TDS5054B-NV-AV too much signal noise?
(1/1)
newtekuser:
Would you consider this too much signal noise? Regardless if it's from the probe compensation hook on the front panel, or acquiring signal from a DUT.
Doesn't matter if I use a more expensive Tektronix P5050 500MHz probe, or a cheap $20 probe from ebay, I get the same result, bad acquisition board, or is this normal for this scope?

I don't have problems acquiring signals from any of the 4 channels and SPC passes. However, if I press the "Calibrate" button and wait 5 mins for full calibration to complete (successfully mind you) and exit out of that screen, tekscope app crashes and forces me to shut down the scope.

xtech:
Your square signal looks fine, don't expect ideal straight line. But scope crash is definitely not ok.
newtekuser:

--- Quote from: xtech on February 20, 2023, 07:37:39 am ---Your square signal looks fine, don't expect ideal straight line. But scope crash is definitely not ok.

--- End quote ---

Thanks! Actually the scope crash isn't happening anymore after reinstalling the scope from the factory restore discs.