Products > Test Equipment
Siglent SDS2000X-E
<< < (8/13) > >>
tautech:
SDS2352X-E -3dB BW check  ~405 MHz   :o

SSG3021X source
N-BNC adapter and Siglent BNC cable.
SDS2kX-E internal 50 ohm termination
Cursors set on p-p @ 200 MHz




---------------------------------------------------------------
SDS2022X-E -3dB BW check = 250 MHz
50 ohm internal termination.
1V p-p @ 200 MHz from HP8654B

tautech:
Defpom has a look at the 350 MHz SDS2352X-E:

https://youtu.be/7zJ8X51l0Go
tautech:
Firmware update for SDS2000X-E models.

Version 1.1.17R6
7.9 MB
https://www.siglentamerica.com/download/8436/

Changelog
1. Fixed the bug: If the oscilloscope is switched off when the CH1 menu is active, it remains without a menu when switching on again.
tautech:
New SDS2022X-E, unpacked and with the new firmware.

520 MHz from HP sig gen. (substantially attenuated amplitude @ some 300 MHz past rated BW)  :o

tautech:
New firmware for SDS2000X-E models.

Version V1.1.19R2
8.3MB
https://int.siglent.com/upload_file/zip/firmware/Oscilloscope/SDS2000X-E_1.1.19R2_EN.zip

Release notes
1. Optimized communication between SDS2000X-E and SLA1016. This revision (and higher) requires the SLA1016
MUST be upgraded to 8.1.16 (or higher) first. Please take a look at ‘Compatibility with SLA1016’ at the end of
the release notes.
2. Optimized Bode Plot: Accuracy and Sensitivity Improvement, New Vari-level Mode for Testing Loop Response, Miscellaneous Function and UI Improvements.
3. Optimized FFT UI: Added Peaks, Markers, and search within FFT.
4. Optimize self-calibration for channels.
5. Optimized accessing USB drive so that when a USB drive is moved to a computer, Windows will not prompt a restoration of the USB drive.
6. Optimize WIFI connection: 63 characters can now be used for a WPA2 PSK key.
7. Added parity options of MARK/SPACE for UART of 9-bit Decode.
8. Optimized behavior of the cursors (x-axis): Added a mode that the cursors remain at the set position on the waveform when changing time base.
9. Added automatic reset to 0 for sample count for statistics after any timebase or voltage range changes.
10. Fixed the bug: The audible beeper activation would affect the voltage waveform displayed for CH2.
11. Fixed the bug of ROV measurement.
12. Fixed the bug: Normal trigger can show more-than-one trigger event on the display at one time.
13. Fixed the bug: Erratic triggering, randomly jumping between the first and the second edge of a signal.
14. Fixed the bug: The acquire time of history is reset to zero if the time reach 1 hour and 11 minutes. The delta time is also wrong if the period is longer than 1 hour and 11 minutes.
15. Fixed The SPI triggering issue: the 16-bit trigger wouldn't work for a gap >360 ns between the two 8-bit packets.
16. Fixed the bug: Save/Recall setup of trigger and scale issue. ( SalesForce ID:P-00199)(SalesForce ID P-00200)
17. Fixed the bug: Reading cursors by SCPI is not accuracy
18. Fixed the bug: In fine adjust mode, some scale such as 302 mV/DIV is in the wrong position.
19. Fixed the bug: LIN decoder doesn't decode frames with zero-length response.
20. Fixed the bug: If the Courser Offset is not a multiple of zoom time base, the cursor will be replaced when zoomed.
21. Fixed the bug: The measurements made with the cursor in Ref. Give wrong values for all the different probes of
X1.
22. Fixed the bug: Measures fairs with GATE give completely random values in signals of low frequency.
Navigation