small siglent scope.
So it will start soon.
One thing I tried out earlier was the meanwell led driver.
You can actually see a significant ripple on the current.
More on this at the weekend.
ddrl46:
Has anyone been successful in performing an adjustment on this unit? The procedure in the service manual is a bit confusing...
SHF:
Hmm black power supply not nice ;D ;D
Martin72:
OK,
Then I already have a black one. ;)
--- Quote ---The procedure in the service manual is a bit confusing...
--- End quote ---
That's why Siglent withdrew it again, there were too many requests regarding the calibration.
However, it must also be said that calibration is only due when the load has run out of its specification.
This in turn can be tested and/or the "Verification Performance Test" from the service manual can be carried out.
theblinkingman:
Can somebody please post the service manual with the calibration procedure again?