Author Topic: Siglent SDS5034X scope: Voltgaes allowed to apply to it?  (Read 1447 times)
Offline FaringdonTopic starter
« on: January 07, 2023, 08:01:30 am »
We want to use a bit of coaxial cable as a DIY probe to scope a leakage inductance spike which goes up to ~250Vpk.
Can we put the coax into this scope directly (with a BNC connector on it)?
AYK, the coax piece doesnt have the 10MEG resistor in it, its just a piece of plain coax
People usually say +/-300V to a scope...but this is with the 10MEG probe resistor in there?
Manual and service manual doesnt say
https://siglentna.com/product/sds5034x-350-mhz-4-ch/
« Last Edit: January 07, 2023, 08:03:05 am by Faringdon »
'Perfection' is the enemy of 'perfectly satisfactory'
Re: Siglent SDS5034X scope: Voltgaes allowed to apply to it?
« Reply #1 on: January 07, 2023, 08:33:07 am »
P9 of the datasheet does under Max. Input voltage
https://int.siglent.com/u_file/document/SDS5000X_Datasheet_EN02B.pdf
The following users thanked this post: Faringdon
Offline Performa01
  • Country: at
« Reply #2 on: January 07, 2023, 08:57:28 am »
The specifications clearly state  max. input voltage @ 1 MΩ ≤ 400 Vpk(DC + AC), (DC~10 kHz).
So the real question is about the frequency spectrum of that signal. If the rise time of that spike is not faster than 35 µs, you should be totally safe. But even if it is faster, it is unlikely that the input circuitry will suffer damage from a moderate overload lasting only a few microseconds.
The 10 kHz apply to a continuous sine wave and not a spike with low repetition rate.
It all comes down to the total energy pumped into the scope input. This in turn quite obviously depends also on the physical size of the inductor...
Offline 2N3055
  • Country: hr
« Reply #3 on: January 07, 2023, 09:56:55 am »
Scope is grounded instrument.
What are you measuring and how?
Please explain exactly.
If you never measured it, how do you know what spike voltage is?
Safe way of measuring high voltage/ high energy is having safety reserve..
If DUT is floating and can be safely connected to scope then you could use normal passive 100x probe rated for 1000-1500V. There are dozens of them to buy, some very inexpensive.
If DUT is part of something grounded, then you need high voltage differential probe.
SDS5000X is not cheap scope. Damaging it because you try to save money on probing is not cheaper long term...
Dr. Richard W. Hamming
The following users thanked this post: luudee, Faringdon
« Reply #4 on: January 08, 2023, 08:15:06 am »
Thanks, this scope must not be menat for use with SMPS, as only giving the voltage rating for up to 10khz, is totally no good for SMPS work....just thinking of eg flyback drain voltages etc etc etc....
Also, i assume they mean its 400vpk when you use a 1:1 probe?
Presumably if you  use a 10:1 probe then its 4000Vpk?
I just connected this scope via a 10:1 probe to a node which was pulsing up to 200V for some 100ns every 3.3us.....i hope ive not damaged it?
« Reply #5 on: January 08, 2023, 09:10:26 am »
Or BNC connection.
Only if the probe is rated for such which is very unlikely for a 10x probe let alone a 100x probe but 1000x probes are in these voltage ratings.
SDS5000X models all come with fixed 10x self sensing SP3050A probes rated to 400V RMS CAT II.
https://siglentna.com/wp-content/uploads/dlm_uploads/2022/08/Probe_DataSheet-E01B.pdf
« Last Edit: January 08, 2023, 05:09:33 pm by tautech »
Offline rf-loop
  • Country: cn
« Reply #6 on: January 08, 2023, 10:25:36 am »
Why you do these things and have oscilloscope if you write this sentence.
Have you ever looked any probe specifications.
Btw, also after DC there is also voltage derate curve vs frequency.
Some examples about other oscilloscopes.
Do you think no one is good or not meant for work with SMPS circuits.
Also one question. Why connect with just plain coaxial to oscilloscope 1M input. Also I may ask: what is coaxial length and how its capasitance affect DUT.
Please explain real reasons. Or is it just for playing fun without enough experience and knowledge until smoke come come out (do you know all electrics and electronics works using smoke and when it leaks out then it do not work anymore).
Tektronix MSO54B
Maximum input voltage
1 MΩ: 300 VRMS, CAT II
derate
at 20 dB/decade from 4.5 MHz to 45 MHz;
at 14 dB/decade from 45 MHz to 450 MHz; > 450 MHz, 5.5 VRMS
R&S®RTM3004
at 1 MΩ 300 V (RMS), 400 V (Vp),
derates at 20 dB/decade to 5 V (RMS) above 250 kHz
Teledyne LeCroy HDO4000A
Maximum Input Voltage 50 Ω: 5 Vrms, 1 MΩ: 400 V max (DC + Peak AC ≤ 10 KHz)
Teledyne LeCroy Waverunner 9000
Maximum Input Voltage 50 Ω: 5 Vrms ±10 V peak; 1 MΩ: 400 V max. (DC + peak AC < 10 kHz)
Rigol DHO4000
1 MΩ CAT I 300 Vrms, 400 Vpk (DC + Vpeak)
Take example this Rigol and connect its input to example 200MHz 200Vrms. Is it ok because well inside data sheet  (except there read CAT I.  Think if it cover also high frequencies)
Many many oscilloscopes do not have any specs about input max V derating vs frequency. Nothing about frequency and sure there still is limits - are there any oscilloscope user who really do not understand this.  Oscilloscopes are made for educated engineers who need understand least simplest basics.
One EEVblog  thread for you to read...
https://www.eevblog.com/forum/testgear/rigol-ds1202z-e-input-voltage-vs-frequency-derating/
Also noted you talk about short peaks, not continuous wave.
But yes, it is sad oscilloscopes inputs are not better specified and a good defense is not that others do the same.
Normally in this kind of test we use probes, not direct coaxial and even more weighted not without any impedance matching.
« Last Edit: January 08, 2023, 11:21:46 am by rf-loop »
EV of course. Cars with smoke exhaust pipes - go to museum.
The following users thanked this post: Performa01, luudee, Faringdon
Offline rf-loop
  • Country: cn
« Reply #7 on: January 08, 2023, 10:30:05 am »
As you can see  there is one very sad thing. It is named as probe DataSheet. And "data sheet" do not include even voltage derate curves vs frequency.
This is not real data sheet at all in my rating scale. It is more like poor sales catalog - brochure.
« Reply #8 on: January 08, 2023, 05:33:33 pm »
Yes of course but sometimes an overview is sufficient however the probe datasheets are not hard to find except on the HQ and EU site that really needs some improvements.
From the NA site finding SP3050A probe datasheet is not hard:
https://siglentna.com/wp-content/uploads/dlm_uploads/2022/08/SP3150ASP3050A-1.pdf
The following users thanked this post: rf-loop
Offline rf-loop
  • Country: cn
« Reply #9 on: January 08, 2023, 07:30:41 pm »
Yes, it is Instructions Guide what can also read as data sheet!
« Last Edit: January 08, 2023, 07:44:02 pm by rf-loop »
Online Martin72
  • Testfield Technician
« Reply #10 on: January 08, 2023, 09:40:13 pm »
Never.
Even a 100:1 probe didn´t reach that value:
https://www.batronix.com/shop/measurement/probes/Testec-TT-HV-150.html
Reasons are isolation, max voltage across one resistor, lack of airgaps...
We always use a differential probe when measuring on smps, especially on primary switched ones.
And/or when measuring on more than one channel - because of the grounding.
Scope is grounded and every of it´s inputchannel too(normally).
If you want to measure on primary and secondary side at the same time and use normal probes, nevertheless if 10:1 or 100:1 or 1000:1, you short the secondary output to ground then - Not good...