Author Topic: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?  (Read 935 times)

0 Members and 1 Guest are viewing this topic.

Offline BionicboneTopic starter

  • Contributor
  • Posts: 21
  • Country: gb
I am just getting used to my new Siglent SDS1104X-E so forgive me if this is a stupid question   :-DD

I have an ESP32 & a Teensy 4.0 powered from the same 3.3 volts bench supply, each is set to output 1kHz pulse, I simply wanted to see the different difference in the pulse and practice with the oscilloscope without keep pressing the default button and auto setup.

I notice that as each pulse changes state there is an interaction seen on the other pulse which is in a stable state.

My question is, is this some strange behaviour with the oscilloscope or is it actually the circuit in some way that I should be aware of ?
« Last Edit: March 24, 2023, 07:12:44 pm by Bionicbone »
 

Offline BionicboneTopic starter

  • Contributor
  • Posts: 21
  • Country: gb
Re: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?
« Reply #1 on: March 24, 2023, 07:30:22 pm »
Intestingly

1) If I only connect one probe groun lead its not as bad but as far as I am aware you should always ground each probe.

2) If I use the internal 1Khz as the other signal that is uneffected thus suggests to me it is the circuit ?
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28379
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?
« Reply #2 on: March 24, 2023, 07:39:12 pm »
There’s nothing strange about channel crosstalk, it can be a real thing just as grounding effects can be too.
For signals with the same reference only one ground clip is required.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6638
  • Country: hr
Re: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?
« Reply #3 on: March 24, 2023, 07:48:37 pm »
Look up the term ground bounce...
 
The following users thanked this post: Bionicbone

Offline BionicboneTopic starter

  • Contributor
  • Posts: 21
  • Country: gb
Re: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?
« Reply #4 on: March 25, 2023, 01:06:11 pm »
For signals with the same reference only one ground clip is required.

Forgive me, can I drill into this a little, two pictures here.

1) Yellow probe in use and grounded.
2) Yellow probe in use, pink probe grounded.

Notice there is significantly less ground bounce. Rise, fall times are recorded as much quicker, and peak to peak voltage lower (probably due to the ground bounce). To be fair it is like there are two different signals.

Why, what do I not understand ?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4105
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDS1104X-E, Dual IC, Dual Pulse with strange interaction ?
« Reply #5 on: March 26, 2023, 05:06:42 pm »
You need understand that every single piece of wire is inductor. Also inside your breadboard these GND and VCC routes are inductors (and also small capacitors).
Also this oscilloscope GND wire you connect to board GND. There wires are inductors.
As long as you are working with DC or low frequencies near DC and audio range and so on, these do dot matter anything or nearly nothing in practice.
Also even bit higher frequencies these do not matter so much when signals are somehow sinusoidal type.
But when we take higher frequencies these start matter more and more. Your square wave can be very low frequency, example 1kHz etc.. but there is these rising and falling edges. So there is also higher frequency components. These frequencies can very easy go to tens of MHz or more and even much more.

If  you want thinking exercise. Draw our circuit to paper but do not draw wire and single wire but draw these as inductors and after then start thinking what happen there. And also not only these inductances but also there is capacitances. Probe tip may have example 15pF between tip and GND point.

Rough  example this oscilloscope probe GND wire inductance can be 200nH
I can see there in your breadboard circuit some wires. You can roughly estimate that every 10mm of these wires may have 8nH inductance.

Now try estimate what is whole GND route inductance from GND jag to Teensy GND can estimate that it somewhere around 300nH

Example in your first message image there is CH3 trace. Rising edge ends to around very roughly ~60MHz oscillating. (overshoot)


And also note. These things you have now seen and displayed here. These all are normal. You have done this "magpies nest" circuit + probing.
 
And also note. These have (in this case) nothing to with oscilloscope internal cross talk etc things. These all happen just externally, and thank your oscilloscope you can see what happens when probing have done wrong way.

When I look this same image 00002.png  it looks like risetime is around 4ns +/-1ns
And when look your circuit and probe connection and specially with probe own long GND wire and also connected FAR away from pulse output GND and OUT... there is really nothing to wonder.

This single _probe_GND wire is around 200nH. Take your multimetes and tmeasure.  Yes your meter tell it is 0 ohm.
But its reactance with example 60MHz is  75ohm! With 100MHz its reactance is 125 ohm. Not 0 ohm. Think about it. Every wire you have there, they are inductors. They have reactance. But then there is also capacitances - everywere. These all start effect when you have these signal rising and falling edges where you can easy find over 50MHz frequency components even if your square wave is 1kHz or even 1Hz or less. Only what matter is frequency components due to these "fast" edges.

These kind of signals need look using short Signal and GND wires. Probe tip (not this hook) directly to output pin and probe GND directly to nearest possible GND point using GND spring or other contact there on signal source board. Many oscilloscope probes have small spring for do this contact (but example PB510 do not have).


Example your probe PB510 may have roughly 22pF input capacitance (when 10x)
If it is 22pF and GND wire inductance is 200nH and IF there is nothing else,  its resonance is around 75MHz. And in your circuit there is more inductance via whole GND route related to signal output point.

And in these cases both probes GND need of course connect. Because you are now working with tens of MHz RF signals not 1kHz or some audio. (due to signal edges)
You must NOT connect these using these long long GND wires if you want avoid these "weird" things you now wonder. These are natural in your circuit when you use probes like you have done. You are not working here with DC where short wire is just near 0 ohm.
It is sad that Siglent PB510 do not have these GND springs. But these are easy to do.

This is how  you need use probe GND for better result.
https://www.eevblog.com/forum/testgear/needle-tip-oscilloscope-probes/msg798352/#msg798352



I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: nikitasius, Bionicbone