Products > Test Equipment
Siglent SDM3045X mA current measurement - weird behaviour
<< < (9/13) > >>
tautech:

--- Quote from: ahope on November 12, 2018, 07:31:07 pm ---The internal fuse is a 12A slow ceramic, the rear fuse a fast 10A. makes absolutley no sense that the internal should blow before the rear fuse.

This is the one and only fuse I have had to replace on this instrument.

--- End quote ---
No sense at all !

As mentioned earlier I've seen this before in SDM3065X.
Questions to the factory have been asked.
nctnico:

--- Quote from: ahope on November 12, 2018, 07:31:07 pm ---A little background on the blown fuse.

The instrument was almost unused, bought new a couple of montgs before.
I was probing inside a chassis and accidentally touched the 220v and poof! it blew. quite a bang actually, those fuses dont go quietly,

The internal fuse is a 12A slow ceramic, the rear fuse a fast 10A. makes absolutley no sense that the internal should blow before the rear fuse.

This is the one and only fuse I have had to replace on this instrument.

--- End quote ---
If you compare the graphs of the fuses you'll probably see the time to open is identical so it is not a surprise they both blew. In mains installations the rule is to go two sizes up when putting fuses in series to get proper selectivity. Over here common values are 4,6,10,16,20,25,30 and 35A. If you have a 10A fuse then the upstream fuse should be at least 20A to make sure only the 10A fuse blows. Siglent probably put the 12A fuse on the board to make sure there is some safety when the user replaced the fuse with a piece of wire. Unfortunately 12A is way to close to 10A for this trick to work. A better way would be to install a user replaceable 7A fuse to make sure the internal 12A fuse doesn't blow. This likely doesn't solve the problem with the current sense resistor becoming damaged though. A fuse isn't the answer to protecting components. A fuse is there to prevent fire.

For example: the VC8145 bench DMMs have a 13A fuse even though they are rated for 20A for 15s (the latter is also printed on the front of the VC8145). Remember a fuse is designed to carry the rated current forever so it will only blow when there is a significantly larger current flowing through it.
tautech:
Any observant hobbyist has witnessed the different characteristics between FB and SB fuses even of the same A rating and IME 12A SB (on PCB) and 10A FB (rear panel) would indeed be sufficient to protect an instrument and give correct cascading of fuse failure/opening.

But here this has not happened (2x IME) so tests must be made to discover why.

Could the fuses be installed incorrectly or the fuse rating characteristics be called into question ?

Careful tests need be made to discover if fuse behavior is 'as designed' and if it isn't, could this have contributed to shunt damage.

Has the factory changed fuse supplier.....there are many things yet to be discovered.

nctnico:

--- Quote from: tautech on November 12, 2018, 08:34:19 pm ---Any observant hobbyist has witnessed the different characteristics between FB and SB fuses even of the same A rating and IME 12A SB (on PCB) and 10A FB (rear panel) would indeed be sufficient to protect an instrument and give correct cascading of fuse failure/opening.

--- End quote ---
That depends entirely on the current and voltage. Compare Littlefuse 216 and 218 series for example. The fast acting opens within a second at 25A and the slow one at 30A. All in all the difference is much smaller than you'd expect. And at mains voltages there is also the effect of arcing which doesn't interrupt the circuit right away. There is a good reason why mains installations require a large margin between downstream/upstream fuses! There is no sense in denying that.
rf-loop:

--- Quote from: ahope on November 12, 2018, 07:31:07 pm ---A little background on the blown fuse.

The instrument was almost unused, bought new a couple of montgs before.
I was probing inside a chassis and accidentally touched the 220v and poof! it blew. quite a bang actually, those fuses dont go quietly,

The internal fuse is a 12A slow ceramic, the rear fuse a fast 10A. makes absolutley no sense that the internal should blow before the rear fuse.

This is the one and only fuse I have had to replace on this instrument.

--- End quote ---

Do you have still available data how these fuses was marked...  All markings what can find in one fuse end or how ever they have marked voltage, current and speed class. Example (F 10A 250V) or how ever they are.

Also, it need note that some glass fuses 5x20mm have really very low interrupt rating what may reach quite easy in total short circuit (depending source impedande)

Then over current - time - break curves can be really unpredictable, there may be really wide tolerances. In one manufacture lot one 10A fuse may break after 10s 15A but and 1s 25A and 0.1s 60A and 0.01s perhaps 200A and 0.001s perhaps 1000A. And then, min and max curves may differ really lot.
Of course these depends fuse type. But fast or ultra fast and slow fuses have not very big difference in some circumstances in practice even if I2t is many fold.
Some 5x20 glass fuse may have as low as 35A - 100A interrupt rating. (means that if current is more big it can not break in worst cases, example Littelfuse type 218   <4A and example 10A  have 100A interrupt rating and some other type have 10kA)

If really want build, using fuses, even some kind of circuit (passive) component protection it really need do lot of study and tests and very extremely carefully and clever select fuses. Mostly no one do it seriously. And it can understand because result is still poor or just borderline And still not for usual semiconductors.
Fuse is, as nctnico told, more like preventing fire etc.  (as told, of course also some amount circuit protection. But as can see if really look fuses true data... 10A fuse make still over 1000A short peak possible.

Why internal 12A blow and not external 10A.  There may be several and also natural reason. One reason may be fuses tolerances in over current time curves. It can be very possible that if short circuit current is example 50A first blow 12A and after it blow 10A is still ok condition if both are not in they tolerances middle ones.. if other is from other side and other is from other side of normal distribution it is very possible even if external 10A is F(ast) and internal 12A fuse is T(race)

Even if there is example under 10A fuse (yes it can and 10A not continuous measurement is possible)
There can be  still very high maximum short circuit current (if source can feed it). In this game it do not mean lot if short maximum  is reduced by fuse to 200A instead of 250A or how ever it is)

Here one tiny example about time current curve (average - and it mean that every fuse have min and max curve and area between these are big.)

But then, here is just ONE example about one type min and max curves. All fuses have different these curves. This may give some imagine how tolerances may affect and give one possible explanation also for this case. As can see differences in min and max is huge.