New KP184 Modbus Software Version 1.0.84.0:
Setup KP184 Software V1.0.84.0 32-bitSetup KP184 Software V1.0.84.0 64-bitUser_ManualChanges:
1. New optional anti voltage dip strategy on battery capacity test. Tries to not interrupt the discharge test and restart electronic load when required where bad electric connections are at play. Used in an attempt to save a very long discharge test from aborting. Always check your electric connections before starting any test.
The KP184 unit may always cut the test short (turning the load OFF) if the voltage dips below the programmed test end voltage setting.
It is a good idea to use solid bolted connections on soldered ring connectors at both ends of the test power cables.
Avoid alligator clips of any kind at both test power cable ends (battery and KP184 binding posts).
Also avoid the usage of 4mm banana plugs at the KP184 binding posts. The 4mm banana jack hole on KP184 is a bit loose and do not provide the required good electric connection.
Bad electric connections will result in short voltage dips and voltage spikes caused by inductive flyback currents. This will mess up your data, test plot graph and cause a premature test ending leading to wasted time.
The optional
anti voltage dip strategy measures and identifies fast voltage dips and tries to avoid a premature test end. If the electronic load turns the load OFF the software tries to restart the load ON.
The voltage dips are identified when the voltage drops below the programmed test end voltage with a slope below -70°. There is a maximum of 3 consecutive attempts to restart the electronic load, there is no limit for the non consecutive load restart attempts. This option can be enabled and disabled during the test progress.