Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1400 on: July 15, 2019, 02:54:44 am »
The inability to make use of “special characters” or blanks hardly qualifies as forcing a “ridiculously restrictive and insecure psk or ssid.”  Most large institutions I’ve encountered avoid using them altogether to avoid the inevitable confusion with visitors or dealing with other non-compliant gear.

63 chars randomly chosen from a set of 36 makes for a pretty sparse possibility matrix.

It's plenty restrictive when the institution's WiFi that you are trying to connect to literally requires special characters in all passwords, and the more serious issue is the lack of spaces allowed in the SSID, which many SSIDs that I have encountered on an institutional basis possess.

Finally, why make something less secure in purpose when it is simple to make it more secure? The argument against it is ridiculous on its face.
« Last Edit: July 15, 2019, 05:33:09 am by SMB784 »
 

Offline tinhead

  • Super Contributor
  • ***
  • Posts: 1918
  • Country: 00
    • If you like my hacks, send me a donation
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1401 on: July 15, 2019, 07:05:18 pm »
right, the best is to use only spaces and two special characters at begin and end.

https://blog.stevepaulo.com/your-password-complexity-requirements-suck-7934c4e4b295
I don't want to be human! I want to see gamma rays, I want to hear X-rays, and I want to smell dark matter ...
I want to reach out with something other than these prehensile paws and feel the solar wind of a supernova flowing over me.
 

Offline ADJohnson

  • Contributor
  • Posts: 17
  • Country: gb
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1402 on: July 17, 2019, 10:37:15 pm »
Siglent America have FW 6.1.33 to download for a SDS1104X-E from May this year, but Siglent EU still has 6.1.26 from September last year.  Is there a reason for this or could I just go ahead and update to 6.1.33 - I’m in the UK with the scope supplied from within the EU?

Thanks,

Andrew
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28136
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1403 on: July 18, 2019, 12:56:14 am »
Siglent America have FW 6.1.33 to download for a SDS1104X-E from May this year, but Siglent EU still has 6.1.26 from September last year.  Is there a reason for this or could I just go ahead and update to 6.1.33 - I’m in the UK with the scope supplied from within the EU?

Thanks,

Andrew
Welcome to the forum.

Dunno what website you've checked but maybe you were looking at the SDS1202X-E 6.1.26 firmware and not firmware for the 4ch X-E models.  :-//
6.1.33 firmware is there, currently it's the 3rd one down in this list:
https://www.siglenteu.com/service-and-support/firmware-software/digital-oscilloscopes/#sds1000x-e-series

Note; the Release notes link only applies to the version atop of the list however Release notes applicable to the version you download are in the zipped package.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline rf-loopTopic starter

  • Super Contributor
  • ***
  • Posts: 4060
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1404 on: July 18, 2019, 04:17:35 am »
Siglent America have FW 6.1.33 to download for a SDS1104X-E from May this year, but Siglent EU still has 6.1.26 from September last year.  Is there a reason for this or could I just go ahead and update to 6.1.33 - I’m in the UK with the scope supplied from within the EU?

Thanks,

Andrew

If you still have SDS1104X-E with old FW 6.1.26  new  FW  6.1.33. have lot of improvements.
Remember read update instructions and use USB flash(of course native FAT32) what you really know working with your scope  and finally do (mandatory) self calibration after continuous power on.

This is official Siglent European area

https://www.siglenteu.com/




Siglent.EU (https://www.siglent.eu) is not at all Siglent. Just one individual distributor who try "looks like" real Siglent.




« Last Edit: July 18, 2019, 04:43:52 am by rf-loop »
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 

Offline ADJohnson

  • Contributor
  • Posts: 17
  • Country: gb
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1405 on: July 18, 2019, 08:29:29 am »
Ah! That’s where I was going wrong - Siglent.eu.  It was/is the one that pops up top of search results and it looks real enough.  Feel like an idiot now :)

Thanks guys.
 

Offline starlight_tools

  • Contributor
  • Posts: 16
  • Country: ca
    • Starlight Tools
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1406 on: July 19, 2019, 05:24:23 am »
[ Specified attachment is not available ]
Initial reply from Siglent indicates 2 units ISFE wouldn't fit on SDS1104X-E however some BNC C-C measurements indicates they will...........just !  :phew:

As plurn has guessed one unit needs to be flipped, 1/ to fit the BNC's and 2/ to expose the USB A socket so to get power to the ISFE.
In an old thread about ISFE we examined their basic performance which is by no means can be called startling however now we have input attenuation to roughly match the ~200x of ISFE reasonable indicative measurements can now be taken and even further improved upon using the custom input attenuation settings X-E DSO's have.

This thread could use some updates however it's a broad look at the capabilities of ISFE:
https://www.eevblog.com/forum/testgear/siglent-isfe-scope-front-end-isolator-any-opinions/


Back to the issue at hand.......well I believe 2 units of ISFE will indeed fit the SDS1104X-E but as I don't have 2 units we need to rely on measurements and here I've marked the center of ch2 and ch3 in pencil and as this photo shows there's basically zero clearance between 2 units.



So a little ingenuity may be required to power both units and still have one USB A socket available for screenshots but I guess any USB phone charger should be adequate if its power is clean enough.

A screenshot
20V p-p 1000Hz from SDG1032X
BNC cable






This confirms that two ISFE's will connect and operate on the SDS1104X-E scope with minimal clearance between the two units, although as stated one is upside downs so care has to be taken not to let the electrons fall out. 

Seriously though, I am a bit concerned about the weight of them hanging off the BNC connectors of the scope and if they get pulled in any way it could end up damaging the scope's front end.  I figure that I will try to find a set of BNC to BNC connector cables which will get the cantilevered weight off.  Especially as they will be used in a shop environment, which is more like field work than bench testing work.

Thanks for all your help gang!



« Last Edit: July 19, 2019, 05:28:25 am by starlight_tools »
Walter Townsend, TTDr
 
The following users thanked this post: iXod, tautech

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28136
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1407 on: July 19, 2019, 08:57:13 am »
Seriously though, I am a bit concerned about the weight of them hanging off the BNC connectors of the scope and if they get pulled in any way it could end up damaging the scope's front end.
Thanks Walter for coming back to us with 4 isolated channels, very cool. Hope they fulfill your needs.

While I understand your concerns for the front end let me explain and show you a little of its construction.
The 4 BNC's pass through part of the steel chassis of which they are fastened to with nuts and the BNC bases are 4 post varieties that are mounted vertically onto the mainboard.
This makes for a fairly rugged construction where damage is more likely to occur to the BNC shell itself rather than its connection or to the mainboard.
Below in Daves teardown from the rear of the mainboard you can see the the 4 posts and center conductor of each BNC and to the far right is the Pass/Fail BNC that's of the same type and fastens onto the rear metal case.

https://youtu.be/Kay4Jk2DHuE?t=313

So yes, maybe for piece of mind short BNC cables will alleviate the risk of damage in a shop environment.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline starlight_tools

  • Contributor
  • Posts: 16
  • Country: ca
    • Starlight Tools
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1408 on: July 19, 2019, 04:47:34 pm »

Below in Daves teardown from the rear of the mainboard you can see the the 4 posts and center conductor of each BNC and to the far right is the Pass/Fail BNC that's of the same type and fastens onto the rear metal case.

https://youtu.be/Kay4Jk2DHuE?t=313

So yes, maybe for piece of mind short BNC cables will alleviate the risk of damage in a shop environment.
Thanks Rob, I had watched the teardown and had seen that, but between Murphy's law and the laws of physics, which say that if you hang something out in the air, it will get damaged!  Like in machining, you never clamp over air, or building a structure, you never have a point load that is not supported by a foundation under it.
 
Walter Townsend, TTDr
 

Offline MikeLud

  • Regular Contributor
  • *
  • Posts: 218
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1409 on: July 20, 2019, 03:59:26 pm »
Is any one having issues with the Web interface on the new FW using a wireless connection. It used to work fine on the old FW. I am able to get it to work intermittently if I set the IP to DHCP but if I set a static IP it does not work at all. On the scope it shows it is connected for both DHCP and static IP. If I use an Ethernet cable cable it works fine. To rule out the WiFi adapter I tested it on a PC and it works fine. Also tried the both USB ports on the scope.

Thanks,
Mike
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1410 on: July 20, 2019, 04:04:52 pm »
Are you using the same ip adress for both wifi and ethernet?
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline MikeLud

  • Regular Contributor
  • *
  • Posts: 218
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1411 on: July 20, 2019, 04:10:18 pm »
For DHCP the IP is whatever gets assigned and using a static IP I tried several different IP's that are not used on my network and are not part of my DHCP pool. Also when trying to use WiFi the Ethernet cable is not connected.

Thanks,
Mike 
« Last Edit: July 20, 2019, 04:12:24 pm by MikeLud »
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1412 on: July 20, 2019, 04:35:00 pm »
Assuming the ip adress is conform the subnet, does a simple ping also not work?
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline MikeLud

  • Regular Contributor
  • *
  • Posts: 218
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1413 on: July 20, 2019, 04:56:38 pm »
HendriXML,

My subnet is set 255.255.255.0 for my network and a simple ping does not work.

Thanks,
Mike
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1414 on: July 20, 2019, 05:18:51 pm »
HendriXML,

My subnet is set 255.255.255.0 for my network and a simple ping does not work.

Thanks,
Mike
Hi,

255.255.255.0 Is the subnet mask, That means xxx.xxx.xxx. must be the same of all the local ip addresses.

Is that the case?
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline MikeLud

  • Regular Contributor
  • *
  • Posts: 218
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1415 on: July 20, 2019, 05:38:55 pm »
HendriXML,

Yes, the subnet for all devices on the network are the same.

Thanks,
Mike
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1416 on: July 20, 2019, 07:08:38 pm »
What is the static IP address and the default router?
 

Offline MikeLud

  • Regular Contributor
  • *
  • Posts: 218
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1417 on: July 20, 2019, 07:27:56 pm »
TK,

Scope Static IP is set to 192.168.1.212 and the router IP is set to 192.168.1.1  I have the DHCP range set in the router to 192.168.1.2 to 192.168.1.199 and use 192.168.1.200+ for static IP's

Thanks,
Mike
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1418 on: July 20, 2019, 07:42:05 pm »
TK,

Scope Static IP is set to 192.168.1.212 and the router IP is set to 192.168.1.1  I have the DHCP range set in the router to 192.168.1.2 to 192.168.1.199 and use 192.168.1.200+ for static IP's

Thanks,
Mike
The scope will have two ip addresses, one for wifi and one for ethernet. Just to be on the safe side I would make them different, whether the ethernet cable is connected or not.
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1419 on: July 20, 2019, 07:50:39 pm »
On my router I can see what devices are connected to WiFi, doesn't the scopes adapter show up there?

Also it is not blocked (or not in the whitelist) by its mac address?
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1420 on: July 20, 2019, 08:14:04 pm »
Maybe even disable the ethernet on the scope (and reboot).
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3212
  • Country: pt
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1421 on: July 20, 2019, 08:34:36 pm »
TK,

Scope Static IP is set to 192.168.1.212 and the router IP is set to 192.168.1.1  I have the DHCP range set in the router to 192.168.1.2 to 192.168.1.199 and use 192.168.1.200+ for static IP's

Thanks,
Mike
The scope will have two ip addresses, one for wifi and one for ethernet. Just to be on the safe side I would make them different, whether the ethernet cable is connected or not.

You shouldn't have both scope's WIFI and ETH addresses in the same network, if both are enabled.
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1422 on: July 20, 2019, 08:41:55 pm »
TK,

Scope Static IP is set to 192.168.1.212 and the router IP is set to 192.168.1.1  I have the DHCP range set in the router to 192.168.1.2 to 192.168.1.199 and use 192.168.1.200+ for static IP's

Thanks,
Mike
The scope will have two ip addresses, one for wifi and one for ethernet. Just to be on the safe side I would make them different, whether the ethernet cable is connected or not.

You shouldn't have both scope's WIFI and ETH addresses in the same network, if both are enabled.
+1
it receives the ping packet but then doesn't know which interface to use to reply... but is there any way to disable one interface?
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3212
  • Country: pt
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1423 on: July 20, 2019, 08:52:00 pm »
+1
it receives the ping packet but then doesn't know which interface to use to reply... but is there any way to disable one interface?

The wifi for sure, just don't insert it. :)

 

Offline HendriXML

  • Super Contributor
  • ***
  • Posts: 1085
  • Country: nl
    • KiCad-BOM-reporter
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1424 on: July 20, 2019, 09:33:18 pm »
TK,

Scope Static IP is set to 192.168.1.212 and the router IP is set to 192.168.1.1  I have the DHCP range set in the router to 192.168.1.2 to 192.168.1.199 and use 192.168.1.200+ for static IP's

Thanks,
Mike
The scope will have two ip addresses, one for wifi and one for ethernet. Just to be on the safe side I would make them different, whether the ethernet cable is connected or not.

You shouldn't have both scope's WIFI and ETH addresses in the same network, if both are enabled.
+1
it receives the ping packet but then doesn't know which interface to use to reply... but is there any way to disable one interface?
Each interface should have it's own ip-address so who would return should not be a problem. How ever I think implementing "services" on multiple interfaces is not trivial. So I can imagine only one (primary) interface get the services behind them. Maybe the first that gets an IP address, hence the difference between dhcp/static. But thats guessing, with a perfect implementation both interfaces could be used.

However I even experienced issues using only ethernet, like a SCPI session that gets broken, when using the web service simultaneously. (Solution: close the web-page)
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
 