minifloat:

--- Quote from: Neomys Sapiens on December 30, 2022, 11:35:46 pm ---
--- Quote from: minifloat on December 30, 2022, 10:13:08 pm ---(...)
If someone needs the shroud, I can share collada or stl files. (...)

--- End quote ---
Please put the shroud .stl here:
https://www.eevblog.com/forum/testgear/replacement-knobs-feet-and-fittings-for-test-equipment
Would help me much with a TekProbe breakout box!

--- End quote ---
done so: https://www.eevblog.com/forum/testgear/replacement-knobs-feet-and-fittings-for-test-equipment/msg4611142/#msg4611142  :-+


--- Quote from: Mechatrommer on December 31, 2022, 07:22:49 am ---
--- Quote from: minifloat on December 30, 2022, 10:13:08 pm ---(...)
Attached you find some pics. (...)
br, mf

--- End quote ---
if BNC cable from that PSU output to scope is long, you risk reduced signal integrity (reflection), if its me, i will make an adapter from tek BNC input directly to normal scope's BNC output with power pins socket on the side, so PSU is external device with wires connecting to the adapter (which will be connected directly to the scope) much like tek 1141a diff probe with external 1142a psu module.

--- End quote ---

Yes I know - the P5205 requires the scope inputs to be set to Hi-Z.
I am using 30cm of BNC cable after my box, since I placed the Hacktronix supply right below my 'scope.

Using the original Tektronix 1103 supply with lengthy cable at the output would lead to the same problem :-//

br, mf