Siglent SDS2000X Plus - Bugs / Missing Features / Feature Requests
2N3055:
--- Quote from: BillyO on March 09, 2023, 05:25:20 am ---Has no one noticed the last "anomaly" I mentioned? Perhaps no one cares. :-//
Well, looks like I found another "anomaly". Have a look at the capture attached. Notice the counts on the statistics. For some really odd reason, the count on the fall time is twice the others even though they were all started at teh same time. :palm:
--- End quote ---
Not an anomaly.
Some measurements are performed once per screen, some will be done as many times as possible.
For instance, RMS is once per screen (one triggered acquisition), period or frequency can be 100 or 1000 times if there are 1000 pulses on screen.
Combine this with track on measurements for seeing how it changes with time.
In measurements Statistics menu you can setup how it works.
There are Count Limit and AIM limit params.
AIM limit controls how much maximum measurements will be performed per screen (for measurements that support it).
IF you set it to 1 you will get only one measurement per screen for every measurement
Count limit is max number of running length of data used for statistics.. If you set it to anything else than unlimited, than that number will be max of measurements used for stats, and there will be no multi measurements either.
Play with that a bit to get an idea..
Performa01:
--- Quote from: BillyO on March 09, 2023, 05:25:20 am ---Has no one noticed the last "anomaly" I mentioned? Perhaps no one cares. :-//
Well, looks like I found another "anomaly". Have a look at the capture attached. Notice the counts on the statistics. For some really odd reason, the count on the fall time is twice the others even though they were all started at teh same time. :palm:
--- End quote ---
An "anomaly" ? :palm:
How many rising edges do you see? And how many falling?
Change the timebase and watch the statistics, then you will understand the implications of deep measurements (over the full record length) with regard to speed and accuracy of measurement statistics.
BillyO:
Ahh ..., I see! Said the blind man to the deaf mute.
Okay, will give that a try. :-+
knudch:
Flash upgrade does not work via WEB interface
When I try to upgrade from the WEB interface, the file (*.ADS) is uploaded and a message is given back
"Fail:File type not supported"
(Firefox web browser, Linux Platform)
When I do the upgrade by USB stick with the same file no problem.
SDS2104x plus with "option" to SDS2504x plus.
Firmware pack 1.5.2R2
(did the upgrade from 1.3.9R10, if I re-try the 1.5.2R2 I get this failure)
Not a big issue but strange
rf-loop:
--- Quote from: knudch on March 20, 2023, 08:10:25 pm ---Flash upgrade does not work via WEB interface
When I try to upgrade from the WEB interface, the file (*.ADS) is uploaded and a message is given back
"Fail:File type not supported"
(Firefox web browser, Linux Platform)
When I do the upgrade by USB stick with the same file no problem.
SDS2104x plus with "option" to SDS2504x plus.
Firmware pack 1.5.2R2
(did the upgrade from 1.3.9R10, if I re-try the 1.5.2R2 I get this failure)
Not a big issue but strange
--- End quote ---
Just tested (Windows). Just mouse click..click..click and ok.
I use "isolated" Opera (do not ask), for only instruments and for www other browsers in same computer. But imho this do not matter anything.
SDS2000XP_V1.5.2R2.ADS (SDS2000X Plus FW update file)
MD5 checksum is: D74B28B943463918B4365A89612BC80F
If this match your .ADS file is ok.
When oscilloscope get this file ok, it can not result as: "Fail:File type not supported"
But why you update to V1.5.2R2. Do you have 06-xx hardware version?