Products > Test Equipment

Siglent SDS1000(X) / SDS2000(X) or LeCroy WaveAce (Blackfin) hacks?

<< < (9/19) > >>

markus_jlrb:
@KeBeNe,

looks very ipressive, -3.6dB at 460MHz.
But I wonder about the increased voltage
level at low frequencies.
You have terminated the dso input with 50
Ohm properly?
Sorry that I ask. It's a big difference if I compare
with my scope. Did you make other mods?

Btw what is your smallest time base value.
1ns or less?

Markus

KeBeNe:
So, the SMT06 is no longer in ALC mode at 1V, the levels are no longer true. I did not pay attention.

We should agree on measurement levels and settings on the SDS (setup file from the SDS).

For the measurements, I activate the internal 50ohm resistor.
1ns is my smallest time.

Other modifications than the 4 capacitors removed I have not done, maybe there is still a calibration file, that would be an explanation

markus_jlrb:
@KeBeNe

I have modified your pic with some markers.
Now It's look for my more realistic.

Thanks for your effort!

Markus

markus_jlrb:
In addition to this BW picture, it would be of
interest if someone could measure the input
impedance of the scope channels when
terminated (internally) with 50Ohm up to
150% of the BW - let say to 500MHz.
The theoretical BW according to Nyquist
with 2Gs/sec is 1000MHz. Taking the
drop of a aliasing filter 600 to 700 MHz
in BW should be possible.
The AD83700 is specified -3dB@750MHz.
I compare this with the Lecroy models
with 2Gs compatible with the SDS230xX
that are  specified up to 400++MHz.

Markus

 

markus_jlrb:
FYI,

I have made the effort and check the reflection on ch1 to ch4
om my SDS2204X model. Looks not very good according to my
opinion. 10dB reflection from 50 Ohm terminated port/channel
means (>3% voltage err or 10% power error) error in maesurement
- right?

I had check the spawn from 0 to 1000MHz.
At the point of 100mV/div coming from the top (10,5,2,1,0.5,0.2V)
the reflection change due to internal switching of components.
All pictures show 50 Ohm terminated channels beside one that
I measure with 50 Ohm disabled (1Meg).

So I present screenshots of my measuring device as well as swr
bridge and cabling and the dso screenshots.

Waiting for your comments.

Markus

Sorry pic. #7 should be labeled with CH3 and not CH1.

 


Navigation

[0] Message Index

[#] Next page

[*] Previous page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod