SPD3303X-E SCPI Interface issues
SpiderKenny:
Hi
I have an SPD3303X-E and I'm trying to write some code for it.
I have written code for SCPI devices before (such as my Rigol DS2202E) without any trouble. I'm connecting over TCP/IP to port 5025, so I'm using SCPIP RAW not VXI-11
The problem I am finding is that many of the commands listed in the QuickStart manual don't work. I think I may have missed something.
For example, if I send "*IDN?" then I get the correct response: "Siglent Technologies,SPD3303X-E,SPD3XHBC3R0662,1.01.01.02.05,V3.0"
And if I send "*READLL?" I get the correct response too: "4,13.90,0.00,0.00,20.06,0.00,0.00,18.00,1.70,32.00,3.20"
But if I send things like "OUTPut CH1,ON" then the power supply just beeps, and reading the error code with "SYSTem:ERRor?" results in: "ERROR: 3 Command keywords were not recognized"
Is there some other step I need before sending some of these commands?
finos:
Try putty with telnet mode with 5025 port.
Output CH1, ON
Be careful with the spacies and formatting.
Try any other scpi command as a test
H.O:
I don't have one of these but aren't you missing a colon in there.
I would think it should be OUTP:CH1,ON or something like that.
SpiderKenny:
--- Quote from: H.O on June 06, 2019, 04:21:42 pm ---I don't have one of these but aren't you missing a colon in there.
I would think it should be OUTP:CH1,ON or something like that.
--- End quote ---
Thanks!
Well, not according to manual - but that is probably wrong.
I've tried all these combinations, but none are working so far..:
OUTPut CH1,ON
OUTP CH1,ON
OUTPut:CH1,0
CH1:OUTPut ON
Maybe I'll just switch over to using the VXI-11 protocol instead - because I can Wireshark that from my PC and sniff it.
I tried using ptty and nc (netcat) but got the same results.
Here's a picture of the manual page:
tautech:
https://www.siglentamerica.com/operating-tip/spd-programming-tips/