Siglent SDS-1202X-e waveform issues

<< < (3/8) > >>

JiaRen:
I see, so I can just auto setup, set the time division then take a screenshot for you?

JiaRen:
It’s set on 1x, initially thought it’s the probes fault. I guess it’s user error now.

Kean:

--- Quote from: JiaRen on December 03, 2022, 11:20:01 am ---It’s set on 1x, initially thought it’s the probes fault. I guess it’s user error now.

--- End quote ---

Just make sure the channel is set the same as the switch on the probe, otherwise voltages won't be correct.

Auto setup should still be able to show the waveform, and you shouldn't have to adjust voltage or timebase to see the signal.

If you have another probe, you might also want to try swapping them to see if that makes any difference.

JiaRen:
I dont know what is going on, when I pressed autosetup, nothing will come up. Only when I start touching the voltage division and the time division, the wave would only appear. So when I replicate your settings, nothing appears.

JiaRen:
Both probes same thing, auto setup wont show the waveform until i touch the voltage and time division. the 1x and 10x are set correctly.