Another question or two for those in the know: At the beginning of the Cal 02 procedure, it states to connect a 0.5 V standard-amplitude signal to CH 1 and to use CH 2 Position to vertically position the "trace" to within 1 division of the center graticule line, and then to use the CH 1 Position and Volts/DIV VAR controls to obtain a 10-division horizontal signal. I don't get a "trace," only 2 dots. I can position the dots with the controls to put one dot on the first graticule line and the other on the 10th.
Question 1: "Is this "normal," or am I doing something wrong?" I've read and re-read the procedures and can't find any references to this.
Also, at the beginning of the Cal 01 procedure, it states that, "Upon entering Cal 01, the Input Coupling is automatically set to 50
DC ....." This doesn't appear to be true for the other Cal procedures (2,3,4,etc) as I had to install a terminator on my generator to get the actual levels called out in the test procedures (monitored on another scope).
Question 2: "Is this correct, or am I missing something here, as well?
Thanks in advance for any input!
Gerry