The manual needs to be updated anyway.
electronics hobbyist:
--- Quote from: temperance on March 22, 2024, 02:16:46 am ---@ Tautech
The probes are set to x1 after you press the default button and they are recognized as x10 probes after unplugging.
--- End quote ---
There is a temporary solution that can be used. If you frequently use the 10X probe, you can use the "To Default Key" function.
1. Set the probes to 10X
2. Click on Save/Recall
3. Select Save mode
4. Set the type to To Default Key
5. Set Default Type to Current
6. Click on Save
After completing the above settings, clicking the Default button will load the 10X configuration.
KungFuJosh:
Yeah, that's what I did. It's annoying though. It should be fixed properly. The coax cables currently connected aren't 10X.
Grandchuck:
New bug or some kind of anomaly. I noticed a displayed dc offset that should not have been so I performed self calibration and all channels then showed offset (see screen grab with all 4 set to GND). Then, I tried measuring a DC voltage and the error was enormous (300% or more). Since I have had issues with this scope before, I tried the usual and disconnected it from the ac line and then repowered and did another self calibration. All OK now :phew:
This is a fairly serious problem. Have any of you seen anything like this?
Martin72:
Hi,
I've experienced something similar before, but I can't find the post(will search for it).
I also had this problem:
https://www.eevblog.com/forum/testgear/siglent-sds2000x-hd-missing-features-and-bugs/msg4601140/#msg4601140
I was able to get both back in order with self-calibration.
And you could also set something on the scope regarding self-calibration, as I said, I'll have to look again where that was.