Author Topic: Tek CSA7404/TDS7000 repair project (Upgraded memory/CPU)  (Read 116902 times)
Online TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
« on: September 16, 2013, 03:59:51 am »
This thread is dedicated to Tektronix TDS7000 platform based scopes (TDS7054,TDS7104,TDS7154,TDS7254,TDS7404,CSA7054,CSA7154,CSA7404)
Some information can be useful for newer TDS7000B platform as well, as they are based on similar modules.
Recently I acquired Tektronix CSA7404 scope, bought as dead for parts, at nice price. Unit come in pretty bad physical shape  :-BROKE:
All plastic exterior parts except front panel were missing. Outside aluminum shielding/case covers were bent on corners, looked that scope was dropped on angle.
Despite of damages and poor look - it still was able to power on and booting Windows 98 OS. Hard drive had bad SMART status and was making clicking noise, so obviously it's going to be replaced with new HDD as one of first steps of scope restoration.
While machine was able to boot OS without problems, main scope application, TekScope was unable to start. It was just showing welcome splash screen and hanging there. Downloading fresh copy of TekScope 2.5.3 Firmware for CSA7000/TDS7000 from Tektronix site and reinstalling application did not help either.
Before we dig in, let's take a look on high-level block diagram of Tektronix CSA7404 from service manual :
There are four main modules/boards in scope.
x86 PC system (NLX motherboard, Socket 370 Celeron 850/128/100 Coppermine CPU, i810 chipset, onboard VGA, network, USB, SDRAM PC133 x 2 slots (only one used with 256MB DIMM)
PowerPC system (Tektronix custom board with interface to NLX and ACQ board (had installed 256MB DIMM on it))
Acquisition board (ACQ, front-end to scope channels, custom Tektronix high-speed ASICs with dedicated RAM and interface to PowerPC board)
Main power supply board (cross compatible, at least with tested CSA7404 and TDS7704B)
I also had access to working Tektronix TDS7704B, which come handy to check some parts and do board-level testing. Obviously both these scopes use common architecture, and share same power supply board and similar (but not interchangeable) PowerPC/x86 combo. As first measure of any electronic device troubleshooting - check power. Swap whole power supply module from TDS7704B to CSA7404 showed no functional issues or differences, which tells me that power supply to other boards from mains PSU is OK.
Initial testing and debug
First step will be visual inspection of all internals of scope to see, if any components or boards are physically damaged or missing  :-/O.
On acquisition board there was missing heatsink on one of big National Semi custom ASICs for channel 3. New heatsink was bought are placed on chip, hoping that it's BGA connection to board is intact.
On x86 board there was leaky output electrolytic capacitor in CPU DC-DC convertor. It was replaced to new one, which is common operation for servicing old computer motherboards already for a while. As a side note, considering cost of such scopes even today it's quite unpleasant to see low-end CapXon capacitors being used.
As was already stated earlier - there were no power supply board problems, as all voltages were present, and no visual damages were observed either. Swapping PSU from another scope did not make scope progress on loading app any further.
Interconnect boards (P200 between PPC board and ACQ, power delivery interconnect boards) were fine too, replacing them to same from TDS7704B did not change anything. So problem lied either in acquisition board, or PowerPC board or interfacing between NLX and PPC boards or PPC/ACQ. Interesting fact, that removing interconnect board P200 from PPC to ACQ on TDS7704B did exactly same - TekScope app stopped working and hung on welcome splash screen. This gives us information, that TekScope app needs to communicate successfully with PPC and ACQ boards to be able to start.
After some checks of x86 side of scope and trying to run few HDDs with Windows 2000 no obvious problems was found. CD-ROM drive was changed to DVD-RW from old laptop. Connection was standard notebook PATA. Original IBM Travelstar DARA-206000 6GB 4200RPM disk was changed to fresh IBM Travelstar IC25N020ATDA04-0 20GB drive. New Windows 2000 SP4 Professional was installed as maximum supported OS on this series scopes. Given current Celeron 850MHz CPU and 256MB this setup should run smoothly for beginning.
Moving our focus to PowerPC board as main suspect. This module There is handy debug 7-segment single digit debug LED, which according to service manual have next codes:
This LED is located on PPC board near riser connector and easily visible even when outer aluminum shell is installed.
Our board during power up and after was rolling from ".8" code to "L", stays there for couple seconds, and then restarting from start again, and doing this infinitely in loops. On working TDS7704B post code LED was going from ".8" to "L", then when OS started goes to "P" and finally to flashing "-", staying there when TekScope application running. Looks like PowerPC board unable to finish it's booting sequence.
NVRAM Failure (Scope boots, OS boots, but TekScope application hangs on welcome splash screens, front panel not responding
Due to age of these machines and usage of battery-backed NVRAM (Dallas DS1245) it's not rare that onboard 3V battery (similar to ones used in usual computers for CMOS data retention) loose it's voltage. Unluckily, boot parameters and configuration is stored in PowerPC NVRAM as well, so after battery failure scope likely to fail, due to inability to load boot ROM from x86 PC system HDD. Visual diagnosis can be done by observing TekScope application to hang on splash screen, and debug POST code indicator rolling codes over and over from 4 to L on PPC board. This is happening because when PPC boot loader lost its boot settings and cannot find correct ROM anymore, it reboots after 10 seconds to try again. That's why it's staying on "L" phase for 10 sec before it repeats loop.
Quickly I found users with very same issues with their TDS7000/CSA7000 series machines here
NVRAM repair solution / fix
First step is to replace faulty battery and allow Dallas NVRAM to be functional again.
Locate NVRAM on PPC board:
On CSA7404 it's near NLX riser connector, and have PowerCAP module on top of DALLAS DS1245YP-100. PowerCAP is just simple PCB with BR1632 battery soldered on and plastic cover. I replaced old battery (voltage on it was 0.7V after removal  :bullshit:) with new Panasonic BR1632A/FA. Make sure polarity is correct when soldering battery on. Latch PowerCAP with new battery back on top of NVRAM.
Now we need to reset again correct boot parameters. To do so there are few extra tools are required:
* Serial interface to your host, or RS232-USB dongle
* Few wires for connecting PowerPC debug serial port
* Any terminal software which can work on serial port
I had no RS232 on my PC, but had few UART-USB interface bridges based on Silabs CP2103. RS232 levels are converted from UART by onboard MAX232 Driver IC , which is located near NLX riser connector on PPC board. I connected GND (Pin 15), TX (Pin 11), RX (Pin 12) to Silabs CP2103 dongle inputs and connected terminal at 9600, 8N1 settings.
First boot output with 9600 baud rate, 8N1 settings :
romPost: MPC106  Walking-one passed.
romPost: MPC106 configuration passed.
romPost: First PCI access  passed.
romPost: PC87560 Walking-one passed.
romPost: PC87560 configuration passed.
romPost: DEC21554 configuration passed.
romPost: PERFORMING ROM CHECKSUM TEST...
romPost: ROM checksum passed.
romPost: PERFORMING DRAM CELL TEST...
romPost: Enabling data cache...
romPost: Data cache enabled.
romPost: Enabling instruction cache...
romPost: Instruction cache enabled.
romPost: DRAM Cell test passed.
romPost: Disabling data Cache...
romPost: Data cache disabled.
romPost: Disabling instruction cache...
romPost: Instruction cache disabled.
romPost: PERFORMING DRAM MARCH TEST WITH CACHE OFF...
romPost: Writing address to each *address...
romPost: Reading address from each *address...
romPost: DRAM march test passed...
romPost: PERFORMING DRAM MARCH TEST WITH CACHE ON...
romPost: PERFORMING DRAM WALKING-ONE  TEST...
romPost: DRAM Walking-one test passed.
romPost: PERFORMING NV-MEMORY WALKING-ONE  TEST...
romPost: NV-MEMORY Walking-one test passed.
romPost: POWER-UP DIAGNOSTIC COMPLETE.
Hel
                            VxWorks System Boot
Copyright 1984-1996  Wind River Systems, Inc.
CPU: Tektronix HPP - PowerPC 740
Version: 5.3.1
BSP version: 1.0.4.0
Creation date: Nov  2 2000, 11:03:34
Press any key to stop auto-boot...
 7 6 5 4 3 2 1 0
auto-booting...
boot device          : ??D?h?
processor number     : 0
flags (f)            : 0x0
Network interface   D¦h  unknown.
Can't load boot file!!
[VxWorks Boot]: Starting at 0xfff00100...
Then it rebooted and repeated above sequence again, and again, and again.
OK, so now we know that PowerPC board works, but obviously boot device "??D?h?" does not look right, thus it cannot load ROM and boot it's OS.
So let's halt auto-booting process by sending any char to terminal. By doing so we can stop loop and try to interact with PPC system. Now can send "?" command:
[VxWorks Boot]: ?
 ?                     - print this list
 @                     - boot (load and go)
 p                     - print boot params
 c                     - change boot params
 l                     - load boot file
 g adrs                - go to adrs
 d adrs[,n]            - display memory
 m adrs                - modify memory
 f adrs, nbytes, value - fill memory
 t adrs, adrs, nbytes  - copy memory
 e                     - print fatal exception
 n netif               - print network interface device address
 N                     - set ethernet address
 $dev(0,procnum)host:/file h=# e=# b=# g=# u=usr [pw=passwd] f=#
                           tn=targetname s=script o=other
 Boot flags:
   0x02   - load local system symbols
   0x04   - don't autoboot
   0x08   - quick autoboot (no countdown)
   0x20   - disable login security
   0x40   - use bootp to get boot parameters
   0x80   - use tftp to get boot image
   0x100  - use proxy arp
   0x1000 - use dio to get boot image
available boot devices: feippc bp sm
After some experimenting and searching web, I found somebody who had similar issues with CSA8000 oscilloscope. Tried very similar parameters and discovered that they work fine after minor modification for my system:
Send "c" command to set settings:
[VxWorks Boot]: c
'.' = clear field;  '-' = go to previous field;  ^D = quit
boot device          : sm
host name            : host
file name            : c:\vxBoot\vxWorks.st
inet on ethernet (e) : 192.168.0.2:ffffff00
inet on backplane (b): 192.168.0.2:ffffff00
host inet (h)        : 192.168.0.1
gateway inet (g)     : 192.168.0.2
user (u)             : anonymous
ftp password (pw) (blank = use rsh): anonymous
flags (f)            : 0x1000
target name (tn)     : target
startup script (s)   : c:\vxBoot\dio_rst.vsh
other (o)            : nvfs=0x1000
I used path to standard location on my x86 HDD partition, which is hidden folder c:\vxBoot. So if your location different - path need to be corrected. Inet/network addresses are used by internal interface between x86 and PPC. No ethernet cable was connected to either network ports on scope itself.
Now reboot scope.
VxWorks System Boot
Version: 5.3.1
 7 6 5 4 3 2 1 0
auto-booting...
Initializing backplane net with anchor at 0x1e80000... done.
Backplane anchor at 0x1e80000... Attaching network interface sm0... done
done.
Attaching network PC1 dio device driver installed
interface lo0... done.
Attaching DIO client... done.
Loading c:\vxBoot\vxWorks.st...1773516
Starting at 0x200000...
Initializing backplane net with anchor at 0x16e0000... done.
Backplane anchor at 0x16e0000... Attaching network interface sm0... done
done.
Attaching network interface lo0... done.
Adding 6008 symbols for standalone.
      ]]]]]]]]]]]  ]]]]     ]]]]]]]]]]       ]]              ]]]]         (R)
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]       Development System
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]       VxWorks version 5.3.1
 ]]]]]]]]]]]]]]]]]]]]]]]]]]       KERNEL: WIND version 2.5
 ]]]]]]]]]]]]]]]]]]]]]]]]]       Copyright Wind River Systems, Inc., 1984-1997
                              CPU: Tektronix HPP - PowerPC 740.  Processor #0.
                              Memory Size: 0x2000000.  BSP version 1.0.8.0 .
                             WDB: Ready.
usrNvFsInit: NAME=/nvram BASE=0xfd0ff100 BPB=512 BPT=8 NB=8 BO=0
nvfs: WARNING: Creating new NVRAM file system
tcsTableUnpack: checksum failed on buffer
tcsTableLoadFromNVRAM: tcsTableUnpack failed
tcsSynchronizeSensors:  ERROR:  Failed to create '/dev/tsAmbAcq'
tcsSynchronize: ERROR: Sensor synch failed
KERNEL: THERMAL CONTROL INITIATED
Executing startup script c:\vxBoot\dio_rst.vsh ...
dbFsIoctl: invalid code 3
    cd "pc:c:/vxboot"
value = 0 = 0x0
    tcsLoadTableToNVRAM "setup.tcs"
value = 0 = 0x0
    # hopefully you get this message at the end:
    # 'value = 0 = 0x0'
    # REBOOT!!!
Great, now PowerPC board boots (diag POST LED shows "P" after "L" which means boot image is running). Now change dio_rst.vsh script to original top-level script which was recovered from native Win98 HDD folder c:/vxBoot. I renamed topScript.hw to dio_rst.vsh to match boot settings:
    ################### topScript starts ###############################
     cd "dbfs:c:/vxBoot"
value = 0 = 0x0
    ld < TekScope.o
value = 23662048 = 0x1690de0
    ld < TekScopeSM.o
value = 23661408 = 0x1690b60
    pDbDev = ramDevCreate(0, 512, 1152, 1152, 0)
new symbol "pDbDev" added to symbol table.
... bla-bla ...
    rtlStartup
value = 0 = 0x0
   tsComp
value = 0 = 0x0
 Done executing startup script c:\vxBoot\dio_rst.vsh
KERNEL:  RUNNING TIME SLICED AT 10 TICKS PER SLICE!
Now scope application able to run successfully, and I got first waveforms on screen :).
Also seem like this particular scope got all options enabled, but some applications required for those options are missing, as I had installed fresh OS.
So system/software spec is:
CSA7404 (4GHz, optical hardware present) running Windows 2000 Pro SP4 Ru
TekScope version 2.5.3
Option 4M: 32 Meg Max 8M/Channel
Disk Drive Measurements 2 (Option 48)
Option 49
USB 2.0 Test package (Option 50)
Option 51
Option 52
Option 53
Option 54
RT-Eye Serial Compliance and Analysis
InfiniBand Compliance Module
PCI-Express Compliance Module
Pretty nice set got there for use in digital/PC hardware development and debug. Scope input interface for all four channels is proprietary TekConnect, and inputs are all 50 Ohm only. So to be able use regular passive BNC probes high-impedance adapter system TCA-1MEG is required, to use 50 ohm direct interface to target environment one can use TCA-BNC or TCA-292MM or TCA-SMA. As drawbacks of high-performance capabilities of such high-speed scopes special probing accessories are required, increasing total using cost. Passive probes cannot be used without TCA-1MEG.
Already got pair of Tektronix P7380 8GHz Differential probes, which are designed for high-speed signal acquisition with special solder-on tip system. Going to buy some lower speed single-ended active probes as well later.
Using this probe is really using for debug hardware, like DDR interfaces on VGA/DIMMs/MB's:
DDR3 DIMM clock measurement
RT-Eye report:
PCI-Express 1.0 report
Next steps:
* Design and assemble few custom probes for high-impedance environment
* Installing missing options software  :-DMM
Bonus:
Replace LEDs behind front panel indicators and push-buttons, just like in newer DSO/MSO series.  :-/O
To be continued...
Original wiki page on my site
Thanks for reading  8)
« Last Edit: December 10, 2013, 03:21:11 pm by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
Offline marshallh
    • retroactive
Re: Tek CSA7404 repair project
« Reply #1 on: September 16, 2013, 04:36:00 am »
Thanks for posting, neat to see it back in action.
Verilog tips
BGA soldering intro
Offline nctnico
« Reply #2 on: September 16, 2013, 07:51:07 am »
Drool...  A wooden casing would be neat  8)
Offline onlooker
« Reply #3 on: September 16, 2013, 08:30:19 am »
Thanks, for sharing an interesting story. May I ask how much did you pay for it and from where?
Online free_electron
    • SiliconValleyGarage
« Reply #4 on: September 16, 2013, 10:01:21 am »
Careful swapping that motherboard out. The motherboard uses a custom bios that allows the VXworks to become bus master and take control over the harddisk prior to boot !
Don't make the mistake of flashing a regular bios in there , it will brick the machine. I've been there and done that ! the correct bios can be found on the harddisk as well as the flash utility for it. a stock motherboard will not work. there are only two motherboards that work in this machine and the 810 is the most performant of the two. the other runs win2000....
let it be ...
The other annoyance is loosing the ram contents. it removes the serial number from the machine and disables all options.
There is a way to restore it but you need a gpib adapter and send a commmand to the machine to contact the security manager and give it the root key.
Online Vgkid
« Reply #5 on: September 16, 2013, 10:47:09 am »
That is a very nice writeup on the repair, very nice scope as well.
Offline marshallh
    • retroactive
« Reply #6 on: September 16, 2013, 12:14:11 pm »
Sooo.. can I send you a board for PCIe compliance testing :)
Verilog tips
BGA soldering intro
Online TiN
  • Country: tw
    • xDevs.com
« Reply #7 on: September 16, 2013, 08:39:13 pm »
Ha, to do so need to have much more than just 4G scope. Even if leave alone BER measurements, just CEM boards and fixtures cost quite a money :). And for PCIe Gen3 at least 16G bw required.
I got scope for just little over one K$, and main reason was because could not resist grabbing it , not because real need right now. Had no scope for personal use, just thought nice to have one at home.
Online TiN
  • Country: tw
    • xDevs.com
« Reply #8 on: October 05, 2013, 12:21:39 am »
Well, I got some update today, because received new goodies for scope.
Got 512MB and 256MB PC133 SDRAM DIMMs. x86 NLX motherboard refused to detect 256Mbit 512MB DIMM, so had to install 2x256MB into PC board
and 512MB on PowerPC board.
And CPUs:
It's Pentium III 1000E (1000/256/100/1.75V, Coppermine-256), original Celeron 850/128/100 from scope and pair of Tualatin 1400S (1400/512/133/1.45V).
Pentium III 1000E drop-in to socket and worked flawlessly, booted right in with 100MHz FSB, 1.75ish Vcore.
For those who want try upgrade their TDS7000/CSA7000's - beware to use 100MHz FSB chip, not 133MHz as such will run on 100MHz only, which will give only 750MHz (100*7.5 ratio = 750 for 133MHz native CPU).
But top glorious Pentium III-S 1.4GHz did not work so easily, Vcore was 1.3V, but nothing happen after power on...
Obviously Intel NLX motherboard in TDS7000/CSA7000 do not support Tualatin natively and I needed to do some surgery to bypass that  >:D
You can see already finished modified CPU bottom left on photo above.
What did I do:
1. Isolate pins AN3 (DYN_OE), AJ3 (RESET#2) from socket (I just desoldered pins right off CPU)
2. Isolate pin AK4 and connect it with AN11 via 1Kohm resistor (on CPU package)
3. Connect AM34 and AK36. (VID25mv signal tie to GND, to have correct Vcore VID operation)
4. Remove pins G35, G37 ? N37.
5. Connect G35,G37 and N37 together on CPU package.
That's it, after these modifications on CPU - it booted up and run smooth on 100MHz BCLK (native ratio on 1400S is 10.5, so 100*10.5 = 1050MHz).
Vcore voltage is correct too, 1.46V, so nothing to worry about.
So far only one thing left to do - receive covers and casing parts from Tektronix to completely restore outer looks on scope.
No more upgrades planned.
Online TiN
  • Country: tw
    • xDevs.com
Re: Tek CSA7404 repair project (Upgraded memory/CPU)
« Reply #9 on: October 09, 2013, 05:54:51 pm »
Got parts from Tektronix today :)
Offline leafi
« Reply #10 on: October 10, 2013, 02:14:46 pm »
so how does your scope run with the upgraded CPU? I might look into do this as my TDS7404 is slow as hell!
Online TiN
  • Country: tw
    • xDevs.com
« Reply #11 on: October 10, 2013, 02:35:34 pm »
Smooth and nice (but I think that's more because I added memory to 512MB total, rather than CPU).
Btw, I still have those CPUs, which I don't really need now ;)
So far:
New covers installed. Missing label for front panel and black cover for aux inputs on front.
Did not installed floppy drive (who needs them today?).
And missing some screws, need find them first.
So far spend around 2.5K$ on everything (including pair of P7380 probes), which is still quite a lot, but considering
this scope level - i think it was worth every penny :)  :-BROKE :clap:
UPD: Went cleap-nut and just printed labels on paper and stick them to panels.  :-DD
Now gotta hunt and wait patiently for good priced TCA-BNC adapters, TCA-1MEG high-impedance adapter. Already won pair of P6245 1.5GHz 1:10 active probes (for parts condition too, lol), and need think about iView (its just GPIB-USB dongle and couple BNC-BNC cables).
Tried iView on my TLA714 with NI GPIB-USB-HS and worklab TDS7704B, was nice neat feature to see everything in TLA software tool, eliminating need to mess with scope controls.
« Last Edit: October 10, 2013, 08:43:34 pm by TiN »
Online TiN
  • Country: tw
    • xDevs.com
« Reply #12 on: November 26, 2013, 12:56:32 am »
Just in case anyone needs drivers for TDS7000/TDS7000B/CSA7000/CSA7000B series scopes:
Display driver from TDS7000B/CSA7000B
USB Front Panel driver for TDS7000/7000B/CSA7000/7000B
PowerPC board driver for TDS7000/7000B/CSA7000/7000B
All drivers except display were tested on TDS7704B and CSA7404, with Windows 2000 Professional english.
Display driver is only for 7000B, i did not needed it on CSA7404
TekScope software itself 2.5.3 for TDS7000/CSA7000 (same as one posted on Tek website)
TDS2.5.3 TekScope
Offline CSmith
« Reply #13 on: December 02, 2013, 12:13:58 am »
BTW, I pulled the following out of a TDS7404 that had been under Tektronix maintenance and should be the correct default values.
                            VxWorks System Boot
Version: 5.3.1
 1
[VxWorks Boot]: ?
 ?                     - print this list
 l                     - load boot file
 g adrs                - go to adrs
 Boot flags:
[VxWorks Boot]: p
boot device          : sm
file name            : c:/vxboot/vxWorks.st
inet on backplane (b): 192.168.0.1:ffffff00
host inet (h)        : 192.168.0.2
gateway inet (g)     : 192.168.0.1
flags (f)            : 0x1008
target name (tn)     : tds7000
startup script (s)   : c:/vxboot/topscript.hw
[VxWorks Boot]:
« Last Edit: December 02, 2013, 12:17:06 am by CSmith »
Offline GreyWoolfe
  • NW0LF
« Reply #14 on: December 02, 2013, 12:48:07 pm »
 :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+  Awesome Job!
Tom, NW0LF
That which doesn't kill you still requires a co-pay.
Offline Khashoggi
« Reply #15 on: December 10, 2013, 10:39:01 am »
Great job!
Where can you find the system restore cd image? For a tds7054. There isn't one posted on the tek website.
Online TiN
  • Country: tw
    • xDevs.com
« Reply #16 on: December 10, 2013, 12:04:45 pm »
I did not had system restore CD image.
Just installed usual Win2000 OS, installed drivers for PPC board and USB Front panel (I had posted them earlier here) and downloaded TekScope 2.5.3 from tek site.
Offline CSmith
« Reply #17 on: December 10, 2013, 12:55:43 pm »
I'll vouch for the drivers that TiN posted are the same as what's included on the hard drive after a restore from original system restore CD's. I double-checked them to see if he happened to have anything newer than what I already had, and they matched. However for those who have expanded their memory to 512MB and would like to jump to Windows XP, I'd suggest hunting down the v2.51c "release" Windows XP driver for the proprietary Chips and Technology LCD interface. The filename is "wxpv251c.zip". The other drivers are generic enough that they run under Windows XP OK. Tektronix went through a number of different NLX motherboard vendors over the production run, so depending upon your particular motherboard, you may also need to track down the Intel Ethernet driver. Replacing it with the Intel SU810 NLX motherboard guarantees Windows XP built-in driver support for all of the on-board peripherals.
« Last Edit: December 10, 2013, 02:11:41 pm by CSmith »
Offline echen1024
  • 15 yo Future EE
« Reply #18 on: December 10, 2013, 01:52:18 pm »
Great job!
YOu won that auction did you? Lucky bastard.
I'm not saying we should kill all stupid people. I'm just saying that we should remove all product safety labels and let natural selection do its work.
https://www.youtube.com/user/echen1024
Offline Carrington
Re: Tek CSA7404/TDS7000 repair project (Upgraded memory/CPU)
« Reply #19 on: January 12, 2014, 05:02:27 am »
Hi TiN.
Do you know if the software Tek App V5.1 [TLA] has support for the TDS7xx series?
Thanks.  :-+
My English can be pretty bad, so suggestions are welcome. ;)
Space Weather.
Lightning & Thunderstorms in Real Time.
Online TiN
  • Country: tw
    • xDevs.com
« Reply #20 on: January 13, 2014, 02:42:18 am »
Seems so, at list TDS7xx scopes and TDS694 are in the list, but probably some functions will not be working (there is extra clock, which is shown as not used in drawing diagram). On TDS7xxx it goes to EXT REF input on scope, but those older scopes don't have EXT REF input.
Take this info as is, as I don't have ability to check if that combo actually work.
In worst case you still can use TLA with scope, just connect trigger output from TLA to scope trigger input (even non-tektronix is fine, lol),
and route analog outputs from TLA7AA4 into scope channels, and you will have signals from LA triggered on scope screen.
Just not as neat as having everything in same TLA app, that's it. (Even using iView one must adjust horisontal/vertical scale on scope itself, cannot use TLA app to do those adjustments).
Offline Carrington
« Reply #21 on: January 13, 2014, 04:28:53 am »
Ok, thanks.  :-+
But if the LA's "wfrm/s" (triggers/s) are greater than the oscilloscope's wfrm/s (rearm time), the oscilloscope will not be able to follow LA's triggers.
This is a problem, because: How do you know what oscilloscope's captures goes with what LA's triggers?
And if the above problem is solved the phase has to be adjusted too. So it's a mess, and better to use compatible instruments.
Because surely they have already thought of this, and have fixed this.
Or am I wrong?
« Last Edit: January 13, 2014, 07:13:50 am by Carrington »
Space Weather.
Online TiN
  • Country: tw
    • xDevs.com
« Reply #22 on: January 13, 2014, 07:56:01 am »
All TLA trigger setup is internal to TLA modules, and then you select output trigger to scope.
Then can use skew adjustment in scope, at least that's available in TDS/CSA7404 and 7704 i tried. Dunno about other models, but should be there too.
Even not adjusted, unless you watch something with hundreds of MHz, will be quite okay.
I shown example non-compensated iView capture here.
TLA7AA4 outputs connected to scope inputs (thru TCA-BNC) via mismatched BNC-BNC cables and SMA-SMA (thru TCA-SMA).
iView setup using recommended connections, and with generic BNC-BNC cables. Interface - NI generic GPIB-HS-USB connected to my remote machine
(as I run TLA only remotely, don't even have NI-VISA on mainframe itself).
Offline Carrington
« Reply #23 on: January 14, 2014, 01:28:39 am »
@TiN:
At least 2us long, in the image above. All was captured without pauses? Ie are not overlapping catches?
And what is the maximum length?
« Last Edit: January 14, 2014, 01:45:17 am by Carrington »
Space Weather.
Online TiN
  • Country: tw
    • xDevs.com
« Reply #24 on: January 14, 2014, 01:09:35 pm »
Capture on post i linked was around 100nS long.
One catch, yes. You have TLA, right? Your questions seem strange, as for owner of TLA frame, as that's
pretty much just TLA operation. :)