Author Topic: Siglent SSA 3021X plus high phase noise  (Read 1940 times)
Offline angelcruizTopic starter
« on: November 24, 2023, 12:19:18 pm »
Hello,
I've got an SSA 3031X which, according to its datasheet and confirmed by several tests that I've found in this blog and somewhere else, should have a phase noise of around 98 dBc/Hz @ 10 KHz offset (specified with 1GHz input signal).
The problem is I can't get better than 85 dBc/Hz as seen in the attached picture.
The tests conditions are:
Unit: SSA 3021X Plus
SW ID: 3.2.2.5.1R1
Source: Siglent SDG 2042X
Input signal: 0 dBm, 10MHz
Span: 50 KHz
RBW: 100 Hz
VBW: 1Hz
15 Averages
Despite the datasheet test conditions being with a 1GHz input signal, several posts in this blog confirm the stated phase noise using input signals of 10 or 20 MHz.
I don't know whether the fact that my unit has been hacked has somehow messed up the internal PLL parameters that shape its phase noise, I have a defective unit or there's something else I'm doing wrong. Still, I've tested with multiple configurations (RBW and VBW of 1Hz, 20, 30 and 100 MHz input signals, etc) and I always measure no less than 85 dBc/Hz phase noise.
Do you have any clue?
Thanks in advance,
Angel.
Offline Orange
Re: Siglent SSA 3021X plus high phase noise
« Reply #1 on: November 24, 2023, 12:55:38 pm »
I just checked on my SSA3032X (non plus)
I get about 100dBc/Hz at 10MHz  @10kHz offset, BW 100Hz
Signal source is a SDG1062 at 10MHz, 0dBm
Mind you that this is not a plus model.
Seems that you have an issue with the SA or maybe the generator is at fault
Offline Orange
« Reply #2 on: November 24, 2023, 01:01:26 pm »
On more thing if you use the SDG2042X with external 10MHz clock it might also increase the noise, the PLL in the generator has a known problem.
Offline tautech
« Reply #4 on: November 24, 2023, 04:11:58 pm »
Sorry, to be specific, the installed FW was SVA1000X_V3.2.2.5.1R1.
I've upgraded to SVA1000X_V3.2.2.6.0R7 and the results are still the same.
I've attached a noise floor test with RBW of 30KHz (50 Ohm termination input) since it seems to be also high for about 5 to 10 dB.
Could it be that the SVA calibrations are bad for an SSA or something like that?
Offline Orange
« Reply #5 on: November 24, 2023, 07:09:53 pm »
From an old SSA3032X, which is properly calibrated within specs.....
Looks almost identical. Compared to the SVA the noise floor of the SSA is about 3.5 dB lower
Offline tautech
« Reply #6 on: November 24, 2023, 10:14:37 pm »
1. OP analyzer settings duplicated best I can see. Source = SDG6022X
2. With BW settings equal.
Try a Preset to factory settings.
Edit to add
Factory fresh SVA1032X
« Last Edit: November 24, 2023, 10:31:27 pm by tautech »
« Reply #7 on: November 24, 2023, 11:33:58 pm »
I think the noise floor is indeed within specs, just a little bit down.
The phase noise is definitely bad, but it gets better going from 10 MHz (-85 dBc/Hz) to 1 GHz (-96 dBc/Hz, attached, ERASynth micro source). The datasheet tests conditions for PN are at 1 GHz but I'd expect it to be as good at lower frequencies.
There's no change after a factory reset or using an external 10 MHz reference either. I think it might be related with the PLL loop filter parameters. Do you know if that could be the case?
« Last Edit: November 24, 2023, 11:39:37 pm by angelcruiz »
Offline tautech
« Reply #8 on: November 25, 2023, 12:44:20 am »
No idea, sorry. Not seen much in the way of discussion about this.
Do consider factoring in the Error and Accuracy spec on P11.
https://int.siglent.com/u_file/download/22_08_23/SSA3000X%20Plus_DataSheet_DS0703P_E04C.pdf
Never looked as to how it compares with a SVA so there's a little exercise for you.
https://int.siglent.com/u_file/download/22_08_23/SVA1000X_DataSheet_DS0701X_E05C.pdf
Then there's the fact you hacked it and saved ~ $ 1800 however you did get a single port VNA for your troubles.
Offline 807
« Reply #9 on: November 27, 2023, 04:11:06 pm »
Hello,
SW ID: 3.2.2.5.1R1
Span: 50 KHz
RBW: 100 Hz
VBW: 1Hz
15 Averages
Thanks in advance,
Angel.
My display at 10MHz looks almost idendical to yours. But I'm using a Marconi 2022E. Don't know what it's phase noise is like.
It's better at 1GHz.
Offline Orange
« Reply #10 on: November 27, 2023, 06:47:09 pm »
Some more pictures from an SSA3032X
Looks better at 10MHz
Offline drew23
« Reply #11 on: November 30, 2023, 01:43:31 am »
Hi All.
I have a PN2060A phase noise analyser and have made a blog post on it, together with phase noise results for the 10 MHz TCXO output of my Siglent SSA3021X and Rigol RSA3015N.
The Siglent is okay at -141 dBc/Hz @ 10 KHz and I think in spec but the Rigol is considerably better at -167 dBc/Hz @ 10 KHz.
In a later, draft post, the phase noise for a Keysight MXA N9020A Vector signal analyser is -163 dBc/Hz @ 10 KHz. This was a very expensive instrument in its day and depended heavily on its TCXO for performance. So the Rigol value is pretty good for a much cheaper instrument.
I am not sure of the method used in the specifications of the Siglent. Presumably, it is the output of the tracking generator at 1 GHz. The PN2060A phase noise analyser can only measure to about 130 MHz at baseband, without a downconverter. I have just moved house and my gear is disorganised. However, I might try measurements at a higher frequency.
Regards Drew VK4ZXI
https://vk4zxi.blogspot.com/2023/07/an-economical-way-to-measure-phase.html
https://vk4zxi.blogspot.com/2023/08/have-pn2060a-phase-noise-analyser-will.html
The following users thanked this post: 807
Offline RoV
« Reply #12 on: December 02, 2023, 10:26:14 pm »
I have both an SSA3021X+ "upgraded" to SVA1032X and an SDG2042X "upgraded" to SDG2122X, so I tried to replicate your configuration.
Additionally, I tested a KVG OCXO, also 10 MHz.
My OCXO is fitted with an output attenuator and gives ~-11.7 dBm on the fundamental, so initially I tried both generators at -11.7 dBm. Same settings as OP, except for input attenuator at 10 dB.
Results were ~identical for the two generators and phase noise at 10 kHz ~99 dBc/Hz
1942980-0 (OCXO)
1942986-1 (yellow OCXO, pink SDG)
By increasing the output level of the SDG at 0 dBm and with SSA input attenuator at 20 dB, results remain consistent (OCXO trace is same as before, taken with 10 dB attenuation)
1942992-2 (yellow OCXO, pink SDG)
« Last Edit: December 02, 2023, 10:28:32 pm by RoV »
Offline 807
« Reply #13 on: December 02, 2023, 11:59:46 pm »
Just measured mine again at 10MHz, but this time using a Marconi 2024 generator which has much better phase noise than the 2022E I used in earlier post.
The close-in noise is better on this generator, but I get similar reading to my previous one at just -86.66dBm/Hz @10kHz.
Is there a problem with mine & angelcruiz's analyzers? Why are we getting such pronounced "hills" compared to others?
edit...I tried the measurements at -10dBm, but results were the same.
btw...mine has been "upgraded", but not to a VNA