Author Topic: troubles with uCurrent GOLD  (Read 1941 times)
Offline TifTopic starter
« on: January 02, 2019, 01:47:33 pm »
I recently acquired a new uCurrent GOLD via welectron. I wanted to use it to do some measurement today, but as it is my first time with this device, I ran some tests before.
The issue I encountered is that when powering on the uCurrent with nothing connected to its inputs, I get 1,7 mV at the output for the mA and uA ranges, and 3,4 mV for the nA range. When I connect a current source at the input, the output almost doesn't change.
I am using a 121GW to measure the output voltage, and tried with an oscilloscope with no improvements.
I checked the battery voltage which is 3,09V.
I visualy inpected the board and I don't see any bad soldering.
May the uCurrent have been damaged during transport? I you have any idea to fix the issue, you are welcome!
Offline SilverSolder
Re: troubles with uCurrent GOLD
« Reply #1 on: January 02, 2019, 03:05:12 pm »
Fresh battery?
« Reply #2 on: January 02, 2019, 03:11:10 pm »
The battery is new (I bought it a few months ago and never used it until now). The voltage of the battery is 3,09 V so I don't think it comes from this...
Offline coromonadalix
« Reply #3 on: January 02, 2019, 03:47:06 pm »
Normally  the current gold is an current to voltage conversion adapter, i dont see any offset adjustment or an zero adjustment
IF you short the inputs, does the output goes to zero ??   and try it with another meter too,  the 121gw is not a perfect meter ???
Specifications – µCurrent™ GOLD
    3 Current ranges:
        +/- 0-1250mA (20µV / mA burden voltage typical) * switch contact resistance plays a role here. 10uV due to the shunt resistor.
        +/- 0-1250µA (10µV / uA burden voltage)
        +/- 0-1250nA (10µV / nA burden voltage)
        NOTE: Higher currents are available by increasing the battery voltage, to a maximum of 5.5V, which would give +/-2250mA/2250µA/2250nA maximum)
    Output Voltage Units:
        1mV/mA
        1mV/µA
        1mV/nA
        NOTE: You simply read your mV multimeter range as if it were mA/µA/nA
« Last Edit: January 02, 2019, 03:52:07 pm by coromonadalix »
The following users thanked this post: 001
« Reply #4 on: January 02, 2019, 03:59:07 pm »
In the original uCurrent article, page 13, it is stated that
With nothing connected you should measure zero on all three current ranges
I agree that there is no circuitry to compensate for the offset, but I didn't think the offset would be that high!
Anyway, it doesn't explain why the output voltage doesn't change when the input current does.
« Reply #5 on: January 02, 2019, 04:00:20 pm »
Unfortunately I don't have any other meter around, but I have an oscilloscope that shows similar results.
« Reply #6 on: January 02, 2019, 04:22:54 pm »
If the ucurrent is under warranty i would not play in it and void anything,  issue an rma ...
You could check U4 pin3 if you have 1/10 th of the same output voltage you have, but were dealing with very low values here, anything can or could interfere
You have 2 successives x 10 amplifications here ...  same check could be done on U1
You could check the virtual ground between the -batt  and R10 (vgnd)  if you have half battery voltage at anytime
J4   ground output is a vitual ground ... for U1 U2 U4 supply lines
Does the switches are physically okay ?
« Last Edit: January 02, 2019, 04:24:38 pm by coromonadalix »
Offline alex-sh
« Reply #7 on: February 16, 2019, 03:51:19 pm »
@Tif
121GW has got a uCurrent built-in, i.e. no need to have a stand alone uCurrent.