Products => Test Equipment => Topic started by: tchiwam on September 15, 2020, 02:57:43 pm
-
I have this unit with a SPC fail, I got the channels 3 and 4 behaving well and the channel 1 and 2 are showing something reasonable at 200mV and up ( upper side of the relays)
Few questions:
Is the ADC the same for both range 10mV to100mV and 200mV to 1V or is it on different pins ?
Anyone having a drawing / diagram for that unit ?
I suspect the ref signals for the SPC or relays to be bad, if I tweak cal, I can see reasonable traces...
-
Don't you love it when they say As Is because we can't test it ...
Then you open it and someone really tried to frill it...
-
Well, I have at least 2 good trigger chips and maybe a 3rd one. Can't test the channel 1 yet...
-
Did you try to check the service manual?
https://www.tek.com/oscilloscope/tds794d-manual/tds500d-tds600c-tds700d-tds714l-service-manual (https://www.tek.com/oscilloscope/tds794d-manual/tds500d-tds600c-tds700d-tds714l-service-manual)
-
Yes, I need something closer to a schematic of the A10 board.
I am suspecting that one of the relay is dead, channel 1, then I think the reference for calibration is out of specs, it is warming up the internal termination.