Super Thread, has always really helped.
I purchased a defective Tektronix 2457B with Option 9 and 10. Error was, that the device turned on and off again and again. Even the smell showed that electrolyte must have leaked out. The device stank like a trash can.
After I opened the device, I discovered the leaked SMD electrolytic capacitors on the A5 controller board. The traces were not attacked and also neighboring components were still OK. After I had cleaned everything and used new Elkos, the device worked again. The error did not occur anymore.
Then my mishap: I soldered out the NVRAM chip to read it out. From now on I made all the mistakes you can only make.
When I put the chip in Programmer, which I only had a few hours, I just wanted to test the chip and erase all the data.
All data gone.
OK, I had the video of Exerciser 02 for safety's sake. The video quality was so bad that it's a miracle that I was still able to write all the data to the hex editor. But the device did not start now, Trigger'D flickered happily, the device complained about corrupt calibration values. Ok, I stayed calm, did not smash the device, I knew I could find the mistake.
I read everything there was to read, read the service manual, nothing. Then I came up with the idea of looking at the solder joints of the NVRAM under the microscope and discovered that I had severed a trace directly at the solder pad while soldering out.
That's never happened to me. I am getting old.
After I fixed that, everything work again. A great device, I've never seen such a good Oscillscop before.
A problem that was already there before: The display, but only the beam, flickers slightly and slow, not always and only in conjunction with my digital function generator. Not for example with my old analog pulse generator. However, the signal of the function generator does not flicker with the DSO and also not with the old Tek 475. The readout I had switched off, even with Cal 08 set the display. All voltage values on the board are OK and look clean, ref values of the DAC are -1.248 and 1.363.