Hi!
Sorry my first post on here is so dumb. In the course of trying to debug a microcontroller project (literally just toggling an STM32 pin high and low), I noticed that my pulse waves are not as sharp as I expected; they're quite slewed looking when you zoom in on them. To try to investigate further, I hooked the probe up to the calibration terminal and noticed the same thing (see attached pic). Since I'm still pretty new to electronics I'm hoping you guys can tell me whether this is normal, whether I have something wrong in my setup, or whether I might have a problem with the scope and/or probe.
Any thoughts/ideas?
May I ask how you have get SDS1104X-E with older than FW 6.1.25R2 what was first version where attenuation was displayed in channel info. So your FW version must be 6.1.25R1 or older.
Is it possible you tell what is your version, least I'm interested where from can get so old version and how old it is. Unbelievable.
Please update it to latest what is today 6.1.33
And follow instructions (pdf) what are inside update ZIP exactly.
And seriously: USB need be native FAT32 format!
Test USB stick with scope before use it for FW update.
Remember after FW and immediately after scope have been
continuously on over 30 min you need do Selfcal before use (and nothing connected to any inputs of course). Because this update change cal related things lot of.
After then you have LOT of improved scope. Totally many things changed.
For me your signal looks like 1kHz calibration square is connected to CH1 using Probe 1x.
(Probe 1x/10x switch. Scope 1x 10x or what ever do not affect freq response at all but probe using 1x is around 5 -7 MHz and 10x something more.
I think your probes what come with scope are PP510 what are "100MHz" (in real world well over) and this is only when probe is set for 10x (and then also scope for 10x so that level numbers are ok)) If any problem you can also PM (finnish if it help)
ETA:
Sorry I forget SDS1104X-E
probe cal output risetime is very slow, reason was not probe possible 1x. (Yes it is ok for
probe LF response adjustment purpose for what it is made)
Difference between 1x and 10x is quite small because output risetime is slow.
Here images what are normal with PP215 probe and SDS1104X-E Probe cal output.
With probe set 1X and 10X. (with PP510 probe situation is quite same in this case.)
Of course risetime may have small differences between scopes because it is not at all critical for purpose it is made (probe LF cal)

Probe 10X Probe Cal output

Probe 1X Probe Cal output