Author Topic: Tek 2755Ap: missing +9V  (Read 3505 times)
Offline Martin.MTopic starter
  • in Tek we trust
    • vintage Tek collection
« on: March 29, 2024, 01:46:51 pm »
hello friends and happy easter,
I have the following problem:
Tektronix Spec: 2755Ap
missing +9V
I have not found application or manual.
There is a rest voltage around +1,2V, so it may be possible anywhere is a LM317 with a bad 240 ohms to ground: the only logical answer to this voltage.
But I don´t want to make shitty work here, this is a wonderful instrument. So please dear insiders, where I have to look for this problem?
It`s a ESD home, so I don`t like unscrew too much parts.
thank you for a tip and greetings from germany
Martin
Offline Testtech
Re: Tek 2755Ap: missing +9V
« Reply #1 on: March 29, 2024, 09:25:27 pm »
Check the fuses under the little cover on top of the PSU.
The following users thanked this post: Martin.M
Offline shakalnokturn
« Reply #2 on: March 29, 2024, 11:04:00 pm »
You'll find at least the 494 schematic on KO4BB's website (think of donating once in a while, all the traffic and site maintenance cost him).
On these SA's à lot of tantalum capacitor problems with age.
Offline KE5FX
    • KE5FX.COM
« Reply #3 on: March 30, 2024, 02:41:38 am »
All of the 275x models are rack-mount versions of one 49x model or another. The 2755AP is a repackaged 492BP if I recall correctly.  Basically a 494AP with 100 Hz minimum resolution rather than 10 Hz.
 I'd probably use the 494AP manuals from ko4bb.com to troubleshoot it; the 494 manuals may be a little too old.
  • in Tek we trust
« Reply #4 on: April 01, 2024, 11:05:37 am »
thank you. 2A was blown.
LM317 at front board was defect, also the driver transistor for ON/OFF the light.
