But there is the thing about the 500Mhz bandwith upgrade and only for 2 channel….what does it means.
Why it is so amazing difficult to understand... it is also told in datasheet. Also in Reply #121 it is said.
System is build so that in 4 channel models there is 2 ADC chips. One for CH 1 and 2 and one for CH 3 and 4.
ADC works interleaved mode when it handle only one channel it can 2GSa/s.
Now if one have 500MHz upgrade it opens 500MHz BW when ADCs are working interleaved mode. (2GSs/s)
It reduce BW back to 350MHz when scope goes to non-interleaved mode (1Gsa/s). It need also note that ADC work always full speed 2G interleaved or 1G non-interleaved independent of if decimated samplerate is slower.
If one ADC is in non-interleaved mode then also other is forced to non-interleaved mode.
Why it is so. Simple. 1GSa/s is not enough for 500MHz BW. Theoretically 500MHz BW minus very little is possible for unknown signals but only if we have theoretically ideal 500MHz lp brick wall filter and nearly infinite amount of samples for do true Sin(x)/x interpolation... just as example HPs Chris Rehorn explain long time ago.
So 500MHz BW (after upgrade) is only for max 2 channels simultaneously on and so that mode is interleaved, so CH1 or 2 and/or CH3 or 4 and upgraded scope have 500MHz BW other ways 350MHz. When ADCs goes to non-interleaved mode BW is then 350MHz. So there is autoswitch what change BW filter.
Interleaving /non-interleaving explained in datasheet page 9
Upgrade explained in page 9 and explanation page 10 (note 1. small font)