We have known this problem for a long time.
The early sales gave a noisy power supply, then if I remember correctly they switched to LiteOn and Lenovo.
I didn't see in the above posts which one was added for those of you who find it noisy.
I use it with a sound technician's 12V power supply, but the LiteOn that comes with it doesn't make more noise for me either.
I tested the LiteOn, I charged a solar powerbank with it, it said 20V and 66W.
I couldn't test it at 15V.
In the OP's first post, I don't see what the channels are set to. 1x 10x
I also don't know if there is something connected to CH1, CH2.
If you have the opportunity to try it with another power supply, perhaps with a car battery, you can find out that everything is fine with the oscilloscope.
If you can't solve it any other way, then return it.
Maybe write to the seller about what can be done in this case.
There are two pictures of the power supply, with 20V 66W.
And the third is just for interest, the ground wire of the probe was oxidized or dirty at the part where it connects to the probe. With a handheld scope, I only measured ~40Vpp with the same probe.
After I cleaned everything was fine.