Hi i would carefull with pulling these eeproms. Inside are your factory cals stored and your scope needs tobe re-calibrated for sure. nowadays no FAS is available anymore and equiptment like a fluke 9500B needs to be part of your workbench or you will get yourself into an extra problems. $$$
is the rest of your scope ok bohrzqc . post log here.
After replaced the X24165 chip, the 521 error still exists. I wonder whether only replacing the X24165 chip is OK or some cal data should be written to the chip. Jay upload the cal data of his 7104. So the cal data should be written to the chip?
I bought some AT24C16B chips, and plan to replace tomorrow.
Maybe I solved the baseline offset is reset to the factory default. First delete the calConStatus = cc_init(2) in the boot file, save the file. Restart the scope. Add the calConStatus = cc_init(2) to the original place again, and save the file, restart the scope.
However, this method is useless for another 7104. Maybe this due to the X24165 chip. Not sure.