Title: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on June 08, 2019, 08:49:35 am
Hi,

just noticed recently that when I did some resistor calibrations with my SDM3065X that You seemingly can´t switch between different displays -say Trend-chart and Histogramm- without loosing the logged data.
Since the data is been logged already (in a volatile memory), I see no good reason at all for this behaviour.
In fact was I quite disappointed about this, as it would mean that You have to repeat measurements to get different presentations  :--
The manual doesn´t help here .... I´m inclined to say of course not, as any online or paper help is usually utterly useless with one´s specific prob.  :palm:
Am I just missing out on something, or is this a real Bug?

regards
Calvin



Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: kado on June 08, 2019, 09:40:55 am
Yes, can double that behaviour and Siglent never developed this great disappointment in any FW since release of this devices !!!! That is a shame!
All other meter with graphic display are able to handle this correct. And when you see what great amount of memory this device have build in one can’t understand why not implemented. [emoji22]

Karsten


Gesendet von iPhone mit Tapatalk
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: tautech on June 08, 2019, 11:02:07 pm
Yes, can double that behaviour and Siglent never developed this great disappointment in any FW since release of this devices !!!! That is a shame!
All other meter with graphic display are able to handle this correct. And when you see what great amount of memory this device have build in one can’t understand why not implemented. [emoji22]

Karsten
So this firmware from earlier this year is no help ?
http://old.siglentamerica.com/USA_website_2014/Firmware&Software/firmware/SDM3065X_3.01.01.06R2_EN.zip (http://old.siglentamerica.com/USA_website_2014/Firmware&Software/firmware/SDM3065X_3.01.01.06R2_EN.zip)
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Electro Fan on June 09, 2019, 03:11:29 am
Yes, can double that behaviour and Siglent never developed this great disappointment in any FW since release of this devices !!!! That is a shame!
All other meter with graphic display are able to handle this correct. And when you see what great amount of memory this device have build in one can’t understand why not implemented. [emoji22]

Karsten


Gesendet von iPhone mit Tapatalk

It's much easier to put a small ton of memory in a box than useful software.

In fairness to Siglent, I think they are trying and doing a pretty good job of developing reliable high performance hardware, pretty good functions and features, and competitive prices.  I don't think Siglent's challenge is writing good software or firmware, I think their challenge is the same as their competitors:  figuring out the functions and features (including User Interfaces) that are the most valuable for their customers.  I think once they identify the market requirements their next challenge is prioritizing the requirements.  At that point it's just a matter of getting those requirements into the upcoming releases (ie, managing resource constraints).

This in turn leads to the value of EEV forums - both for the manufacturers and the users.  To the extent we can describe with reasonable detail the required functions and features, and the reasoning/justification/benefits of those functions and features (especially the user benefits that could possibly produce a ROI for the users), the manufacturers can better prioritize their dev releases.  Might come down to which platforms have the most horsepower for performance and flexibility for expansion and which manufacturers have the best insight in terms of who to listen to.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Electro Fan on June 09, 2019, 03:16:38 am
And yes, what you see in a trend chart should more or less be viewable in a histogram without recreating datasets.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: kado on June 09, 2019, 10:29:58 am
They copy most of the GUI and the layout from the well known Keysight devices but if you think one can get the same working power for less money you feel hit yourself with a knife in a finger.
Think you run a test measurement over 12 hours see the trend chard and you push one key to see the same data in a histogram all data are lost (futsch) and you start a new measurement! What a shit! I feel like a blizzard hit me at the first time this happened to me ( 2 years ago, but no one rant’s about this since release of this series ).

Karsten


Gesendet von iPhone mit Tapatalk
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Electro Fan on June 10, 2019, 05:00:02 am
Maybe tautech can get Siglent to fix this.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on June 10, 2019, 08:43:10 am
Hi,

@kado ... this thread is not about Keysight or another A-brand, nor is it about Siglent-bashing.
One could certainly expect a better behavour for the amount of premium the A-brands ask for .... but they aren´t free of bugs and failures also.
Btw. I´m still waiting on an A to my (honest) Q to the -here also active- Keysight staffer what the added value could be in buying a Keysight device instead of a Siglent.
No answer is an answer too :--

Back to the Prob.
I generaly try to have the latest FW installed on my Siglents and it is rev. 3.01.01.06R2 on HW 01-01-00-05-00.
I appreciate very much that Siglent ´listens´ to what´s going on here and the help and support offered by Tautech.
Let´s just hope that Siglent fixes the Bug soon and releases a new FW rev.

regards
Calvin



Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: tautech on June 10, 2019, 09:15:02 am
Hi,

@kado ... this thread is not about Keysight or another A-brand, nor is it about Siglent-bashing.
One could certainly expect a better behavour for the amount of premium the A-brands ask for .... but they aren´t free of bugs and failures also.
Btw. I´m still waiting on an A to my (honest) Q to the -here also active- Keysight staffer what the added value could be in buying a Keysight device instead of a Siglent.
No answer is an answer too :--

Back to the Prob.
I generaly try to have the latest FW installed on my Siglents and it is rev. 3.01.01.06R2 on HW 01-01-00-05-00.
I appreciate very much that Siglent ´listens´ to what´s going on here and the help and support offered by Tautech.
Let´s just hope that Siglent fixes the Bug soon and releases a new FW rev.

regards
Calvin
I haven't had the time to get a unit out and check this behavior...........been away.
It needs writing up properly with examples or a vid before it's sent to the factory with a high priority for fixing.

If anyone wants to also do this it will add weight to the recommendation and assist in getting a faster fix.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: BillB on June 10, 2019, 01:56:04 pm
I don't have a SDM3065X but did test out the SDM3045X.  The device will reset data only if the data option is set to recent rather than all.  This option is the second soft button when in trend display mode.  When all is selected, I have no problems with accumulating data when I switch display modes.  I also tested this using EasyDMM and it behaved the same.

A couple of caveats that I didn't care for with EasyDMM:
1)  It will reset the device accumulation upon connection in Data Mode.
2)  It seemed to cause my device to hang at one point, where I had to cycle power.  (I could not get the device back to local control mode)
3)  At one point a few minutes in, the EasyDMM reported a query error and was not updating.  The device was still collecting, but upon reconnecting with EasyDMM, it reset the data.  Not cool.
4) EasyDMM (and maybe other EasyTools) are saving files with a .bmp extension, but are really .jpgs.  This causes problems with this forum, as it thinks the files are corrupt!

The first image shows where the option is.  The second shows the data accumulating across the display mode change.  The other images show Trend in EasyDMM and data and graphing using the app as well.

*** UPDATE ***
Testing the SDM3055 as well (both having latest firmware), it behaves just like the SDM3045X.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on June 10, 2019, 04:52:41 pm
Hi,

Quote
The device will reset data only if the data option is set to recent rather than all.
With the SDM3065X the data is reset always when You change the display mode, regardless if set to Recent or All.
Even if You press the Diplay/Trend button out of the trend chart display and then return to trend chart all data is lost and restarts logging new data.  :wtf:
So it doesn´t even keep the data when you remain in the same display  :palm:
Haven´t checked with easy-DMM so far.

regards
Calvin
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: BillB on June 10, 2019, 09:38:01 pm
Even if You press the Diplay/Trend button out of the trend chart display and then return to trend chart all data is lost and restarts logging new data.  :wtf:
So it doesn´t even keep the data when you remain in the same display  :palm:
Haven´t checked with easy-DMM so far.

Well, as I updated below, the SDM3045X and SDM3055 keep the data just fine.  So, it seems like Siglent should be able to easily fix this for the SDM3065X.  The other meters do not lose the data even when going back to Number display mode, in fact, I almost feel like I'd like to know it is still trending all data.  But, there is no indicator anywhere that it is still capturing all the data.

While waiting for the fix, EasyDMM could allow you to change views on its local set of data as its capturing, but I hesitate to recommend it given how it handles query errors and interruptions in connection.
 
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on June 12, 2019, 05:45:20 am
Hi,

yesterday I contacted the european Siglent support and reported the problem.
I already got an answer, that they forwarded the note to the SDM project manager to look into that problem.
Now that was fast and just what one could wish for.  :-+

regards
Calvin
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: tautech on June 12, 2019, 05:55:11 am
Hi,

yesterday I contacted the european Siglent support and reported the problem.
I already got an answer, that they forwarded the note to the SDM project manager to look into that problem.
Now that was fast and just what one could wish for.  :-+

regards
Calvin
:)
Reported 2 days earlier.  :P
Reply:
Quote
if it is the bug, we will fix it
.
Added BillB's findings yesterday.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on June 12, 2019, 04:46:08 pm
good
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on July 30, 2019, 08:22:55 am
Hi,

got no news so far .... does anybody else?

regards
Calvin
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: tautech on July 30, 2019, 04:51:00 pm
Hi,

got no news so far .... does anybody else?

regards
Calvin
No news yet and I check here everyday:
http://www.siglent.com/ENs/gjjrj.aspx?id=15 (http://www.siglent.com/ENs/gjjrj.aspx?id=15)
If I remember to post it in this thread I will otherwise it will be announced in the SDM3045X, 3055 and 3065X thread.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: bson on July 31, 2019, 07:38:26 pm
I'm pretty sure they eat their own dog food - meaning they use their own instruments internally.  But maybe what's missing is a culture of engineers simply submitting a ticket when they see something that looks or works wrong or could be improved on.  Their engineers after all are no different from any other engineers, and it's always better to discover subtle problems or glitches internally than having your customers find them for you.  As usual there's only so much QA/QE can do, and advanced products need whole another level of vetting to become reliable and usable.
Title: Re: Siglent SDM3065X - Loss of logged data when switching to another display/chart?
Post by: Calvin on September 14, 2019, 05:26:49 am
Hi,

got a note from Siglent.eu support, stating that "The issue has been reproduced, reported to R&D and will be fixed with the next FW. But I didn’t get a release date for the firmware."
Now let´s just hope that the next FW update is soon to b released.

regards
Calvin