Spikes and noise on Rigol DHO814
« on: March 12, 2024, 12:56:11 pm »
Hi there,

i bought a Rigol DHO814, which is my first > 250 bucks Oscilloscope. While trying it out on my prototype, i discovered spikes (and noise?) coming from the power supply. It has a frequency of 54-70 Khz. Connecting earth does help, if its not to far away.

I connected a coil on the leads and held it to the cable. Interestingly, the amplitude changed with position on the cable.

I found out, that i have to connect all GND, in order for the spikes to be lower than with the yellow and blue curve on the picture. But it was still about 22mV. With a little bit of aluminium and ferrite, i could lower this to just under 10mV. One turn in the ferrite does a little, two turns does better, but three turns makes it worse.
Now i wonder if i have a bad power supply?

Is there also something i can do / could try, to clean up the signal?
« Last Edit: March 12, 2024, 01:02:29 pm by eTobey »
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #1 on: March 12, 2024, 01:08:53 pm »
What probes were connected to the scope inputs when you captured those screenshots, and what were the probes connected to on the other end? (Including ground connections of rhe probes, if any?)
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #2 on: March 12, 2024, 01:12:55 pm »
I had 2 probes connected, but only one GND of them. Another probe was laying unconnected.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #3 on: March 12, 2024, 01:16:46 pm »
Does the noise go away when you disconnect the probes from the scope , i.e. is it picked up by them rather than within the scope? Does it help if you move the power supply far away from the probes and your prototype under test?
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #4 on: March 12, 2024, 01:38:44 pm »
The noise goes away from the probe that is beeing disconnected, but it then adds up on the other channel!

The noise does not change, if i put the powersupply with the foil near the probe, but if the blue probe has a disconnected GND, then the noise gets from about 25mv to about 20mV, when the powersupply is near... wtf?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #5 on: March 12, 2024, 01:44:24 pm »
Assuming that you use the little ground strap on the probe (rather than the short ground spring), the probe & strap form a loop which can pick up external fields. Depending on how your prototype is built up and wired, it may enlarge that loop cross section.

It is good practice to move the power supply as far away from the scope and the device under test as possible, and to definitely ground the scope (both for EMC and safety reasons). If you do that and start from "no probes connected at all" to "probe connected and grounded, with minimal loop area" to "probe connected to the device under test, grounded via ground spring", what does the noise look like? Still unacceptably high? 
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #6 on: March 12, 2024, 02:33:08 pm »
I now measured 13mVpp with GND of the probe connected wthout ground spring. With additionally Earth closer to the probing point, it goes down to 4mVpp. With groundspring i got 0.9mVpp.

Connecting earth pretty much elimates the spikes from the power supply, without really affecting the lower noise. But removing the ferrite makes it worse again (the spikes).
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #7 on: March 12, 2024, 03:27:27 pm »
So, with the scope grounded and proper probing technique the power supply noise is negligible? That sounds good -- although it is a bit strange that the additional ferrite is required. Maybe some other DHO owner can chime in here with comparison data?

If you want to test the intrisic noise of the scope, and check whether any PSU noise couples into the scope directly via the power line, you can disconnect all probes and shorten the inputs, or connect a 50 Ohm terminator right to the input BNC.
 

Offline jim_griff

  • Regular Contributor
  • *
  • Posts: 81
  • Country: gb
  • Personal Text goes here.
    • Principal Audio
Re: Spikes and noise on Rigol DHO814
« Reply #8 on: March 13, 2024, 12:21:35 am »
Hi there. My DHO804 arrived today. I found the same issue with the switching USB-C PD PSU supplied with the Rigol. Haven't tried any other supplies as I don't own any, but will be figuring out a way of dealing with it. It shouldn't be this way.

I use 20MHz BW limit (doing audio work), but pics are my measurements *NOTE: noise is only slightly higher without 20MHz BW limit.*

All tests done with max vertical magnification (500µV/div, 1ms/div, 10M memory depth, 20MHz BW limit). Persistence time: Infinite. Probe set to 1x and properly compensated.

Pic 1: Probe connected to scope; has "clamp" cover on it. [~4000µVp-p]
Pic 2: Probe connected to scope; no clamp cover on it, no ground spring.  [~900µVp-p]
Pic 3: Probe connected to scope; has ground spring connected.  [~850µVp-p]
Pic 4: Probe connected to scope; has ground spring shorted to probe pin.  [~250µVp-p]
**Test 4 is lower in noise than with an unterminated (floating) BNC with no probe connected, which makes sense.

All tests were done with the probe sitting in exactly the same position. Probe technique will count for a lot, but the switching noise is still visible in some instances. No other hardware in the vicinity (switching supplies, computer, TV, monitor, speakers, studio equipment, psu) is causing any noticeable noise. I have a fair bit of equipment. ONLY the supplied Rigol PSU is imparting visible spikes/ringing artifacts.

I suspect the psu is built down to a price point and will be finding a linear USB-C PD supply for future use.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #9 on: March 13, 2024, 04:59:56 am »
Try switching off the lights and unplug all the wall-warts in the area:



I suspect the psu is built down to a price point and will be finding a linear USB-C PD supply for future use.

Not unless it's defective. The majority of people don't get this noise.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #10 on: March 13, 2024, 05:00:47 am »
It is good practice to move the power supply as far away from the scope and the device under test as possible, and to definitely ground the scope (both for EMC and safety reasons).

Ground won't make any difference to this.

It's usually something external.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #11 on: March 13, 2024, 07:51:43 am »
Ground won't make any difference to this.

Wrong! It made it better. Altough it wasnt huge.

I have already proven, that the problem is the powersupply, so turning anything else would make no difference.  :palm:
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #12 on: March 13, 2024, 09:08:21 am »
I have already proven, that the problem is the powersupply, so turning anything else would make no difference.  :palm:

Where did you prove that?
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #13 on: March 13, 2024, 09:15:25 am »
Just in my first post! I found the same frequency on the cable, and doing measures for this, got rid of that spikes quite a bit!

Did you actually read what i wrote, and looked at the pictures???
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #14 on: March 13, 2024, 09:19:40 am »
Hi there. My DHO804 arrived today. I found the same issue with the switching USB-C PD PSU supplied with the Rigol. Haven't tried any other supplies as I don't own any, but will be figuring out a way of dealing with it. It shouldn't be this way.

I only just watched your pictures, but there are no spikes like mine. You should scale it like i had, to see if this is really the same problem. You could also measure on the cable with a coil.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #15 on: March 13, 2024, 10:40:01 am »
Just in my first post! I found the same frequency on the cable, and doing measures for this, got rid of that spikes quite a bit!

Did you actually read what i wrote, and looked at the pictures???

That doesn't prove anything.

It can be coming through the power supply from the mains.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #16 on: March 13, 2024, 01:17:24 pm »
I happily would try to investigate this, if you tell me how i could track the sourc of it.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #17 on: March 13, 2024, 01:24:16 pm »
I happily would try to investigate this, if you tell me how i could track the sourc of it.

You could try and block interference on the primary (mains) side, before the power supply: Use a plug-in mains filter on the 230V side, or put ferrite cores on the 230V input cable instead of the 15V output. Or move the scope to a different mains phase (different room), or better yet to a friend's house, for comparative testing.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #18 on: March 13, 2024, 06:33:57 pm »
Shouldnt the powersupply filter this all out and not generate it either?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6506
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #19 on: March 13, 2024, 07:11:30 pm »
Shouldnt the powersupply filter this all out and not generate it either?

It's not black and white...

All switching power supplies generate some level of switching noise on their outputs. (There is a reason why good laboratory power supplies, which you may want to use e.g. for sensitive analog prototype circuits, still use big fat transformers running at 50 or 60 Hz and linear regulators, rather than the much cheaper switchers.) The output of a switching supply does not have to be perfectly clean if the device that you power from it does some filtering of its own.

The power supply may or may not have some internal input filters. In any case it will suppress most of the incoming noise by charging a large capacitor, then chopping the voltage up again for transfomation. But it does not suppress 100% of the incoming noise: If a particularly nasty source of voltage spikes is operating on your mains, some of that may come through.

(A motor with bad interference suppression is a classical source of powerline interference. Very cheap switching supplies are a more modern, but rather common source. E.g. I originally used a cheap no-name 12V supply to power the LED light strips on my bench, and that created nasty interference spikes everywhere...)

Finding the source of interference signals, and the path how they couple into your measurements, is not an easy task. Small changes can have a big impact, and there may be influences which you are unaware of and/or which are outside of your control -- e.g. some device which only runs part of the time and which causes interference. Be diligent and systematic, and take good notes: short description what you did and changed, note what was connected to the scope, photos of shielding or probe arrangements, screenshots of scope data. You will track it down eventually!
 
The following users thanked this post: egonotto, DaneLaw

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #20 on: March 14, 2024, 04:43:31 am »
I happily would try to investigate this, if you tell me how i could track the sourc of it.

As noted in the video I posted. Try switching any LED lights off, unplug everything around it.
« Last Edit: March 14, 2024, 04:45:10 am by Fungus »
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #21 on: March 14, 2024, 05:40:59 am »
I have done this. Im thinking about talking it out ona field with my camper ;-). It has a 220V system.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #22 on: March 14, 2024, 01:02:38 pm »
Try it at a friend's house, or at work...
 

Offline jim_griff

  • Regular Contributor
  • *
  • Posts: 81
  • Country: gb
  • Personal Text goes here.
    • Principal Audio
Re: Spikes and noise on Rigol DHO814
« Reply #23 on: March 14, 2024, 03:32:19 pm »
Pic 1: Probe sat by scope with clamp connected.
Pic 2: Probe sat by scope with clamp disconnected.
Pic 3: Probe hanging next to the Rigol PSU brick.

The noise is definitely coming from the Rigol PSU brick. The waveform shape is different, and the lower frequency ringing is likely due to parasitics in the power cable going to the scope.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #24 on: March 14, 2024, 03:45:37 pm »
It would really help, if you would scale it, to see its frequency. Without this, it could be another source too, as stated by the others here.
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline jim_griff

  • Regular Contributor
  • *
  • Posts: 81
  • Country: gb
  • Personal Text goes here.
    • Principal Audio
Re: Spikes and noise on Rigol DHO814
« Reply #25 on: March 14, 2024, 04:11:21 pm »
~17µs / ~58kHz switching frequency. The same frequency, but much higher in amplitude, when putting the probe next to the power brick. It's definitely the culprit.
 

Offline jim_griff

  • Regular Contributor
  • *
  • Posts: 81
  • Country: gb
  • Personal Text goes here.
    • Principal Audio
Re: Spikes and noise on Rigol DHO814
« Reply #26 on: March 14, 2024, 04:22:12 pm »
~17µs / ~58kHz switching frequency. The same frequency, but much higher in amplitude, when putting the probe next to the power brick. It's definitely the culprit.

Here are measurements with two probes. One sitting on the desk (yellow trace), the other by the Rigol PSU brick (blue trace).

Yellow is 500µV/div; Blue is 100mV/div
EDIT:(20MHZ B/W limit.)
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #27 on: March 14, 2024, 09:38:26 pm »
The noise is definitely coming from the Rigol PSU brick.

...or through it from a dirty mains supply.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #28 on: March 14, 2024, 10:09:43 pm »
The noise is definitely coming from the Rigol PSU brick.

...or through it from a dirty mains supply.

So then where in the world can this signal come from? It must be somewhere between germany and UK? Maybe its the wind turbines in the sea??  :palm:
« Last Edit: March 14, 2024, 10:12:18 pm by eTobey »
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 
The following users thanked this post: egonotto

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #29 on: March 14, 2024, 10:56:29 pm »
So then where in the world can this signal come from? It must be somewhere between germany and UK? Maybe its the wind turbines in the sea??  :palm:
Nah it must be coming from the outer space, as I'm seeing it too:

2072561-0

This was captured with the probe in the 10:1 switch position and the ground attachment shorted to the tip, creating a loop, lying right on top of the PSU.
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #30 on: March 14, 2024, 11:00:14 pm »
Definitely not the cleanest PSU. I tried a car cigarette lighter plug with USB-C outputs, and it was actually better than this. So it makes sense to look for a less noisy power supply, but the question is how do we tell them apart from each other without buying and returning a dozen or two before finding a better one.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #31 on: March 15, 2024, 07:02:56 am »
Definitely not the cleanest PSU.

Mine is perfectly clean and so is just about every other person on here.

So unless yours is defective then the spikes are coming from the environment.

Spikes like these can definitely come from the mains. Dave's done videos on tracking these problems down, I posted one above.

I tried a car cigarette lighter plug with USB-C outputs, and it was actually better than this.

In your car?
 

Offline jim_griff

  • Regular Contributor
  • *
  • Posts: 81
  • Country: gb
  • Personal Text goes here.
    • Principal Audio
Re: Spikes and noise on Rigol DHO814
« Reply #32 on: March 15, 2024, 07:41:32 am »
Definitely not the cleanest PSU.

Mine is perfectly clean and so is just about every other person on here.

So unless yours is defective then the spikes are coming from the environment.

Spikes like these can definitely come from the mains. Dave's done videos on tracking these problems down, I posted one above.

I tried a car cigarette lighter plug with USB-C outputs, and it was actually better than this.

In your car?

The evidence all points towards the PSU bricks of some people's devices. Occam's Razor:

1. Some people who own this scope are measuring the exact same signal, and in totally different parts of the world.
2. These signals are all being tracked down to the Rigol PSU brick (inductive coupling tests).
3. These signals are not able to be measured through the mains 120V/230V supply (unless Rigol PSU switched on).
4. Other devices and power bricks do not exhibit this exact signal and have their own switching signature, but much lower amplitude.
5. The signal disappears when using other power supplies to power the scope.

It's 100% noisy Rigol power bricks, albeit not all of them. Some have obviously been engineered better than others - different revisions, perhaps.
 

Offline csuhi17

  • Regular Contributor
  • *
  • Posts: 215
  • Country: hu
Re: Spikes and noise on Rigol DHO814
« Reply #33 on: March 15, 2024, 08:38:07 am »
We have known this problem for a long time.
The early sales gave a noisy power supply, then if I remember correctly they switched to LiteOn and Lenovo.
I didn't see in the above posts which one was added for those of you who find it noisy.

I use it with a sound technician's 12V power supply, but the LiteOn that comes with it doesn't make more noise for me either.

I tested the LiteOn, I charged a solar powerbank with it, it said 20V and 66W.
I couldn't test it at 15V.

In the OP's first post, I don't see what the channels are set to. 1x 10x
I also don't know if there is something connected to CH1, CH2.

If you have the opportunity to try it with another power supply, perhaps with a car battery, you can find out that everything is fine with the oscilloscope.
If you can't solve it any other way, then return it.
Maybe write to the seller about what can be done in this case.

There are two pictures of the power supply, with 20V 66W.
And the third is just for interest, the ground wire of the probe was oxidized or dirty at the part where it connects to the probe. With a handheld scope, I only measured ~40Vpp with the same probe.
After I cleaned everything was fine.
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #34 on: March 15, 2024, 09:32:30 am »
Mine is perfectly clean and so is just about every other person on here.
No switching supply is perfectly clean. Did you actually try to capture its EMI by shorting the probe's ground strap to the tip and using it as an antenna to sense noise close to the PSU?

I'll tell you more, the scope itself is also emitting like crazy. Easily sensed in the vicinity of the input power connector, where the several switching converters are located. But at least it doesn't seem to couple into the analog input lines.

So unless yours is defective then the spikes are coming from the environment.

Spikes like these can definitely come from the mains. Dave's done videos on tracking these problems down, I posted one above.
Are you kidding me? Two people thousands of kilometers apart show an identical switching noise waveform, (which in my case raises in amplitude dramatically as the ground loop antenna gets closer to the PSU brick), and you still tell that this may be coming from the environment?

I tried a car cigarette lighter plug with USB-C outputs, and it was actually better than this.
In your car?
No, on my bench. Cars aren't the only source of 12V DC power in the world.
« Last Edit: March 15, 2024, 09:40:16 am by shapirus »
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #35 on: March 15, 2024, 09:38:28 am »
The early sales gave a noisy power supply, then if I remember correctly they switched to LiteOn and Lenovo.
Mine is LiteOn.

I didn't see in the above posts which one was added for those of you who find it noisy.
Mine isn't particularly noisy and isn't a problem in practice, as the emitted interference quickly drops down to undetectable levels as you get farther away from it, but its noise is easy to detect when you look for it.

Some other switching supplies around me are noisier. All kinds of noise can be seen when probing with an alligator clip ground strap. Using a spring attachment, or, if possible, a coaxial adapter, helps.

I use it with a sound technician's 12V power supply, but the LiteOn that comes with it doesn't make more noise for me either.
How did you capture the attached waveforms (how was the probe connected)? They don't look like typical switch mode PSU EMI spikes.
 

Offline csuhi17

  • Regular Contributor
  • *
  • Posts: 215
  • Country: hu
Re: Spikes and noise on Rigol DHO814
« Reply #36 on: March 15, 2024, 10:05:19 am »
I use it with a sound technician's 12V power supply, but the LiteOn that comes with it doesn't make more noise for me either.
How did you capture the attached waveforms (how was the probe connected)? They don't look like typical switch mode PSU EMI spikes.

I charged an Allpowers 288Wh solar power bank/generator with the LiteOn power supply.
Between the two is a USB-C male female test connector.
I measured between Vcc and Gnd, with 10x.

If I have time, I'll try with an adjustable linear load, I couldn't find anything else at the moment.
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #37 on: March 15, 2024, 10:21:31 am »
I charged an Allpowers 288Wh solar power bank/generator with the LiteOn power supply.
Between the two is a USB-C male female test connector.
I measured between Vcc and Gnd, with 10x.
That makes sense then.

What we've been discussing is this:



To capture the radiated interference.
 

Offline csuhi17

  • Regular Contributor
  • *
  • Posts: 215
  • Country: hu
Re: Spikes and noise on Rigol DHO814
« Reply #38 on: March 15, 2024, 11:07:27 am »
To capture the radiated interference.

Everything is clear now.
I misunderstood, I thought he could see the noise from the power supply on the channels.

I checked my nearby switching power supplies using that method.
For me, they were all much noisier than the Rigol power supply.
I measured 2 power supplies of Micsig, an LED light source, a computer, the power supply for a 12V refrigerator, and even my Chinese metal housing power supply.

In all the videos I've watched so far, the power supply had a metal housing in the scopes. Maybe they could have added a properly insulated laptop charger with a metal housing, if such exists at all.

Wouldn't a thicker metal box have been better than the aluminum foil?
 
The following users thanked this post: Fungus

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #39 on: March 15, 2024, 11:23:58 am »
Everything is clear now.
I misunderstood, I thought he could see the noise from the power supply on the channels.
It's possible to see this noise during normal measurements too. The signal being measured must not be too noisy itself, the ground strap instead of the ground spring must be used. Sometimes it's possible to catch it even with the ground spring.

I checked my nearby switching power supplies using that method.
For me, they were all much noisier than the Rigol power supply.
I measured 2 power supplies of Micsig, an LED light source, a computer, the power supply for a 12V refrigerator, and even my Chinese metal housing power supply.
Yeah, even if we power the scope with the cleanest PSU possible, then there still will be a lot of crap around emitting all sorts of switching noise, so really the solution is to use probing techniques that help minimize their influence, and to get rid of the worst offenders: for example, in one case, when measuring how well an RC filter on a DC power rail worked, I saw some switching noise and thought that it was coming from a switching PSU powering the circuit, until I saw that the noise remained even after switching the PSU off. It turned out that the interference was coming from a generic chinese DC/DC step-down module used in a homemade device sitting ~3 meters (!) away from the probe. That one must be really poorly designed, I'll have to replace it some day, or make one myself, or maybe just wrap it in foil.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16672
  • Country: 00
Re: Spikes and noise on Rigol DHO814
« Reply #40 on: March 15, 2024, 11:49:40 am »
To capture the radiated interference.

Everything is clear now.
I misunderstood, I thought he could see the noise from the power supply on the channels.

Yeah, that wasn't made clear.

Simple solution: Don't do that!
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1355
  • Country: ua
Re: Spikes and noise on Rigol DHO814
« Reply #41 on: March 15, 2024, 12:12:18 pm »
Well, it's not fully clear how exactly OP measured/captured that noise and in what situation it was causing trouble.

I assumed it was EMI, because I've seen that myself. With the probes disconnected this switching noise doesn't make it to the displayed waveform in my case. Maybe the OP's situation is different.
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #42 on: March 15, 2024, 05:46:28 pm »
I measured it on my prototype, which is powered through a cheap power supply. Maybe it came through this. But that Rigol powersupply should not spew all that noise all over the place anyway right?
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 

Offline csuhi17

  • Regular Contributor
  • *
  • Posts: 215
  • Country: hu
Re: Spikes and noise on Rigol DHO814
« Reply #43 on: March 15, 2024, 06:34:09 pm »
Not in my case.
With the above method, I have to put the probe quite close to the power supply, I have to almost touch it. Also, there are parts and positions where less noise is detected.
If I touch the probe with my hand or its wire, the noise increases.

If you measure the noise on the cheap power supply, you will be able to determine whether it is bothering you.
The power supply I bought for my Riden is so noisy that it was also visible on the regulated output.

When I measure, I always place such power supplies as far away as possible.

In the very first post, the channel setting and the probe were set the same, both 1x or 10x?
 

Online eTobeyTopic starter

  • Frequent Contributor
  • **
  • Posts: 557
  • Country: de
Re: Spikes and noise on Rigol DHO814
« Reply #44 on: March 15, 2024, 07:24:17 pm »
Yes
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant." (Maxim Gorki)
 


Share me