dianzimi:
yes
dianzimi:
yes
tonyM:
I think these 3 black parts (excluding input capacitor) at FSH3 input signal path are resistors, which are making Pi-pad attenuator.
As you can see, TSMU (ver.1 and ver.2) lacks Pi-Pad attenuator. Only one Zero resistor (see pictures below):
If you put Zero resistor in the right place (as in TSMU-ver.2), I think your FSH3 will work (if next components/ICs are OK), but the measured levels will not be correct.
As I don't have FSH3, I will propose the following experimental approach for calculating the Pi-pad resistors values (and calibrating the instrument):
1. Put zero resistor at the input signal path
2. Plug-in signal generator with known level at RF-input
3. Check the level, measured by FSH3, and calculate the difference in dB
4. As you know the difference (and input impedance=50 ohm), use calculator https://www.pasternack.com/t-calculator-pi-attn.aspx to find the values of the resistors.
hgl:
There is an interesting detail, the pin 2 of the synthesizer chip is isolated on the TSMU version 2. (Not properly visible in the photo, you can see only a small piece of brown foil under the pin). This is not the case with version 1 and FSH3. On the photos you can not see whether the synthesizer chips of FSH3 and TSMU are really the same.
dianzimi:
who knows the components detail information in the red rectangle bellow??[img][img]