Been considering the SPD3303X to replace my aging EMCO linear supply and also use as a somewhat precision variable voltage source. So interested in things like measured stability, noise levels and such. I didn't see anything in the data sheet that indicated small temperature change (+-10C) performance around room temperature.

Noted during Dave's teardown, what was Siglent thinking by using non-standard plug spacing :palm:  but also on the + side they are using the AD7792 24 bit ADC & PGA and a decent reference voltage source. If these are utilized in the direct control of the output I suspect the stability & temperature performance should be pretty good.

Anyway, I know this is a "Power Supply" but seems it potentially could be good enough to consider use as a moderately accurate and stable variable voltage reference, thus my inquiry.

Best, 
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SPD3303X stability
« Reply #1 on: November 06, 2020, 07:56:02 pm »
Tapped secondaries and a smart fan makes for quiet use. Essentially completely quiet at low loads.

Sold a good few over the years and just that one has had a problem....operator induced !
A customer popped a channel while powering a motor and trying to do regenerative braking without a protection diode .....but he is a young bloke and learnt that lesson: how high can back EMF go...as high as it needs to until something breaks down !  :-DD

Member HendriXML has had some fun with his:
https://www.eevblog.com/forum/testgear/battery-charging-using-a-siglent-sds1104x-and-spd3303x/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: mawyatt

Offline mawyattTopic starter

  • Super Contributor
  • ***
  • Posts: 3267
  • Country: us
Re: SPD3303X stability
« Reply #2 on: November 06, 2020, 08:27:00 pm »
Good info, thanks :-+

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: SPD3303X stability
« Reply #3 on: November 06, 2020, 08:59:07 pm »
Just recently my unit failed to manage CC where my other power supplies did not so I'd recommend getting one to try before you buy.
 
The following users thanked this post: mawyatt

Offline mawyattTopic starter

  • Super Contributor
  • ***
  • Posts: 3267
  • Country: us
Re: SPD3303X stability
« Reply #4 on: November 06, 2020, 09:02:27 pm »
Just recently my unit failed to manage CC where my other power supplies did not so I'd recommend getting one to try before you buy.

Not sure I understand what you are saying here, could you please provide more details?

Thanks,

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: SPD3303X stability
« Reply #5 on: November 06, 2020, 09:35:24 pm »
I connected the output of channel 1 or 2 to a MAX 60W 12V DC load. It can accept 12V/5A or any lower current with no issues. It can also accept up to 24V also up to 60W.

Rather than cut voltage back it would drop to 0V once it hit 3.2A and reset. So back to 12V and 0 A. Ramp to 3.2A then V drops to 0V and resets. None of my other power supplies exhibited the same behaviour.

As far as being a reference... I don't think so. Even using it as the supply for my LTZ1000 or even my DMMCheck was visible in the recordings but I don't know what you're looking for in terms of performance.
« Last Edit: November 06, 2020, 09:39:23 pm by maginnovision »
 

Offline mawyattTopic starter

  • Super Contributor
  • ***
  • Posts: 3267
  • Country: us
Re: SPD3303X stability
« Reply #6 on: November 07, 2020, 02:10:04 am »
Thanks for the explanation on the way it works, agree that is not Constant Current as I would have thought where the voltage is modulated to keep the current constant within a max voltage level.

Wasn't thinking about LTZ1000 or LM399 performance levels, but if your seeing effects using this as a supply for the LTZ1000 that's something to consider, as well as if it's not good enough for just a quick DMM validation where it just needs to maintain a voltage at a fixed level and not drift or have too much noise. I recall some old lab power supplies we had while I was working that could be dialed to levels of 10uv out of 10 volts, were very low noise and stable, so was hoping for something along these performance levels, maybe at the mv level.

The specs don't say much about short or long term drift, or temperature performance, so don't have much to go on. The use of the mentioned 24 bit ADC and a good reference one would think should produce a stable output. 1mv out of 32V is only 15 bits resolution, well within the capability of the ADC.

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SPD3303X stability
« Reply #7 on: November 07, 2020, 04:47:10 am »
I connected the output of channel 1 or 2 to a MAX 60W 12V DC load. It can accept 12V/5A or any lower current with no issues. It can also accept up to 24V also up to 60W.

Rather than cut voltage back it would drop to 0V once it hit 3.2A and reset. So back to 12V and 0 A. Ramp to 3.2A then V drops to 0V and resets. None of my other power supplies exhibited the same behaviour.
Nor should a SPD3303X-E.

SPD3303X-E firmware version please ?

12V halogen headlight lamp with both filaments in series draws 3.1A @ 12V and SPD3303X-E goes into CC with a 13V setting.
FW version  V1.01.01.02.05

Mucked around for ages with a falling trigger trying to discover your drop to 0V and there isn't one !
Even with a single shot trigger set barely below the output voltage captures of any significant voltage foldback where unable to be captured. A play with Normal triggering didn't prove anything either, certainly nothing repeatable or consistant so I'm calling BS on your reset to 0V.

Of the many attempts to capture bad CC behaviour this is the very worst I could manage which is hardly a foldback and die:

Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: mawyatt

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: SPD3303X stability
« Reply #8 on: November 07, 2020, 05:31:18 am »
I'll get you the FW version when I have a chance.

EDIT: FW 1.01.01.02.05. HW 3.0. In about a week when tests are done I can make you a video. Until then the DUT can't be touched.
« Last Edit: November 07, 2020, 06:06:15 am by maginnovision »
 
The following users thanked this post: tautech, mawyatt

Offline mawyattTopic starter

  • Super Contributor
  • ***
  • Posts: 3267
  • Country: us
Re: SPD3303X stability
« Reply #9 on: November 07, 2020, 08:43:19 pm »
If someone has the SPD3303X or -E  PS could I request you setup a test to supply say 10.000VDC and maybe 1.000VDC with no load and monitor the voltage & noise with a good hi res DMM over an hour or so to see how much it drifts and how much output noise unloaded (maybe look with a DSO). If you have the time and want to also load the supply with a very heavy load to increase the internal temperature, then remove the load and repeat the above test to see how the voltage stabilizes as the temperature slowly returns to normal.

Thanks in advance for any help,

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6660
  • Country: hr
Re: SPD3303X stability
« Reply #10 on: November 08, 2020, 12:39:22 am »
I don't have Siglent PSU, but here is Rigol DP831, set to 12V. No load, about 55 mins, temp was around 23C.
Histogram in attachment.
 
The following users thanked this post: mawyatt

Offline doppelgrau

  • Contributor
  • Posts: 37
  • Country: de
Re: SPD3303X stability
« Reply #11 on: November 15, 2020, 04:14:32 am »
Didn't look at the clock, guess was about 20-30 minutes each for the SPD3303X-E. Sigent SDM3055 measured it.
For comparison a few minutes a SPF1168X. Hope that helps.

Edit: Removed the screenshots from the Rigol, that noise seems to have an other source, a test with the power supply disconneted showed similar results.
« Last Edit: November 15, 2020, 10:50:50 am by doppelgrau »
 
The following users thanked this post: mawyatt

Online Mortymore

  • Frequent Contributor
  • **
  • Posts: 444
  • Country: pt
Re: SPD3303X stability
« Reply #12 on: February 07, 2021, 01:24:59 am »
New firmware released: 1.01.01.02.07R2 (Release Date 05 Feb. 2021 )

https://siglentna.com/service-and-support/firmware-software/dc-power-supplies/#spd3303xx-e-series

Revise History
1.Add the OCP function.  -> I would love to know what they mean with this (OCP = Over Current Protection?)
2.Different  font  colors  when  the  power  is  ON and OFF.
3.Some  improvements  about  the  USB’s stability.

The only thing I notice at first glance is that the output values (V, A and W) font colour is grey when the output is OFF, and turns to white when ON
Apart from that, I don't now if it solves any issue, because I have none.

EDIT: How to access to OCP function. Thanks to mrprecision  :-+
« Last Edit: February 07, 2021, 06:14:02 pm by Mortymore »
 

Offline Alectricity

  • Newbie
  • Posts: 4
  • Country: at
Re: SPD3303X stability
« Reply #13 on: August 05, 2021, 07:46:44 pm »
New firmware released: 1.01.01.02.07R2 (Release Date 05 Feb. 2021 )

Just updated - i have the feeling, it introduced a bug:
pressing the "IP/Store" button for long time (while on the "Home screen") => it doesn't open the store page. it just enables the LED for the "Fine" button.
as far as i can remember, it opened the store page before the update. now i have to press 1x "IP/Store" to enter network page and press "IP/store" then for long time to enter store page.

in addition to that: OCP function is cool feature but:
- i can set max current to something that is higher than OCP (does this makes sense?)
- if OCP is enabled, this is not visible on the home screen
- missing documentation of SCPI commands related to OCP

sorry for hijacking this thread - but it was the only one i have found about the new firmware update :)
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SPD3303X stability
« Reply #14 on: August 05, 2021, 08:16:24 pm »
New firmware released: 1.01.01.02.07R2 (Release Date 05 Feb. 2021 )

Just updated - i have the feeling, it introduced a bug:
pressing the "IP/Store" button for long time (while on the "Home screen") => it doesn't open the store page. it just enables the LED for the "Fine" button.
as far as i can remember, it opened the store page before the update. now i have to press 1x "IP/Store" to enter network page and press "IP/store" then for long time to enter store page.
Confirmed and reported to Siglent.
User installed IP is NOT saved and returned at next boot as it was in the earlier firmware.  :palm:
Rolling back to the previous V1.01.01.02.05 version will resolve this until the next FW is available.

No update as yet to when updated firmware might be a available.  :(