On another SDS800 thread (see #340) Bill was investigating probe & DSO effects.
https://www.eevblog.com/forum/testgear/sds800x-hd-bug-reports-firmware/msg5757649/#msg5757649This got our curiosity and we did a couple quick measurement setups wrt the probe performance an it's "
effect on the DUT".
Setup is crude (not technically correct and calibrated) but good enough to show the effects. An AWG signal generator (SDG6022X) is set to 0dBm and swept from 1MHz to 300MHz. For the AWG 50Ω termination we utilized the input of a Spectrum Analyzer (SSA 3121X +) so we could "see" the probe effects on the SA since it's now the DUT for the DSO probe. A BNC "T" was used to allow the DSO Probe BNC Probe Adapter to be inserted and the various test probes inserted into the adapter.
The SA was set to 1MHz to 300MHz range, peak detect, and +2dBm Reference level with 1dB/div. A reference was captured to show the results without the DSO probe and should show ~0dBm across the 300MHz span, see PNG0.
The SDS800X HD was setup with an FFT of similar parameters, 1MHz -300MHz Span, +2dBm reference and 1dB/div Scale. Other FFT parameters, Flat Top Window and 1us/div TimeBase. CH1 was set to 10X for probes and 100mv/div since 0dBm equates to ~316mv peak (note: don't saturate the DSO Input).
PNG112 shows SDS800 result with PP510 Probe and PNG1 shows SA result.
PNG113 is with SP3020A Probe, SA result is PNG2.
PNG 114 shows Rigol PVP3150 Probe with PNG3 SA result.
Note the "Effect" the probes have on the DUT (Spectrum Analyzer acting as DUT) which is different than what's displayed on the DSO as the back end of the probes and DSO characteristics influence the measurement at this stage as shown by the DSO.
One important point we should mention. Our SA (SSA 3121X +) has a Tracking Generator (TG) output which one might consider utilizing with these measurements as the source instead of the AWG. This particular SA
"Normalizes" the SA result from the TG by taking a
"Reference Sweep" which is used to Compensate for the TG amplitude variations across the frequency sweep. The result shows on the SA
Display as a nice level amplitude across the entire sweep implying a constant level from the TG!! This is
NOT the case here as the TG hasn't been leveled or compensated in any way, just the SA
Display is
Normalized!! One can easily be fooled by this and think the SA TG has a constant level output
Dedicated RF/MW sweep generators usually have a proper leveled output, and suspect more expensive SA Tracking Generators also have a leveled output. Leveling involved actual measurement of the generator output and feeding back the result for corrections, AWGs generally don't do this and just rely on an accurate generation of the waveform without output feedback assuming a
prefect load (either 50Ω or infinite).
Anyway, just thought we should mention this in case some folks might decide to use the SA TG as a source for DSO bandwidth measurements, here small amplitude variations shouldn't matter much since the DSO bandwidth is so highly influenced by the selected probe and the DUT characteristics.
Best and Happy Holidays