Products > Test Equipment
Siglent SDS1104X-E In-Depth Review
tautech:
25R1 XY teaser.
https://youtu.be/wUCdaVq3kL0
Performa01:
--- Quote from: toli on May 22, 2018, 06:29:06 am ---thank you.
any idea about the resolution of the display in this mode?
--- End quote ---
Resolution is the same as the scope. The web server does not do anything different than what the scope does when operated and viewed locally.
Yet depending on screen size and resolution, the DSO screen usually appears enlarged on the PC (I'm using 27" @ 1920 x 1080).
Performa01:
Someone was interested in getting a TESTEC TT-HV250 probe and after reading my reply #62
https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-in-depth-review/msg1435196/#msg1435196
could not believe that there actually was no means for a proper ground connection at the probe-tip end.
The following photos are intended to show what I have. First with the condom fitted :)
TESTEC TT-HV250 01
And without:
TESTEC TT-HV250 02
It should be immediately obvious, that neither the supplied probe BNC adaptor nor a ground spring (not supplied) could make a contact here- The white part is non conductive and too small in diameter on top of that.
I’ve purchased this probe a long time (>5 years) ago, so maybe the probe design has changed since then.
It might also well be that I just got a lemon.
2N3055:
I have same probe but sold as Hameg HZ53..
You can pull the outer black plastic tube all the way... This is how it should look..
Regards,
Sinisa
Performa01:
--- Quote from: 2N3055 on May 25, 2018, 03:46:44 pm ---I have same probe but sold as Hameg HZ53..
You can pull the outer black plastic tube all the way... This is how it should look..
Regards,
Sinisa
--- End quote ---
Thanks for that. Yes, that makes more sense. That metal sleeve is definitely missing on my probe, as can be seen in my picture. I've already suspected something like that...
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version