Offline capt bullshot
One thing to note is that in those two ranges the X10 attenuator kicks in while in other ranges either there is no attenuator or only the X100 attenuator is in the signal path.
Any suggestions?
Yes, should be quite obvious: There's an issue with the X10 attenuator. Maybe a capacitor gone bad. The usual scope attenuator is a parallel connection of a high impedance resistive divider and a capacitive divider of the same ratio. If DC is OK, I'd believe the resistors are still OK, but if you've got a deviation at AC, there's a mismatch in the capacitors. Older Tek scopes had the capacitors as trimmer caps, while in these ceramic hybrids the resistors get adjusted during the manufacturing process.
I guess you'll have to remove the ceramic hybrid from the PCB to gain access to the capacitors, either they are underneath the relays, or on the other side of the ceramic.
Offline analogRF
Thanks @Jim Narem for the detailed response. Yes, I can see either of the optocouplers can cause the offset problem
right now the offset is small and I can deal with that later but my major concern is the messed up frequency response of channel 1
it  seems the input capacitance changes to wrong value in two vertical ranges when the X10 attenuator kicks in
my suspicion is also the relay (the middle one) but I cannot quite figure out how/why the relay can cause
capacitance imbalance? I can see it might cause unwanted contact resistance but capacitance? how?
on the other hand, as you said, there is not much else on the attenuator board since the elements are all laser etched or something
It's also hard to imagine one of those integrated capacitors has changed its value over time
Offline analogRF
Any suggestions?
don't you think the capacitors are also integrated inside the ceramic pcb? I mean like they are eatched into the circuit board?
if they are underneath the attenuator, there  is hope !
also, can it be due to the relay as @Jim Narem suggests? Although i cannot figure out how the relay can cause imbalance capacitance
Offline analogRF
....I think this might be how it works: there is a series C (or RC) between the input and output of the relay. when the X10 atten is out, the relay is closed and that series RC is short circuited and when the X10 atten kicks in, the relay is open and the series RC (or C) between the I/O of relay comes in to the signal path (parallel to the series branch of the attenuator ). So, that capacitor is probably at fault here not the relay itself.
If that capacitor is a discrete component underneath the hybrid, i might have a chance but i don't know what the value should be.
this problem seems very similar the issue people have had with Siglent SDS1202X-E early models for which a remedy was also discussed extensively on the forum but I dont know how exactly they modified the circuit....i do have that scope myself and it has that issue but I didn't care to fix it. never affected my work really...
Offline Zucca
  • EE meid in Itali
The following users thanked this post: Tiberius1
I've never taken a look into the shielded attenuators of these THS scopes myself, so I just was guessing. One can see a few capacitors soldered to the ceramic, especially the big one for AC coupling as I suppose. I've replaced the relays in my TDS580 attenuators (found some pictures here: https://www.eevblog.com/forum/testgear/tektronix-tds520b-attenuator-swap/ ), here one clearly can see the large metallized areas - these are the capacitors for their attenuators, see the laser trimming. I don't know wether the THS ceramics are made the same way or they've soldered discrete capacitors. For sure, one can see laser trimmed resistors there (Edit: looked again at the high res picture, looks like a trimming scheme applied to whatever is hidden under the blue coating, can't tell if it's a capacitor or resistor).
From my opinion / experience, a weak relay should cause DC errors. Within the TDS 5xx / 7xx attenuators, weak relays cause SPC errors, and/or complete loss of signal for some attenuation settings. Exercising the relays can cure that for a while.
« Last Edit: September 11, 2018, 01:18:56 pm by capt bullshot »
Offline analogRF
You know we have the schematics, don't you?
https://www.eevblog.com/forum/testgear/tektronix-ths700-tekscope-component-level-information-package-and-schematic/
of course. But my frequency response problem is inside the attenuator hybrid (marked as AT1 on schematics) so the schematics don't help ....see the pictures posted by @Jim Narem  at http://w140.com/tekwiki/wiki/THS720
Offline analogRF
@capt bullshot: I also believe the caps are etched into the ceramic. I highly doubt there is anything underneath that board.
SPC passes on my scope with no problem and DC is perfectly fine on all vertical ranges. Only when I apply a square wave with long enough period I can see on those two ranges the pulses have a exponential riding on top of them with overshoot and then settling down to the correct value.
Scope calibration didnt help either as most of the process only uses DC voltages.
I wonder what might have happened to a ceramic etched capacitor to go bad?
Overvoltage? Environment (humidity)?
One of the laser cut links may have gone short again for whatever reason?
I've had some defective ceramic caps in an old 7 series plug-in: http://wunderkis.de/tek-7/ Don't know what caused them to go bad either.
Online Ice-Tea
    • Freelance Hardware Engineer
Joining the club: I got 6!  :clap:
First job: batteries!
eBay shop with all the gear you need!
FS: Agilent 54825A, R&S CMU200, CRTU, Marconi 6201B, Lecroy WP 950, 9354TM, 9354M, 9374L LC584A, Tek THS720P, TDS7154B, Anritsu MG3671A 2.75G I/Q RF gen, Keithley 238 SMU 707 matrix, HP 8642B, 8903A, 8110A, 8156
Offline HighVoltage
Wow, are all 6 of them THS720P?
Are they supposedly broken or in working condition?
Offline Zucca
  • EE meid in Itali
Holly Molly... welcome to the club.
I suggest you to build some DIY LI ION 2S battery packs, as discussed before here. If not I can sell mine NiMh pack used 2 hours in total.
« Last Edit: October 09, 2018, 11:38:47 am by zucca »
Online Ice-Tea
Yes, all 720P. Should be all perfectly fine, clearly labelled "excess", not broken or anything. And not a scratch on them. Just ordered a stack of 4500NiMh's  :clap:
Offline Zucca
  • EE meid in Itali
Online Ice-Tea
Works like a charm!
Online Ice-Tea
First unit on the Bay. This one includes 2 P5102's and a funky Fluke RF probe!  :popcorn:
Offline Maxis
Gentlemen,
I've got TEK THS720 without A. I wonder whether by a simple (nor not so simple) FW change it can be converted to THS720A. By looking at the main board is seems like the main components are the same (even down to the ID resistors in question). Have anyone done anything like this?
Thank you!
Maxim
Offline hardcoreman
  • Country: sk
same idea... but no free time to test.... i think software will run.
Offline heize0
  • Country: cn
Has anyone still interested in topic? I have a broken THS710A due to flash memory's content being corrupted and stop from booting. I use the firmware of V1.08 been uploaded here to repair it. Now the machine can boot and operate. But it still seems to have some problem. When the time scale change from 500u/s to 200u/s and so on, the base line comes to be huge thick almost near 1div. I had checked the broken flash memory. It had the firmware version of V1.11, yet some content was different every time been read. Does anybody has newer version firmware?
Offline heize0
  • Country: cn
Finally, problem got be solved. It was not caused by the version of firmware, but due to uncalibrated. I did it and then everything went well.
Offline superman13
  • Country: md
heize0 Do you have dump of V1.11 firmware?
Offline heize0
  • Country: cn
Yes. But it may be wrong because this is dumped from the one i repaired that could not boot.
Offline TekWal
I need some advice from the forum.
Just got a nice THS720A for my daughter project. It is experiencing some weird power up behavior.
The scope works fine but occasionally, it will just refused to boot up. Pressing the on/stdby button will just light up the backlight and nothing happened. I need to unplug the AC adapter and then plug in again with the on/stdby button pressed. After that it will works fine. Self-test PASS and there is not error messages on the error log status. Whenever the problem return, I will have to do the AC adapter unplug then plug-in + on/stdby button pressed fix. Just to add, there is no battery in the scope.
Anyone experienced this problem?
Anyway comments/feedback is much appreciated.
Wally
« Last Edit: April 08, 2019, 02:07:21 am by TekWal »
Offline HighVoltage
Are you using the original TEK charger?
I had the same issue, when I used a cheap switch-mode charger and no battery in the scope.
The original TEK charge is a heavy linear charger in good quality and no leakage.
Online mk_
  • Country: at
I had no problems with any of my 3 different THS7xx -oscilloscopes with any charger - switchmode or linear - as long as it can deliver the required power.