In relation to my previous forum I make another one.
Now I've replaced both the TVS diode and Q9 and now I have a bias voltage of 15, sweat!.
Unfortunally the psu still doesn't work, it have no output from the transformer and for some reason the 12V bias is 15V. This is no strange because the service manual refer to 12V bias voltage to a rail that than is ratet to 14.65V. (See the picture below for more detail).
I also post the A19 board schematics.
There some common failure on the A18 board that is the switching part of the psu?
There some else to check to make sure that everything is ok?