34465A is set to 10V range, input Z auto.
B2987A provides voltage to 34465A, and measure current.

V(volt)   I(pA)
0      7.1
1      9.1
2      10.2
3      11.1
4      13.1
5      15.2
6      15.5
7      15.7
8      14.8
9      13.8
10      12.2

It's strange that the current gets smaller when V>8.
 
The following users thanked this post: Dr. Frank

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14214
  • Country: de
Re: strange 34465a IV test result
« Reply #1 on: March 09, 2022, 12:16:30 pm »
The input current does not have to be a linear function of the voltage. Especially the current from charge injection due to Az switching can be nonlinear and possibly getting smaller with higher voltage.
 
The following users thanked this post: Kean

Offline analogNewbieTopic starter

  • Regular Contributor
  • *
  • Posts: 54
  • Country: cn
Re: strange 34465a IV test result
« Reply #2 on: March 09, 2022, 03:37:45 pm »
Thanks
The data is much better after turning off auto zero
V   I
0   1.0
1   1.9
2   2.0
3   2.0
4   2.7
5   4.1
6   5.3
7   6.5
8   7.6
9   8.6
10   9.9

 