HP8590 series reference in/out
« on: March 07, 2020, 06:38:43 pm »
I have a question about the reference in/out of the HP8590 series spectrum analysers. Can I split the 10MHz using a simple T adapter? I want to use it for an external tracking generator. By my measurements, the input is not 50 Ohm anyway and the frequency is low enough. What do you think?
Fear does not stop death, it stops life.
 

Offline ps

  • Contributor
  • Posts: 44
  • Country: de
Re: HP8590 series reference in/out
« Reply #1 on: March 08, 2020, 10:40:37 am »
In the counterlock assemby (A25) the reference signal is AC coupled (47 nF) and clamped to 0.6 V before driving the OSCIN input of a MC145146 PLL. Reference Output is driven by a NE5212 RF OpAmp, also AC coupled (10 nF). So I assume there will be enough power available to drive a second input, and there is definitely no 50 Ohm matching present.
 

Offline MitiTopic starter

  • Super Contributor
  • ***
  • Posts: 1324
  • Country: ca
Re: HP8590 series reference in/out
« Reply #2 on: March 09, 2020, 12:32:05 am »
In the counterlock assemby (A25) the reference signal is AC coupled (47 nF) and clamped to 0.6 V before driving the OSCIN input of a MC145146 PLL. Reference Output is driven by a NE5212 RF OpAmp, also AC coupled (10 nF). So I assume there will be enough power available to drive a second input, and there is definitely no 50 Ohm matching present.

Good info, thanks ps! I will use it as reference to an AD4350.