delvo:
How do I enable the Trigger Out? According to the Manual it should be in Utility -> IO Set -> Trigger Out. The Trigger Out option is not in it.
Martin72:
Hi,
Which version is your manual ?
Page ?
In the current manual (EN05B, 08/2021) there´s nothing to find.
My guess is, the output is trigger out as long as you don´t activate pass/fail mode.
delvo:
I had an old version of it and it looks like its just always outputting. Thanks
Martin72:
No problem, next step would be to test if I´m right, but this I can do later.
Another thing:
If it´s so, then it´s not mentioned in the manual or I couldn´t find it.
1audio:
I'm having an issue with the probe shield resistance. I have an 1104-XE that came with 4 probes.They have been working fine for most tasks however I'm looking at low level signals now and noticed a lot of noise that change when I move probe cables. I chaced it down to the shields on the probes. They vary fom around 4 Ohms from clip to BNC to over 20 Ohms. When looking at low level signals the shield can become an antenna if its not well grounded. Do other users have this problem? I tried some other probes in my collection. The miniature HP probes do'n seem to connect on the center pin. Te lare hP and Tek probes I have are lower DCR shield (1 Ohm) and seem to be better.
Do I cave some defective probes? Who is making good low level probes today? My other probes are all 20+ years old.