Author Topic: Tektronix 016-0829-00 Blank Plug-In for 11K Series  (Read 604 times)
Offline calibrationfixtureTopic starter
« on: May 20, 2024, 07:13:33 am »
Hi to All,
Seeing some Lab Tours on YouTube I noticed Tektronix 11K Mainframes working without the empty slots populated with these blank filler Plug-Ins. These are crucial in directing the airflow. These Blank Plug-Ins are only designed with this in mind. Not to create a platform to make your own Plug-In like the Kits for the 7000 and TM500 Series. Those have PWB's. The 016-0829-00's didn't have these and they never existed.
https://w140.com/tekwiki/wiki/016-0829-00
Calibrationfixture
« Last Edit: May 21, 2024, 05:23:00 am by calibrationfixture »
Offline coromonadalix
Re: Tektronix 016-0829-00 Blank Plug-In for 11K Series
« Reply #1 on: May 20, 2024, 01:06:59 pm »
some places offer them in USA,     but at 95$ USD ???
Offline kbrill
« Reply #2 on: May 21, 2024, 03:26:32 pm »
Not sure if this response is appropriate , if not advise
I have a very clean 016-0829-00
Offline Testtech
« Reply #3 on: May 22, 2024, 11:36:07 pm »
You can also use the 7K blank filler, it is just a front with plastic guides, works fine. 016-0155-00.
Offline edavid
« Reply #4 on: May 23, 2024, 12:05:40 am »
Is there any evidence for this idea that the filler panels are required?  I have not noticed any temperature increase with my DSA602 or 11302 with some plugin slots empty.  There is not a lot of air leaking out either.
Offline colorburst
« Reply #5 on: May 23, 2024, 01:00:48 am »
The 11302 manual says the following on page 1-5:
A variable speed fan controls the air flow as temperature varies. To maintain
adequate cooling, all plug-in compartments must have an 11000 series plug-in
unit installed. If active plug-ins are not available, install blank plug-in units
(Tektronix Part No. 016-0829-00). Failure to comply with these ventilation
requirements violates conditions to which the advertised specifications (i.e.,
accuracy) apply.
And later on,
Blank plug-in units must be inserted in unused compartments to maintain proper
ventilation and compatibility with electromagnetic emissions,
susceptibility and accuracy specifications. Refer to the Optional Accessories
list in Appendix B at the rear of this manual for ordering information.