| mlloyd1:
Robert: Thanks so much for your VERY informative investigation. :-+ My V-665 has this exact problem and I'd like to repair it so my grandson can have this scope when he is ready for one. However, I think the exact OEM replacement is actually the Fujisoku ALE2G-2M4-10-Z, right? Looks like the ALE2S-2M4-10-Z is a rocker. mlloyd1 |
| rbm:
Hi, Either model can be equipped with a toggle or rocker actuator. The difference between the G and S models is that the G has 2.54mm between contact rows and the S has 5.08mm between contact rows. The S is the correct original designation. |
| old gregg:
Hi, I'd like to change the push/pull in my hitachi VC6023, they're all dirty and the switching isn't done properly anymore (the trigger becomes quite tricky). The service manual gives as references EVH - YK3325B14 (10K push pull). Does anyone know where I can find a replacement or a sibstitute ? (It's annoying to not being able to use the X10 mag) I googled the refrences with no success. |
| Ether:
Hi, is it possible to repeat the frequency mod on a V-1560. Also, is there some way of preserving the holdoff function while getting the frequency counter? |
| grumpydoc:
--- Quote ---Hi, is it possible to repeat the frequency mod on a V-1560. --- End quote --- Probably - have a look at your CPU board and see if the jumpers correspond to the "no frequency counter" setting. If so swap them. --- Quote ---Also, is there some way of preserving the holdoff function while getting the frequency counter? --- End quote --- Not sure what you mean by preserve? |