I agree about the switch. It will sometimes switch to the other setting while you're not paying attention. I'm pretty sure switchable probes existed back when HP and Tektronix were the major scope brands and were manufacturing probes in the US, but they were much more of a niche item back then. Probably bought for specific applications like field service. Often their performance was worse than contemporary 1x probes (switchable Tek P6062 had 7 MHz bandwidth with a 6 ft cable, while 1x Tek P6101 had 15 MHz with a 2 m cable). Somehow by late nineties they became the default probes for many scopes with up to 200 MHz bandwidth.
By the way, one thing to keep in mind about probe input capacitance: A higher input capacitance does not just increase circuit loading, it also makes the probe more susceptible to series inductance (ground lead or a wire between the probe tip and the circuit). Resonance frequency is inversely proportional to sqrt(L*C) after all.