The second test (1:10):
a) 27 year old probe - 10 ns;
b) PP6100 100MHz Chinese probe - 10 ns;
c) PP215 200Mhz - 22 ns;
The result still speaks for itself and PP215 will be the last probe I choose from those 3 probes. And no, I did not say that PP215 unusable, but I am saying that in this particular test PP6100 100MHz probe outperformed it.
You are right that the PP215 are certainly not the best probes money can buy; just look at the input capacitance, which is specified 16–20pF. Compare that to the Keysight N2889A specifications, where it is just 11pF – at 5 times the price, that is. If Siglent is going to ship the DSO with probes of that quality level and the price of the 4-channel model increases by $400,- because of that, most folks will be very unhappy and in fact, essentially no one would really appreciate that. The bells and whistles is what sells, not inconspicuous quality probes. The original probes are cheap and will serve most folks well and the ones who really need or want something better can always invest some serious money to buy whatever they want whenever they want.
It’s not that easy though and price is not the ultimate performance indicator, as the probe should be a good match for the rather complex input impedance of the scope. For instance, I have tested a bunch of probes on the Siglent SDS2304X, including a Keysight N2843A 500MHz probe, which happened to perform very similar to the old bulky (nameless!) 300MHz probes that shipped with the SDS2000 (without X), whereas I clearly prefer the characteristics of the Siglent SP2030A (shipped with the SDS2304X) over the Keysight – for that particular scope, that is. After all, the SP2030A extends the scope ±3dB bandwidth up to 450MHz! So one should be careful and keep in mind that higher price and/or more specified bandwidth does not automatically mean better performance … but more prestige at the most

Back on topic - PP215. The 200MHz Siglent scopes have a specified (calculated!) rise time of 1.8ns and since the actual bandwidth is more like 240MHz, the real rise time is 1.4~1.5ns (measured!) accordingly. The rise time of the probes has to be added, but is not specified for the PP215. Since these are only 200MHz rated, we have to assume their rise time to be also some 1.8ns, which already means a total of about 2.4ns rise time. Then the signal source sure has some rise time too, which needs to be added as well (of course this is always the same so it could be ignored when just comparing probes).
What’s more important is probe input capacitance and source impedance. There we could get some unlucky combination that makes the probe look worse than it normally is – but by no means in the realm of double-digit nanoseconds when probing low impedance circuit nodes!
From previous tests I know that the PP215 does not limit the frequency response of a 200MHz frontend as in the SDS1202X-E. It performs pretty much the same as a direct coax connection and we get >240MHz bandwidth. From that measurement, its rise time has to be better than 1.5ns.
Just to be sure, I did a quick test right now to verify the performance of the PP215. Signal source for all following tests is a pulse generator with precisely 1ns rise time, repetition rate 1kHz, pulse width 50ns, amplitude 600mVpp, with a 50ohm through termination directly fitted to its output. Probe is connected via its BNC adapter accessory, ground lead is disconnected of course.
First the PP215 on the SDS1202X-E it was shipped with:

PP215_x10_SDS1202X-E_1ns_Risetime_T
We get 2ns rise time and doing the math, knowing the signal source rise time to be 1ns and assuming the SDS1202X-E has 1.4ns, then there is just 1ns left for the PP215. Not bad at all…
Now let’s verify this on a faster scope, the SDS2304X, same test setup otherwise:

PP215_x10_SDS2304X_1ns_Risetime_T
Doing the math again. We measured 1.75ns total rise time (toggling between 1.7 and 1.8 ), scope and signal source both have 1ns and there’s once again just a round single nanosecond left for the PP215.
Finally I wanted to know what does it look like with the (rather nice) 300MHz SP2030A probe, shipping with the SDS2304X?

SP2030A_x10_SDS2304X_1ns_Risetime_T
It looks definitely different with less bad ringing. Now we get 1.6ns total rise time, minus 1ns for scope and signal source each, there is only 750ps left for the SP2030A. Not a significant difference on paper, but in practice, as the screenshots are illustrating.
In general, dealing with transition times close to 1ns, matters start getting a little tricky